Test Equipment / Re: SDS800X HD - Issue disscusions - is it me, or the scope?
« Last post by ebastler on Today at 07:20:54 am »I just gave the Qualified trigger a quick try and cannot reproduce your issue so far.
Using 0V .. 3V pulses on both, the Qualified and Edge channel, I can happily overdrive the channels at 100 mV/div and higher sensitivity. Qualified trigger remains stable for me. Please provide further details on your setup and signals.
One caveat: In Qualified trigger mode, pushing the trigger threshold encoder has a different function than normally. It does not set a default (halfway) threshold, but toggles the threshold knob between controlling the Qualified and Edge channels. Did you possibly set unintended threshold values due to that?
Using 0V .. 3V pulses on both, the Qualified and Edge channel, I can happily overdrive the channels at 100 mV/div and higher sensitivity. Qualified trigger remains stable for me. Please provide further details on your setup and signals.
One caveat: In Qualified trigger mode, pushing the trigger threshold encoder has a different function than normally. It does not set a default (halfway) threshold, but toggles the threshold knob between controlling the Qualified and Edge channels. Did you possibly set unintended threshold values due to that?