Test Equipment / Re: SDS800X HD Bug Reports + Firmware
« Last post by 2N3055 on Today at 07:47:31 pm »Level indicator is wrong in zoom modus. It is not right in the bottom window (find the arrow).
I can confirm this one. It applies to Qualified and Delay triggers; seems to work OK with other trigger modes.
Steps to reproduce:
Trigger threshold for Qualified and Delay triggers is displayed incorrectly in Zoom mode
Test signal:
None required; leave inputs open.
Scope settings:
- Start from Default mode
- Set trigger mode "Qualified" or "Delay". Leave all other trigger settings at default values.
- Switch axis labels on -- just to visualize the problem more clearly
- Switch zoom mode on
- In the lower window, increase vertical amplification -- e.g. two steps, to 200 mV/div
- Adjust the trigger threshold via the rotary encoder, and observe the screen
Observed behavior:
- The trigger threshold is displayed as a dashed line, both in the upper and lower window.
- Its position is correct in the upper window.
- In the lower window, its position is incorrect. It ignores the different vertical scale in the lower window.
Will check .. Thanks.