Author Topic: Ultra Precision Reference LTZ1000  (Read 1682206 times)
0 Members and 155 Guests are viewing this topic.
Offline branadic
  • www.qoool-sensing.org
Re: Ultra Precision Reference LTZ1000
« Reply #275 on: August 05, 2013, 08:09:15 pm »
http://www.edn.com/design/other/4326640/DC-accurate-32-bit-DAC-achieves-32-bit-resolution
Interesting about that is, that on volt-nuts the mistakes in the articel were never really identified and that W Stephen Woodward, Chapel Hill, NC; Edited by Martin Rowe and Fran Granville never showed a picture of the circuit (pcb-layout or photo of the pcb) or sold ready to use pcbs. Does anybody have the full text article?
@ DiligentMinds.com
You might want to draw your idea into a real world circuit?
« Last Edit: August 05, 2013, 08:58:26 pm by branadic »
Measuring is like guessing, but more advanced.
Offline Andreas
« Reply #276 on: August 06, 2013, 06:15:10 am »
Hello together,
there are other articles from Woodward who explain the circuit more in detail:
http://www.edn.com/design/analog/4323340/Fast-settling-synchronous-PWM-DAC-filter-has-almost-no-ripple
http://www.edn.com/design/analog/4329365/Combine-two-8-bit-outputs-to-make-one-16-bit-DAC
formerly it was possible to download the whole article as pdf and not only the schematics.
I have built the 32 Bit DAC cirquit mentioned above with slight variations. And could by far not reach the the values mentioned in the article.
That what I got is around 4 ppm INL (ok maybe only 3 ppm since my ADC is only linearized to around 1 ppm).
But when thinking twice: Woodward does not mention the switching times in his accuracy formulas neither he mentiones how he gets to the 23 bits as the effect of charge injection of the switches.
R7 has to be carefully adjusted to reach the 4 ppm INL. And naturally should be tracked with the chip temperature of the MAX4053 chip. With a simple 5.1 Ohms Resistor INL was around 13 ppm (65uV change) as can be seen on the diagrams of June 28th in this thread.
Also noise with 2uVpp only seems to consider the OP-Noise of the chopper. And not the additional effects of the switches and the error of the integrator producing some kind of "staircase noise" with much larger amplitude (around 15uVpp measured with a 4th order 10 Hz low pass filter) in my case.
So all in all the circuit idea is not bad. The settling time is excellent. But there is much room for enhancements.
So I for my part will rely on a 2 DAC solution with a precision ADC in loopback as in AN86.
With best regards
Andreas
Offline Mickle T.
  • Country: ru
« Reply #277 on: August 06, 2013, 07:37:36 am »
This article is like a joke. Even ancient (30-years old) Datron 4000 calibrators have a two PWM DACs and provide 0.03 ppm of internal resolution with 0.1 ppm INL. But schematics is far more complicated, "The devil is in the details."
http://www.ko4bb.com/Manuals/09)_Misc_Test_Equipment/Datron/Datron_D4000A/Datron%204000A%20Ref%20Div%20Text.pdf
Offline Dr. Frank
« Reply #278 on: August 06, 2013, 10:03:56 am »
Well, I think, that the steering and auto-calibration software for that dual-DAC concept is substantial to get low INL figures.
A very good description is contained in the service manual of the Fluke 5440/5442, which use also this same approach.
These instruments (like its successor, the actual 5720A) have an internal resolution of 27 bit (0.005 ppm), 0.1ppm front panel resolution, and 0.5 (0.3) ppm +/- 1.5µV output linearity.
The trick is, not to use very low PWM rates (<10 units) on the upper DAC, this nonlinear region is covered by the lower DAC.
All offset and gain errors /drifts are measured and compensated first, and then only those ultra linear specifications are possible, but realistically not much below those given linearity figures.
Therefore, claiming 32bit resolution for a DAC sounds great, but is totally nonsense for physical reasons.
To be checked is the characteristics of the much more simply built DYI divider, i.e. if it's possible to to achieve < 1ppm INL or output uncertainty by appropriate steering of both DACs.
