Hello Forum.
34401A 2W Ohm strange phenomenon
I already fixed previously
https://www.eevblog.com/forum/testgear/34401a-cal-error-707/Thanks again Dr. Frank, free_electron, robrenz
New problems seem to come thick and fast
I would like to do Fix this problem before Calibration Request
Symtoms
0. Self Test Pass.
1. Power on, 2W Mode button Press, OVLD OHM is Normal Display (Attached Photo : pic2)
2. Touch my finger Hi Input Jack, "1xx M OHM" Dislay (Attached Photo : pic3)
or, Power on, after a 2~3 minute lapse "1xx M OHM" Dislay (Attached Photo : pic4) * bottom unit
3. Known good meter (Attached Photo : pic5)
* upper unit
Add:
4. Ohm OFFSET Error 722 (Cal 2-wire ohms function offset out of range)
- Short front Input Jack then CAL-> CALIBRATE -> OHM 2,3,4 Beep
* I am speculating phenomenon as the reason.
* Other function normal
In my experience, Agilent/Keysight do not repair, only Change Mainboard
understand that, but It's too expensive, too harsh
How can I understand This situation?.
Please share knowledge
Thank You!~
Reference. (an excellent article)
https://www.eevblog.com/forum/testgear/34401a-cal-error-707/https://www.eevblog.com/forum/repair/hp-34401a-drift-in-the-2w-r-mode-in-the-100-ohm-range/Thank you Z, for sharing your knowledge