So I've started to tear down and repair the loot from the Duesseldorf trip
http://wunderkis.de/mg3633a/index.html

This is work in progress, first unit is fixed, two more to go, stay tuned for updates.

Some of us might remember DC1MC and me picked up some stuff there:
https://www.eevblog.com/forum/buysellwanted/%28deeu%29-group-sale-of-5-anritsu-mg3633a-signalgenerators-%2810-khz-to-2700-mhz%29/
« Last Edit: March 22, 2018, 11:27:00 am by capt bullshot »
Safety devices hinder evolution
 
The following users thanked this post: bdivi, DC1MC

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair
« Reply #1 on: March 21, 2018, 02:10:46 pm »
Update #1:
Some more circuit analysis and fault location happened to unit #2, not yet repaired but root cause of malfunction located:

http://wunderkis.de/mg3633a/repair2.html
Safety devices hinder evolution
 

Offline DC1MC

  • Super Contributor
  • ***
  • Posts: 1882
  • Country: de
Re: Anritsu MG3633A signal generator teardown and repair
« Reply #2 on: March 21, 2018, 07:01:39 pm »
Excellent tear-down and repair detective story, recommended for everybody who wants to see a properly done and documented repair  :-+.
Also the generator construction is really a wonder of Japanese technology, I had some kind of electrogasm looking at that gold attenuator !!!

Keep on the good work,
DC1MC

 

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair
« Reply #3 on: March 22, 2018, 11:24:29 am »
One more update for today:

Unit #1 is no more wacky: http://wunderkis.de/mg3633a/repair1.html
Unit #2 is "undefekt" now: http://wunderkis.de/mg3633a/repair2.html
Unit #3 is diagnosed: http://wunderkis.de/mg3633a/repair3.html

Two out of three, that ain't so bad?

« Last Edit: March 22, 2018, 11:28:29 am by capt bullshot »
Safety devices hinder evolution
 

Offline DC1MC

  • Super Contributor
  • ***
  • Posts: 1882
  • Country: de
Re: Anritsu MG3633A signal generator teardown and repair
« Reply #4 on: March 22, 2018, 11:48:47 am »
One more update for today:

Unit #1 is no more wacky: http://wunderkis.de/mg3633a/repair1.html
Unit #2 is "undefekt" now: http://wunderkis.de/mg3633a/repair2.html
Unit #3 is diagnosed: http://wunderkis.de/mg3633a/repair3.html

Two out of three, that ain't so bad?

Hey capt, are you taking apprentices ? 'Cause my unit is still in need of some love, I can haul it back to you some time for some spiritual healing  ;D
But excellent job indeed, done in a very German way  ^-^ !!
 
I have this strange idea, if you mix the outputs of the two guys that are OK, will this become an even higher frequency generator ?
And second, less wacky idea, if those guys will be feed in their reference input from a rubidium 10MHz source, will they become some craz stable generators ?

 Cheers,
 DC1MC
 
The following users thanked this post: capt bullshot

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #5 on: March 22, 2018, 01:52:03 pm »
I've added some thoughs about the rather subtle cause of the ceasing oscillation within this PLL (refering to repair #2):

Quote
In the end, we've got a nasty but innocent looking electrolytic capacitor causing all this inconvenience. I'm slightly not amused now, but happy because the unit is up and running again. Alas, the culprit suffered from desoldering and cutting it off the board (the bottom side of the PCB isn't accessible in an easy way), so I couldn't investigate further on the failure mode of this component. For sure, it'll be properly widlarized then.

I guess that little electrolytic was dried to some extent, and changed its ratio from C to ESR / ESL somewhat. The ESR must have been still low, otherwise there would have been too much damping, but the ESL apparently formed a resonant tank with the paralleled ceramic cap, resonating in the ballpark of 300MHz. This is the low level oscillation that I could observe when the desired 95MHz oscillation ceased. The damping ratio of these concurring tank circuits must have gone critical here, so the parasitic one could take over.

I remember Jim Williams warning about paralleling electrolytic and ceramic caps for decoupling purposes (it's written up in one of his famous app notes), just because this tank formed
by the high Q ceramic and the electrolytic's ESL could create serious ringing.

Safety devices hinder evolution
 
The following users thanked this post: DC1MC

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #6 on: March 23, 2018, 09:17:24 pm »
I've done some timenuttery with the two working ones now:

Using a simple power splitter, one can add (not mix) the outputs of two signal generators to see the beat frequency on a scope.
Both units set to 100MHz, the scope shows about 2Hz - meaning the OCXOs match to 2*10E-8

Adjusting one generators frequency to 100MHz - 2.05Hz results in a constant amplitude (meaning the frequencies are the same), resulting to 2.05*10E-8 oscillators.
The reference oscillator of the CMU200 is just about the same accuracy - quite nice for about 1992 vintage instruments.

I guess, now I need a GPSDO ...
Safety devices hinder evolution
 

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #7 on: April 18, 2018, 06:31:02 pm »
Update on unit #3:

http://wunderkis.de/mg3633a/repair3.html

Short story: It came to life again, I've replaced some blown up and obsolete components by modern stuff.
Safety devices hinder evolution
 

Offline DC1MC

  • Super Contributor
  • ***
  • Posts: 1882
  • Country: de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #8 on: April 18, 2018, 06:50:12 pm »
Capt, wie immer (as always), an excellent repair job, nicely illustrated and with good and cost effective solutions found.  :clap: :clap: :clap:
This puts my lazy old a$$ to shame, you've repaired 3 and I was only able to repair the PS of the SA and order a CRT rejuvenator to see if the tube can survive a bit more or it's the LCD time.
I definitely have to meet you again or put my behind in movement.

 Congratulations and keep up the good work,
 DC1MC

P.S. If you need rejuvenate some of your CRTs, I have the professional solution coming  8).
 

Offline capt bullshotTopic starter

  • Super Contributor
  • ***
  • Posts: 3033
  • Country: de
    • Mostly useless stuff, but nice to have: wunderkis.de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #9 on: April 18, 2018, 07:11:01 pm »
That was a pretty quick reply, have you been sitting all the time and waiting for this post?  :-DD
Safety devices hinder evolution
 

Offline DC1MC

  • Super Contributor
  • ***
  • Posts: 1882
  • Country: de
Re: Anritsu MG3633A signal generator teardown and repair log
« Reply #10 on: April 18, 2018, 07:26:50 pm »
That was a pretty quick reply, have you been sitting all the time and waiting for this post?  :-DD
Told you', lazy a$$  :-DD
I was just logging in and looking at the replies to my posts while crunching some toxic blue chips from Kaufland  >:D, we probably reach home around the same time.

 Cheers,
 DC1MC
 