Author Topic: FIXED: Agilent E4418B repair: possibly require the FLASH EEPROM  (Read 7127 times)
Offline analogRFTopic starter
Re: FIXED: Agilent E4418B repair: possibly require the FLASH EEPROM
« Reply #25 on: September 10, 2018, 08:13:10 pm »
...the other side of the new key pad.
Online TheSteve
  • Living the Dream
« Reply #26 on: September 12, 2018, 06:43:27 am »
just a quick update on the final state of the E4418B. I received the key pad from china ($46 incl shipping with UBI which is fairly fast). Photo is attached here. Replaced it and the unit works like a charm now. Slight calibration of the ref. osc. was needed which I crudely did with the help of another E4418A and E4412A power sensor I borrowed again from work. The actual cal process is more complex and requires HP 432A with thermistor which i don't have. but it all works ok now
One thing I noticed and I hope someone could check it, is that if you go to Systems and then to Version, it shows up as E4418A.
Here is the version screen from my E4418B:
VE7FM
The following users thanked this post: analogRF
Offline pigrew
« Reply #27 on: September 12, 2018, 01:43:52 pm »
The model ID can be changed over GPIB:
serv:vers:proc "C"
serv:vers:syst "2"
The A and B models have different I/O ports on the back (B has "RMT I/O" and RS232). Switching the model number as above might enable these... I'm not sure. The A and B models use the same firmware. There are also options (listed in the manual) to set the serial number and installed options.
« Reply #28 on: September 12, 2018, 01:48:56 pm »
serv:vers:proc "C"
serv:vers:syst "2"
This is great. I'll look into this. Thanks!
« Reply #29 on: September 12, 2018, 08:50:36 pm »
@pigrew suggestion worked great. here are the pics before and after I run those two SCPI commands
but it seems I should upgrade the firmware as mine is one version old...maybe later...
EDIT: the power LED in the replacement keypad is so shiny and bright I cannot stare at it more than 2 seconds  :o
« Last Edit: September 12, 2018, 08:54:29 pm by analogRF »