No help needed on this one, just hoping to help someone else out.
The series of GW Instek, PSP-405 lab bench power supplies has capacitor issues. We have about ten of these power supplies. In the past year, three of these power supplies have failed. They start acting up by the output not turning on consistently. The next stage of failure is all of the LCD segments turning on when the PS powers up or the PS being completely dead.
The GW Instek, PSP-405 is a 40V, 5Amp, programmable bench power supply. The programmable part of the power supply is though a 9-pin RS-232 connector on the back of the supply. Amazon has the power supply new for $310 USD, which seem really high to me. The power supply was also found under another brand: Iso-Tech ISP-405. (At least, the faceplate and specs look identical) I’m not sure if any others in GW instek PSP series have the same cap issues. They are the PSP-603 and the PSP-2010.
Two of the failed power supplies were 9 years old, with the third failure being a 5 year supply. All of the power supplies had issues with the four capacitors (see "four caps.jpg")
On the 5 year old power supply, it was showing the first stages of failure with the output not turning on consistently. There were no blown tops on the four caps in the 5 year old supply. (but there were vented caps on the previous power supplies) After the 4 caps were removed, here are the ESR readings:
Capacitor Printed value Measured Value Measured ESR ESR of a Good Cap
C25 470uF, 25V 409uF 4.1 ohms 0.25 ohms
C29 470uF, 25V 414uF 3.9 ohms 0.25 ohms
C33 470uF, 25V 434uF 4.3 ohms 0.25 ohms
C49 100uF, 100V 86uF 21 ohms 0.4 ohms
Digikey has some good replacement caps:
C25, C29, and C33: Part number P12389-ND, CAP ALUM 470UF 25V 20% RADIAL
C49: Part number P5313-ND, CAP ALUM 100UF 100V 20% RADIAL
So far, all three power supplies with the new caps are performing well. Anyway, I hope this helps someone out.