I have measured the linearity of my 5442A by means of the 3458A, and present them here in two versions:
The first version is the linearity referred to the input (13V = 2 x SZA263), that should be the same as the INL, and is well below 0.05 ppm.
The second calculation is related to the output, and tells you, what uncertainty of output you really get.
(That's the specified linearity parameter of 0.5ppm + 1.5µV for the 5440A.)
It's clear, that this parameter diverges towards zero output (covered in the spec by '+/- 1.5µV'), but between 1V and 11V, the deviation is less than about 0.3ppm (w/o offset).
This level of uncertainty is really necessary for usage as a precision divider in a LTZ1000 based reference, otherwise a resistive divider would be more precise.
Frank
« Last Edit: August 06, 2013, 10:13:15 am by Dr. Frank »
Offline bingo600
« Reply #279 on: August 06, 2013, 01:47:38 pm »
Hello together,
Could be those attached ?
/Bingo
Offline Dr. Frank
« Reply #280 on: August 06, 2013, 08:40:01 pm »
This design can be further improved by having more PWM circuits that divide the ~7.2V to control the oven temperature of the LTZ1000(A), and the temperature of the multi-stage Peltier refrigerator that houses the entire circuit.  If the refrigerator is set at 0-deg-C, then we can run the LTZ1000(A) at 10-deg-C, and (hopefully) it will show very little drift in comparison to current state-of-the-art voltage references.  WIth the microcontroller in control of the LTZ oven and the refrigerator temperatures, the temperature cycling (ala the Pickering patent) can be used to condition the entire circuit if there is ever an extended power failure.
This is about as far as I have come with this design.  I am doing some experiments with an LTZ-based reference to see if I can coax the device to operate in a mode where "very good" temperature compensation is achieved at the chosen die temperature-- (which is not easy, as the data sheet is devoid of much needed information).
A Peltier element, great idea, at least this should make the Pickering patent work, finally..
Have you mentioned those bugs in his patent, i.e. that asymmetrical temperature interval, which will not work theoretically for 'degaussing'? Neither in practise, in the 7001, as 45°C is far too near room temperature.
I still think, the 7001 box was not successful, as the hysteresis effect is so low for a power down situation, or buried under other effects, that this technique does not give real advantages.
Below 20..30K temperature intervals, the hysteresis is very small.
How do you plan to realize the duty cycle programming?
Any calibration features?
Btw: The famous Datron 4910 reference, doesn't this device contain PWMs for the 7.2V => 10.0000V transfer also, instead of a resistive divider?
How's this realized?
Frank
« Last Edit: August 06, 2013, 08:46:52 pm by Dr. Frank »
Offline Andreas
« Reply #281 on: August 06, 2013, 09:27:13 pm »
The 500KHz is used on the final output to provide a gain of 2.
Really 500 kHz? = 2000ns.
how much will the break/before make time > 15 ns and the difference between rise and fall time (30-40 ns) affect the 2:1 ratio?
One other question: Why do the calibrators use 125 Hz as PWM. It's neither a multiple of 50 nor 60 Hz line frequency and might give a beat frequency with both NPLC integration times.
With best regards
Andreas
Offline Mickle T.
  • Country: ru
« Reply #282 on: August 07, 2013, 10:26:50 am »
No problem. 4910 schematic is well-known.
The following users thanked this post: guenthert
Offline Dr. Frank
« Reply #283 on: August 07, 2013, 11:17:39 am »
Mickle,
thank you very much!!
Partly I've seen the schematics, as I remember now..
The content of the Vishay resistor array still is not known yet?
Frank
Offline Mickle T.
