So I ended with this small circuit to see and hear with a SDR any disturbance/noise in 2KHz-2MHz range injected on mains by an electronic light bulb under test.
Now, after simulation and testing with a DMM to see that the output is safe for the instrument (circuit output in normal conditions is 124mVpp while instrument input is MAX 0dBm/225mVRMS, PEAK +10dBm/710mVRMS, that for extra safety will be used with a 10dBm attenuator and a DC block), I'd like to receive some advices and observations about the safety of the SDR/SA instrument connected.
The critical component for the instrument safety connected at the output of R4 is of course C2 since is the primary connection with mains.
I've been teached that that "Y caps are designed to be safe, metallized paper and film caps tend to open circuit IN FAIL MODE."
If C2 fails, since is a Y metallized paper type, it's engineered to go open, so normally all the circuit beyond this component is considered safe.
But since catastrophic failure is always an option and it would fry my SDR/SA, let's consider the worst case:
First of all: it is possible that a 300V metallized paper Y cap go short on failure mode?
If C2 goes short R2 would pull 100mA of current thru earth tripping the 30mA RCD (usually in 35mS). In this amount of time 240VRMS from mains would be present on R3.
EDIT: R6 and R7 would dissipate 10A and 2.451W so they will fry in normal conditions but I think they can keep that power for 35mS. In the worst case, they go open and most of the voltage will go on the ouput of R4, frying all.
Am my thoughts right? What would you modify to have absolute safety of the SDR/SA instrument connected to R4?