Author Topic: Agilent 16702 logic analyzer PSU fail  (Read 1485 times)

0 Members and 1 Guest are viewing this topic.

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Agilent 16702 logic analyzer PSU fail
« on: January 08, 2017, 10:00:43 am »
PSU autec 552-1362A pcb rev. .X3  clicking wont start , any schematic beforte to check ?
 

Offline bsalai

  • Regular Contributor
  • *
  • Posts: 87
  • Country: us
Re: Agilent 16702 logic analyzer PSU fail
« Reply #1 on: January 08, 2017, 11:33:37 am »
Alredy replaced all diodes mosfet etc marked in the image





separated vertical module (i m the one broken the small cap ..isn't the cause of failure)

I've not worked on this one, but I've solved a similar sounding problem in other power supplies in HP scopes by removing the electrolytic capacitors, cleaning the PC board under them (even if you don't see anything, and installing new caps. Apparently some of the caps leak.
Good luck.


Sent from my iPhone using Tapatalk
Brad, W2EK
 

Offline mos6502

  • Frequent Contributor
  • **
  • Posts: 537
  • Country: aq
Re: Agilent 16702 logic analyzer PSU fail
« Reply #2 on: January 08, 2017, 04:07:54 pm »
Clicking means that some protection circuit is active, either overcurrent or overvoltage. You need to trace the circuitry going to the control pin of the TOPSwitch and see what is making it stop. Randomly replacing parts is unlikely to result in success.
for(;;);
 

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Agilent 16702 logic analyzer PSU fail
« Reply #3 on: January 15, 2017, 02:51:08 pm »
Tried to draw schematic from board , maybe there is errors , got short SCR1 , SCR2 and D3
5 modules input 300Vcc



« Last Edit: January 15, 2017, 05:43:06 pm by Daxxin »
 