Repair / Re: Series defect on agilent 167xx boards?
« Last post by FrodeM on Yesterday at 08:36:07 pm »
I spoke too soon, after removing the rails it no longer pass selftest.

Today I went over with flux, a soldering iron and a microscope. Three of the thin lines that just looked discolorated at a spot ended up disintegrating, so I had to fix that. I also reflowed most components with visible corrosion on the solder pads. Two vias had corrosion into them, but both are connected to planar layers and seems to make connection I figured after a little bit of digging.

I will test it tomorrow, since I need to fix the power-supply of the analyzer again. Despite the grid voltage should be 230V according to the power company, during nighttime it can at times be as high as 240V. Either it's that that blows the regulators, or the PSU has an inherent design defect that triggers if you turn the analyzer off and then on again only a few minutes later.
92
https://linux-hardware.org/?id=usb:174c-55aa

That's what I found in another (linux related) VID + PID data base:

Code: [Select]
174c  ASMedia Technology Inc.
07d1  Transcend ESD400 Portable SSD (USB 3.0)
1151  ASM1151W
1153  ASM1153 SATA 3Gb/s bridge
2074  ASM1074 High-Speed hub
3074  ASM1074 SuperSpeed hub
     -> 5106  ASM1051 SATA 3Gb/s bridge
5136  ASM1053 SATA 3Gb/s bridge
51d6  ASM1051W SATA 3Gb/s bridge
     -> 55aa  ASM1051E SATA 6Gb/s bridge, ASM1053E SATA 6Gb/s bridge, ASM1153 SATA 3Gb/s bridge, ASM1153E SATA 6Gb/s bridge

The read out PID via flash tool seems to be wrong (55AA indicates ASM1051E) but in reality it is a ASM1051 without E.
Does it matter if 55AA or 5106 is used? Should I correct it (5106 resp. ASM1051)?

I would suggest that you identify the serial EEPROM/flash and short the DI or DO pins while powering up the enclosure.
This will invalidate the external memory and cause the bridge to operate as a generic device until the next power cycle.
If the result is what you want, then you could permanently disable the chip.

It's a serial EEPROM 25LC512:



Can shorting SI or SO to GND kill the IC?

93
Test Equipment / Re: Test Equipment Anonymous (TEA) group therapy thread
« Last post by unseenninja on Yesterday at 08:34:33 pm »
Hi! I am a fellow TEA sufferer, with a focus on the voltnuttery side of things. A particularly deep 



I thought I should say hello!

Must be an ePay thing then.

I'm many times too far from Japan, or something like that.

I always thought it was some kind of error. But I also saw a lot of listings with free shipping from Japan. I Never tried to buy one though.

I was also surprised when I saw free shipping from Japan. I needed a programmable, high precision current source and I saw an Advantest R6144 in mint condition from a Japanese seller for a reasonable price a couple of months ago on eBay. It was sent with FedEx at no extra cost and it arrived very quickly. It works perfectly. I was charged import VAT and the FedEx service charge of course, but that was quick and painless.

So if there is something in Japan that interests you, these free shipping deals seem to be genuine. These free shipping deals make US sellers very expensive option now if you are in the EU. I noticed that shipping prices from the US seem to have nearly doubled recently.

I thought people might like to hear from someone who has taken advantage of free shipping from Japan. Just make sure that what you buy can be reconfigured for 230V operation if you are in Europe! It cannot be taken for granted.


94
Because I'm not afraid to use the common words for the sounds I'm trying to mask:  urine hitting the toilet when I'm standing up while urinating and farting.

One of the sounds you're trying to mask is probably similar to this.

Well, you could just "double down" and play the above or something similar. Of course, this may be counterproductive.

