Hello oscilloscope friends
I plan to show the rise times of a HP8133A by Tek 11403 (11A72, 1GHz), TDS 820 (6GHz) and Tek 11801 (with SD-20, 20GHz).
Why this?
The HP 8133A can provide pulses between 33 MHz and 3 GHz frequency,
with up to 3.0 Vp-p amplitude and transition times of <60 ps (20% -80%).
It is at least interesting to me which oscilloscope can show the form of pulses, and how close to reality.
I hope that these experiments are not considered absolutely trivial and can be of interest.
I am curious about your comments, tomorrow morning I will start
regards
Physikfan
Online edavid
Re: rise time of pulses of a HP8133A by Tek 11403, TDS 820 and Tek 11801
« Reply #1 on: September 16, 2017, 09:26:38 pm »
Don't we use 10%-90% here?
How will you interconnect the HP8133A and the scopes?  (I would suggest a short hardline.)
« Reply #2 on: September 17, 2017, 07:18:16 am »
Hi edavid
HP specification for 10% up to 90% amplitude (rise time) is <100 ps.
I will not use a rigid cable, because I want to use the same cable for all three oscilloscopes, for the 11402 I will use an adapter SMA to BNC.
But I will try to find out a possible difference between a flexible and a semi rigid connection.
« Last Edit: September 17, 2017, 08:16:35 am by Physikfan »
Offline PartialDischarge
« Reply #3 on: September 17, 2017, 07:33:57 am »
You should visit "show us your square wave" !!
https://www.eevblog.com/forum/testgear/show-us-your-square-wave/
« Reply #4 on: September 17, 2017, 03:21:11 pm »
Hi mastertech
Thanks for giving me the hint to visit "show us your square wave".
I have learned a lot.
Unfortunately most of the pictures were not readable for me:
"please update your account to enable 3rd party hosting
for important info, please go to
www.photobucket.com/P500"
Regards
Physikfan
« Reply #5 on: September 17, 2017, 08:29:38 pm »
These are the first results:
Square wave of the HP8133A with the Tektronix sampling oscilloscope 11801C, x-axis 50ps / div, y-axis 200mV / div, with head SD26, rise time 17.5 ps, bandwidth 20 GHz.
200 ps wide pulse of the HP8133A with the Tektronix Sampling oscilloscope 11801C,  x-axis 50ps / div, y-axis 200mV / div, with head SD26, rise time 17.5 ps, bandwidth 20 GHz.
The displayed rise time is about 50 ps in both cases.
« Last Edit: September 18, 2017, 08:55:01 am by Physikfan »
Offline David Hess
  • DavidH
« Reply #6 on: September 17, 2017, 11:20:47 pm »
Sometimes you have to use 20%-80% because 10%-90% is unrealistic when the cable suffers from dribble up.
« Reply #7 on: September 18, 2017, 08:24:16 am »
Here are the results measured by a TDS820, 6 GHz oscilloscope with 44 ps rise time:
Square wave of the HP8133A,  x-axis 50ps/div, y-axis 200mV/div:
200 ps wide pulse of the HP8133A,  x-axis 50ps/div, y-axis 200mV/div:
The displayed rise time is about 65 ps in both cases.
« Last Edit: September 18, 2017, 01:53:04 pm by Physikfan »
Offline Leo Bodnar
« Reply #8 on: September 19, 2017, 08:05:04 pm »
Why don't you use internal 11801 risetime calculation?
If you select the risetime measurement then tap on "Rise" rectangle then you can adjust Proximal and Distal values to 20% and 80% using the knobs - like below.
Let the machine do it for you.  Tek engineers designed 11800 series 30 years ago so it's time we finally figured out how to use it.
Cheers
Leo
« Last Edit: September 19, 2017, 08:11:11 pm by Leo Bodnar »