  • Country: ru
« Reply #284 on: August 07, 2013, 11:26:11 am »
PIN
1-2:48.7795K  1-3:48.6536K  1-4:48.6543  1-5:47.9947K  1-6:38.9950K  1-7:38.9952K  1-8:39.0165K
2-3:127.422ohm  2-4:127.326ohm  2-5:786.248ohm  2-6:9.78619K  2-7:9.78629K  2-8:9.80740K
3-4:0.3380ohm  3-5:659.267ohm  3-6:9.65917K  3-7:9.65927K  3-8:9.68048K
4-5:659.163ohm  4-6:9.65910K  4-7:9.65920K  4-8:9.68033K
5-6:9.00017K  5-7:9.00028K  5-8:9.02140K
6-7:0.3355ohm  6-8:21.5500ohm
7-8:21.4520ohm
Offline quarksTopic starter
« Reply #285 on: August 07, 2013, 11:28:01 am »
which one do you mean (part# 315532)?
« Reply #286 on: August 07, 2013, 12:02:44 pm »
PIN
7-8:21.4520ohm
I guess that answers my question.
Your meassurement looks close to what it really is (an array of 8 resistors some in series and some parallel).
If there is interest I can probably deliever all details you like to know, because I have the original datasheet somewhere.
I also was in contact with VPG and they offered me to make my own personal part number, if I want this obsolete part.
But I stopped there, because I do think this is not as good as you can do today.
« Last Edit: August 07, 2013, 12:14:10 pm by quarks »
Offline Dr. Frank
« Reply #287 on: August 07, 2013, 01:05:56 pm »
you have the original datasheet of that Datron specific part.
Well, should be 5 different resistors only, i.e.
21.2, 127, 39k for the reference, 21.2 for improvement of T.C., 127 for the 4mA Zener current, 39k as collector resistor.
Then, 9k over 659, as divider for the heater control, but regulated additionally from outside the module.
So, the complete schematic may be sketched in a nicer way.
Also, the points which have that guard ring, might be drawn in a schematic... I still would like to understand, what Datron has done there, and which influence this might have.
I agree, todays components have improved a lot, especially the oil filled resistors are superior, and allow a more stable design also without the elaborate heater control in the 4910.
Anyhow, the schematic is interesting, for improving the sensitivity of the regulation part. Datron has included some caps there..
Frank
« Last Edit: August 07, 2013, 02:49:37 pm by Dr. Frank »
« Reply #288 on: August 07, 2013, 01:54:52 pm »
Then, 9k over 659, as divider for the heater control, but which regulated additionally from outside.
Frank
there are 8 resistors in 4 diff. values (5x 64R, 659R, 9k and 39k)
Tolerance of the individual resistors is only 1% and abs. TCR is 3ppm/K.
Tracking is mentioned to be 0.5ppm/K for all 64R and for the others only 4ppm/K
my copy (Rev E from 11/92) is not very good but see the schematic
there is nothing about the guard
Besides the surprise that it is not as percise as I guessed, I did not really understand why it is as it is.
Maybe you or Mickle or anyone else can explain what could be the idea behind it.
bye
quarks
edit: I just saw, I am probably missing pages, because I only have page 1/4+2/4
« Last Edit: August 07, 2013, 02:09:43 pm by quarks »
Offline branadic
« Reply #289 on: August 07, 2013, 07:24:14 pm »
Wouldn't it be enough to use two 16bit pwm, one for pre-dividing and the second for boosting it to 10V?
Offline Andreas
« Reply #290 on: August 08, 2013, 05:04:09 am »
The analog switch I am looking at is the Analog Devices ADG1419-- which has 2.1-ohm switches, and will transition in about 200nS.
Did you already do some linearity checks with the ADG1419 for your 2 first stages?
How does it compare against the results (13 ppm INL un-tweaked and 3-4 ppm INL tweaked circuit ) that I got with the MAX4053A? see Diagrams (x-axis digital code y-axis error voltage in mV against 5000 mV full scale) of  June 28th in this thread?
With best regards
Andreas
Offline Andreas
« Reply #291 on: August 10, 2013, 02:06:45 pm »
I have only done simulations at this point,
Hello,
which simulation program do you use?
and from where do you get the models of the ADG1419?
I am not the simulation expert. And doing simulations mostly with LTSPICE.
My experience with simulations is that the models are only simplified.
Simulation of deviations below the 100 ppm level seem for me to be
very hard to simulate with a spice based simulator.
with best regards
Andreas
Offline branadic
« Reply #292 on: August 10, 2013, 03:54:20 pm »
Looks like he used LTSpice. My guess is, that he used a simple Voltage Controlled Switch model behind the ADG1419 since there is no spice model available from Analog Devices.
