Hi,
I just tried a couple of my meters.
First a Fluke 87I took at 1 Meg 1/4 metal film resistor and a 0.47uF film capacitor.
I measured them individually, in series and in parallel.
If I measure just the capacitor I get a reading of 450nF which is close enough.
If I measure the 1 Meg resistor I get a reading of 140nF
In parallel I get 0.6uF and in series I get around 0.1uF
All of which seems about right for series and parallel capacitors.
However, the Fluke 87 Manual says:
Fluke 289I measured the capacitor first. I got a reading of 452nF.
I tried to measure the resistor on the capacitance range, I got an unstable reading.
I tried the series combination. The Fluke 289 reads the correct value of capacitance, even with a 1 Meg resistor in series.
This just shows, that some meters don't work with a parallel resistor when measuring capacitance.
I will try the same components on an LCR meter, HP4274A, but it is too late....
Regards,
Jay_Diddy_B