Author Topic: Agilent 33210a Components not identified.  (Read 1579 times)

0 Members and 1 Guest are viewing this topic.

Offline jocelioTopic starter

  • Newbie
  • Posts: 3
  • Country: br
Agilent 33210a Components not identified.
« on: January 15, 2017, 01:32:36 pm »
After an error of operation from my part grounding damaged the 3 components as shown in the photo it is working from to see that it is normally generating the components then short .
 

Offline bktemp

  • Super Contributor
  • ***
  • Posts: 1616
  • Country: de
Re: Agilent 33210a Components not identified.
« Reply #1 on: January 15, 2017, 01:46:17 pm »
One part seems to be a tranformer or a common mode choke. So measuring a short accross its pins could be normal.
The other ones are TVS diodes, based on the marking GFT they should be SMCJ43A.
 
The following users thanked this post: jocelio

Offline jocelioTopic starter

  • Newbie
  • Posts: 3
  • Country: br
Re: Agilent 33210a Components not identified.
« Reply #2 on: January 15, 2017, 02:17:54 pm »
What seems to be a type of choke is giving short circuit in all the send and when I put it back simply it eliminates the pwm it is in short posivo with negative preventing any type of pwm when I take it back the pwm and the generator Works normally the diode seems to be more of a type of varistor and is in short. And has no polarity.
 

Online Andy Watson

  • Super Contributor
  • ***
  • Posts: 2085
Re: Agilent 33210a Components not identified.
« Reply #3 on: January 15, 2017, 02:38:04 pm »
The manual for the 33220A indicates that the four-pin device is, as BK suggests, a common mode T1101. The manufacturer's (TDK)  part number is ACM4532-801-2P, I would expect a similar device in the 33210A.
 
The following users thanked this post: jocelio

Offline jocelioTopic starter

  • Newbie
  • Posts: 3
  • Country: br
Re: Agilent 33210a Components not identified.
« Reply #4 on: January 15, 2017, 02:49:19 pm »
The manual for the 33220A indicates that the four-pin device is, as BK suggests, a common mode T1101. The manufacturer's (TDK)  part number is ACM4532-801-2P, I would expect a similar device in the 33210A.
    Here in Brazil ACM4532-801-2P does not exist for sale what can I do? Already the vr602 FAD27 would be possible a substitution of the components for another type of component would have some type of solution for me being that I wanted to preserve the safety of the equipment .. because here in Brazil the prices are abusive.
 

Online Andy Watson

  • Super Contributor
  • ***
  • Posts: 2085
Re: Agilent 33210a Components not identified.
« Reply #5 on: January 15, 2017, 03:52:12 pm »
Looks like it's obsolete!
 http://uk.rs-online.com/web/p/signal-filters/5474850/
Go with the nearest part that you can find. 4532 is the physical size of the device. 801 is the value of interest, it indicates it is 800\$\Omega\$ impedance at 100MHz. 2P indicates two pole (rather than 3 pole/lines).
For low quantities ebay can sometimes be useful.

 