« Last Edit: August 10, 2013, 04:00:48 pm by branadic »
« Reply #293 on: August 12, 2013, 10:08:25 am »
because of new interesting discussions here, I kind of paused my DIY LTZ1000 project, to wait and see what I can possibly adopt.
But yesterday I saw something totally unplanned but interesting hapen.
It is topic related, because the used gear (a calibrator and two DMMs) has internal LTZ1000 references.
Here is what I did/observed:
- all three units were started at the same time (no warm up time at all)
- all were connected together and setup to measure 10VDC with 6 digit resolution (for fast reading)
- no adjustments (like Zero or offset corrections) were made
Result was stable 10.000 00 V reading from a few minutes after start until around 2 h when I stopped
This morning I repeated it but with 8 digit resolution and this time with Peak to Peak (PKPK average stat) monitoring running.
Start value was almost spot on 10.000 000 0V
difference between the two DMMs readings was most of the time <1µV (=0,1ppm of 10V)
Overall PKPK was around +10µV  (=1ppm of 10V) running >2h
Final value after >2h was around 10.000 010 0V
Ambient 24,3°C to 24,9°C at 57% rel. humidity during measurement
That altogether is an agreement of within 1ppm of all three units from switching on to >2h.
BTW the gear is of different age and origin and the DMMs are not tweaked/adjusted/calibrated to show this behavior.
Besides I am very pleased with that result, I really wonder how this can statistically be.
Is this possibly only by accident/random, although I alredy repeated it?
Or is it really possible that the warmup drift of all three units almost match identical?
If that could be the case, in future I could save a lot of time, because so far (for serious measurement) I always waited the recommended at least 4h to warm up the gear, before I started to measure anything. But with that finding, I would only have to make sure to start at the same time and could expect to be within 1ppm.
bye
quarks
« Last Edit: August 12, 2013, 04:24:48 pm by quarks »
« Reply #294 on: August 12, 2013, 05:14:39 pm »
Can you tell us what equipment you have?
In the test I used Wavetek 4808, 1271 and Fluke 8508A
« Last Edit: August 12, 2013, 05:16:41 pm by quarks »
Offline branadic
« Reply #295 on: August 12, 2013, 05:58:18 pm »
Can someone please link to the "Pickering Patent", would be great if all mentioned patents would be linked so it's easier to follow.
Offline Dr. Frank
« Reply #296 on: August 12, 2013, 08:48:46 pm »
Hello DiligentMind, nice thoughts!
Just turning off the 3458A, and waiting 1 day, then turning it back on (and waiting 4 hours), then doing an auto-cal, I noticed that the reading at 10V can be different by as much as +/-0.2ppm.
I see the same fluctuations of 0.2ppm in the monthly  measurements of 3458A vs, 5442A, as I also switch them off. (Energy is very expensive here in Germany).
That's perhaps not a pure hysteresis effect, as the difference drifts forth and back.
Pure hysteresis would add up or stick to fixed output values, as both oven see two different temperatures only, i.e. RT and 65°C only.
It is much more difficult to get the last 2 digits to behave well, but the next to the last digit can be made to be very stable (long term and through power cycles) if the room temperature is kept at 23C +/-0.5C, and there is at least a 4-hour warm-up period before I auto-cal.  So, with care, 7.5 stable digits are possible-- the last digit is always bobbling around-- and I have an idea that it is DC-10Hz noise from the 732B (combined with the DC-10Hz noise of the DMM).
My measurements (10min stability) of the 5442A and of the DYI LTZ1000 also show jitter on the last digit only, i.e. on the order of 0.01ppm, @NPLC100, const. temp.
I also relate that to the Zener noise only, after I improved the circuitry by better PSU and shielding. The 3458A itself should be more stable / less noisy than that.
