all esr meters are impedance meters aren't they ?
how do you make the difference between an impedance meter and an esr meter
apart the fact that esr meters are impedance meters testing at 100KHz ?
Using a synchronous (phase sensitive) detector allows the reactive and resistive components to be resolved.
For instance this example from your excellent site http://kripton2035.free.fr/analog%20esr/esr-lowohm-silic.html
Here the switches IC3a & b are driven in phase with the test current form IC3c & d, this will reject the quadrature component caused by the reactance of the capacitor under test.
Jim
The quadrature detector only works well if your test signal is a low distortion sine wave. If you look at the schematic circuit of a handheld LCR meter, there is usually something in the order of an 8th order bandpass filter after the signal generator to wipe out all the harmonics. The final signal applied to the device under test has a distortion of something like 0.01% or less. Really accurate bridges go a step further - they also use a bandpass filter after the device under test, partly so they can cope with distortion caused by the device being tested.
If your test signal is a square wave, the harmonics will make the quadrature detector fairly inaccurate in terms of measuring capacitance.
Also, ESR is most easily measured at high frequencies like 100KHz. Capacitance of large electro's needs to be measured at low frequencies like 120Hz. To get a useful esr reading at the same frequency that you can measure the capacitance value, you would have to start designing to a true LCR meter quality. Otherwise, you will have to have two test frequencies to measure the C and ESR.
Nice idea, but is it too much work for results that may often be deceptively wrong? How much time are you going to waste chasing up capacitor problems that are the result of a dodgy meter?
The significance of an ESR test is that with electrolytics, ESR rise is usually the first sign of something going wrong in the capacitor - ESR will rise before capacitance starts to fall. ESR will go up if the electrolyte starts to evaporate, so it is a sign of a capacitor that no longer as a true gas-tight seal. So a basic ESR meter that purely measures the voltage across a capacitor when fed with 100KHz on its own does a pretty good job. Quadrature switching? Don't know why you would even bother.
Richard.