You can try to mask that with pink/brown noise. (Use in a second tab to play around with what that might sound like!) Frankly, I wasn't impressed. This seems slightly less objectionable, but IMHO the most pleasing results come from trying to harmonize with the running-water sounds rather than simply drown them out. To that end, I came up with this (which includes running-water noise; I'd keep that because, even though it "adds" the very noise you're trying to hide, having those sounds always present should make the actual noises you're trying to disguise less noticeable).

Unfortunately, the... other problem noise is, well, more of a problem. Because it's subject to much more variability in terms of volume, pitch and duration, it's hard to blend it with something else. You potentially need a fair bit of raw volume if you're going to effectively mask it, and the sorts of sounds that will do the job are much less pleasing than the semi-pastoral sounds from my suggestion for dealing with the first sort of noise.
95
Test Equipment / Re: Rigol DMM DM858 comparison thread
« Last post by UFOdriver on Yesterday at 08:33:19 pm »
Not working upgrade from DM858E to DM858
96
Hardware version 3.0 has that rectifier bug.
If it's still under warranty, revert to old FW and have it repaired.

Mine is version 3.0 as well, but i never had any problem.

Do you happen to know where we can have more information about this 'rectifier bug' ? Is that something easy to fix by ourselves ?
Easy fix for any capable hobbyist...although the soldering might be challenging for those not experienced in lead free solder rework.
This will provide guidance on the rectifier bridge failures and fixes:
https://www.eevblog.com/forum/repair/siglent-spd3303x-e-reboot-problem/
97
Test Equipment / Re: Tektronix TDS5034B, TDS5054B, and TDS5104B. Any Good?
« Last post by nicnac117 on Yesterday at 08:30:26 pm »
Thank you . Would appear its not on  :-DD
98
Repair / Re: Help needed for repair of Efratom FRS-C rubidium oscillator
« Last post by edpalmer42 on Yesterday at 08:27:37 pm »
Hello Ed,

After giving the situation some thought, I think I will construct a replacement three-output supply first, delivering +5, +15, -7.
The +5 V can be fixed, that is needed more often. The other two voltages variable, so I can use it in other situations too.

(Side note: I happen to have a Fluke 7261A that needs the same treatment. Same transformer problem: short in the primary windings.
Completely different transformer, though. Also three output voltages to replace. With a bit of work, I should be able to rewind the primary.
Transformer disassembly started, E-I core removed. But before going any further, I want to know if it is worth while!)

So, now I am thinking about constructing a fixed+variable power supply. That would come in handy more often, I'm sure.
I have some surplus computer supplies, that should take care of the +5, and gives a decent +12 for free.
It is the variable outputs that require some thought.

There are two ways to look at it.  You can build or buy 3 power supplies with the hope that the board powers up and works properly and then those power supplies become the permanent replacement for the entire blown U39 module.  If the counter is dead, you repurpose those supplies as lab supplies.

You could also build or buy three lab supplies that cover the voltages necessary.  In that case, you end up with three nice lab supplies and then, if the board works, build or buy 3 dedicated supplies to repair the counter.  You end up with three lab supplies.  You can *NEVER* have too many power supplies in your lab!  :)

One complication to keep in mind is that the input to the U39 module is high voltage DC rather than AC.  I don't know if typical AC switching power supplies will tolerate high voltage DC input.  You might have to rearrange the input power to feed them with AC.

Quote
Coming back to the Rb oscillator.

Magnetics:

I have the same situation: the bolts for the outer shell are non-magnetic, the bolts for mounting on the base plate are magnetic.
And additionally, and this is where I had not taken care enough to keep things separate: the same is true for the washers!
I have four magnetic ones, and four non-magnetic.  I have sorted them using a weakly magnetized screwdriver.

The mix of magnetic and non-magnetic screws didn't make much sense to me.  I dug through my stash of non-magnetic screws and found four more non-magnetic ones so now I've got a full set.