That in itself makes me confident that if the DMM manufacturers put a carefully designed LTZ1000A based reference in their 6.5 digit DMM's, then they could be made stable to 1ppm per year (ie., the last digit would change no more than +/-1 per year) on their best DC accuracy figures.  I would pay a lot of extra money (up to $500 extra) to have that in a 6.5 digit DMM.
Well, I prefer the LTZ1000 @ 45°C, and this reference can be built much better than HP by yourself for about 100$.
Indeed, using highly selected components together with the special hand tuned wire-wound resistors in the boost circuit of the 732B are key to the unit's stability.  I'm not certain that these techniques could be applied by the home experimenter-- you need a vast inventory just for the selection process, and this can become very expensive very quickly.  ...
To my opinion and experience, instead of hand made / selected wire wound resistors, those hermetical oil filled VHP201Z will do the same job, perhaps even more stable.
Note that resistors will have double the drift rate for each 10C rise in temperature-- so keeping them cool (but at a constant temperature) is best.  How cool?  I don't know-- but I imagine that there is a point of diminishing returns-- so there is probably an optimum temperature to keep all of the circuitry at.
Room temperature is sufficient for 2ppm/6yrs for the VHP201Z resistors.
And that level of stability is over sufficient for the LTZ1000 circuitry.
Any heating or cooling will lead to more problems.
Another problem is temperature hysteresis-- and this occurs in the LTZ1000, as well as any op-amps that you are using-- and as Dr. Frank has proven-- in foil resistors too.  (I'm not certain what hysteresis effects there are in wire-wound resistors, but I am willing to bet a premium beer that they at least have some hysteresis effects).
The hysteresis effect of the metal foil resistors is relatively low (5ppm after 125°C), and due to the 100:1 attenuation effect, does not play a big role.
On +/- 15K, there is no appreciable hysteresis, so if you take care, there's no problem at all.
Afaik, the wire wound types have neglect-able hysteresis effect, only the tightness of the winding has to be relaxed first.
Some time, we all should meet for a beer, anyhow.
The only way I can see around this is to run the LTZ1000 at cryogenic temperatures-- like in liquid nitrogen (LN2).
Will cause problems due to the thermal shock, and will stress the tightness of the package.
Also, as the internal gas will dilute and create a vaccum, the package might implode, or will get a leak and suck LN2 inside.
 This would result in an absolutely stable temperature, because the LN2 always boils at a specific temperature--
No, that's not really the case. (I worked with cryogenic liquids for years, and was also responsible for air and nitrogen liquefaction in our institute for some time).
LN2 will quickly soak oxygen from the surrounding air, so the boiling point will rise from 77.2K (or so) to above 80K within a few hours, depending on the amount of liquid you use.
I used LN2 as a reference bath for a low temp thermo couple, and it was important to always use 'fresh' LN2 from the Philips liquefier.
1 l of LN2 used to cost around 50 Cents, or so, LHe4 was 10-20 times more expensive.
A usual 50l cryostate (Dewar) will cost 500-1k $/€ only,  but a complete JJ  is how much? 100k, I think.
A lot of first class calibrations can be paid from that.  8)
I think, the stability of the LTZ1000 circuitry at RT is fully sufficient, and putting it completely in an oven or a crogenic bath will only cause problems and complications
Chose those ultrastable external components instead. Compared to that LN2 steadily boiling off, it is a one time investment only.
Frank
« Last Edit: August 12, 2013, 09:07:45 pm by Dr. Frank »
Offline Andreas
« Reply #297 on: August 12, 2013, 09:24:31 pm »
http://www.google.de/patents/US5369245
Offline branadic
« Reply #298 on: August 13, 2013, 07:08:55 am »
Offline babysitter
  • pushing silicon at work
« Reply #299 on: August 13, 2013, 01:20:17 pm »
I propose a volt-nuts beer meeting at the Weinheimer UKW-Tagung ham festival (September 13th-15th). I will be there, Quarks is living less than 20 km from there, affordable trip for branadic and Dr. Frank too... bargain hunting on the flea market is possible, at the camping site a few km away we could sit down for some $BEVERAGE.
Greetings
Hendrik
I'm not a feature, I'm a bug! ARC DG3HDA
