Hi Group,
Here are some pictures from my experiments:
First I have a Fluke 87 connected to a 0.47uF Film capacitor. The more experienced readers will recognize the Phillips (Mullard) capacitor. The meter reads the right value.
Then I connected a 1M Ohm metal film resistor. The meter is reading 0.147 uF
So I tried the 1M resistor in series with the capacitor and I got 0.098uF:
If I check (C
1 x C
2) / (C
1 + C
2)
I get (450n x 147n) / (450n + 147n) = 110nF
This is close to the 0.098uF measured.
I then tried the parallel combination:
I got 0.59uF
I then tried the same experiments with a Fluke 289:
First the 470nF capacitor:
I tried the 1M resistor but I did not get a stable reading, so there is no photograph.
I then tried the 1M and 470nF in series:
I got a reasonable reading for the capacitor.
HP4274A LCR MeterThe HP 4274A is a very accurate LCR Meter that was probably introduced in the late 1970's but a very capable instrument.
Here is the open circuit reading after zero calibration:
That is zero fF. a fF is a femto Farad 1
-15 F or 1/1000 of a pF
Here is the same meter with the 1M Resistor in the test fixture.
The value is 160 fF or 0.16pF. This is very similar to the 0.11pF predicted by Ian.M in an earlier post. There are reasons why the value is higher. One is the metal end caps on the resistor construction reduce the distance.
Comparing the measurements from the various meters, I would conclude that even some very good DMMs are fooled by a high value parallel resistor into giving erroneous readings. The more specialized LCR meter give the correct answer. The correct answer matches the calculated value of the capacitance from the physical dimensions and the material properties.
I learned from this thread to be careful measuring capacitors with a DMM.
Regards,
Jay_Diddy_B