Quote
Root cause of the failure:

The signal that I saw on the output before the unit repaired itself, was a completely different and well-defined signal.
That tells me that the connector itself is good. I remember reading that there are some analog switches that the 10 MHz signal must go through
before it reaches the output buffer on the supply board. That means that other signals can be routed to there as well.
If that is correct, the failure was in that switching part! Would you agree? Was it truly a failure to begin with?
What is the purpose of the analog switches? (I do not recall reading anything about that.)

I'm not seeing any analog switches.  The oscillator itself is U1, 54LS320 which is a dedicated quartz oscillator chip, followed by a 54HCT393 which is a ripple counter, i.e. a divider.  In any case, you had a solid 10 MHz signal coming out of the oscillator board and I don't see anything between the oscillator board and the power supply board.

Quote
The signal observed earlier is some kind of modulation signal. From the frequencies observed, most likely it was the 127 Hz signal.
If I am not mistaken, the 127 Hz signal is used to modulate the Ghz resonator frequency. But is that 127 Hz signal itself modulated as well?
I think I need to go back and read more carefully about how the modulation and locking mechanism works.

I don't think you stated what the amplitude of the bad signal was.  Did it look like a solid signal or something that was basically noise being coupled in to the output line and then amplified by Q3?  Bad solder joints or other circuit faults anywhere in the system are a possibility, but the clean 10 MHz coming out of the oscillator board but not on the output connector points toward the power supply board rather than the oscillator board.

Quote
Calibration:

In order to get the oscillator out for inspection, I also removed the 20(?)-turn Bourns pot from the front panel.
(So now I know how to circumvent the calibration sticker, which is still very much in tact!) After removal from the front panel,
I carefully measured the position of the pot, so that I could always put it back in position in case it would be accidentally turned.
The resistance measurements have in 5 digits, thanks to the 8842. The unit was originally calibrated to 0.2 mHz.
I have not touched the Bourns pot, nor have I touched the setting on the aux power supply of the unit (the one that delivers 24 V).
Would that suffice to get me somewhere close to the originally calibrated value, or is that just wishful thinking?

Regards, EJ

Rubidium standards do drift over time.  Considering the age of the unit, recalibrating it by comparing it to your GPSDO isn't a bad idea.  It will be challenging to do it with a scope, you really need a time interval counter.  The frequency will also shift slightly with temperature and input voltage.  But since you've got that pot, that's a good starting point.

FYI, the PM6681 power supply is very close, if not identical to the PM6685.  The service manual warns that if you touch the +5V trimmer, you have to adjust 'the complete instrument', whatever that means.  The PM6685 service manual doesn't say that, but since it's so similar, keep that warning in mind.

Ed
99
General Technical Chat / Re: Does the electrical grid acompany network lines?
« Last post by G7PSK on Yesterday at 08:25:50 pm »
Quote
when the solar farm was built the planning application was put in by a UK firm which granted me permission to photographically document the build but just before the build started the UK firm sold off the whole thing to a Chinese company that rescinded the permission. They then imported labour from China (one of the planning requirements was as much local labour as possible) and when I took photos from outside the site they got very angry and shouted at me in Chinese. 

Fascinating! I wonder if that ever made it to the pages of the Register.  Guardian

Not that I know of, in any case the Chinese are heroes to the likes of the Guardian and can do no wrong.
100
Projects, Designs, and Technical Stuff / Re: (Yet Another) DIY Multislope ADC
« Last post by iMo on Yesterday at 08:25:32 pm »
Now, when we agree the runup's end voltage may end up anywhere (positive or negative), would be above "positive" and "negative" symmetrical rundown sequences work fine?

As I indicated above a recommendation I saw somewhere said the rundown sequences (starting at positive or negative voltage) have to approach (at their end) the comparator always from the same side (ie positive) in order to compensate for comparator's hysteresis (and perhaps for other errors too).. That means they cannot be symmetrical.

An example is the Landsberg's rundown NNNI implemented in his MS II - you go always up into positive (positive or negative runup's end voltage regardless) and then down to zero (crossing zero from the same positive side of the comparator).