Questions so we can further diagnose.
1) With the probes shorted and rotary dial set to ohms, what does the LCD show? Do the annunciators look correct?
2) With the probes open and rotary dial set to ohms, what does the LCD show? Do the annunciators look correct?
3) With the probes shorted and rotary dial set to volts, what does the LCD show? Do the annunciators look correct?
4) With the probes open and rotary dial set to volts, what does the LCD show? Do the annunciators look correct?
5) Did you check the input protection components like the MOVs, PTC and fusible resistors, given that this is an used and abused meter, to make sure they are okay?
1) The analog bargraph shows 1 bar up and down. The reading is blank, never settles.
2) On first power up, it goes straight in OL, then slowly starts creeping down, to a around 6 Megs, this is with no probes. Connecting probes it jumps all over the scale and autorange never settles.
3) First overloads (OL), then displays around 1000 volts and starts going down (looks like a capacitor discharging) until it reaches 0.000 in about 10 seconds.
4) Displays -300 volts, steady if not touching probes. When touching probes, lecture jumps around.
5) This is what each component measures (using my trusty Fluke 77 IV):
- 10 Amp Range Fuse: open, shorted it temporarily for now.
- mA Range Fuse: short, its a glass fuse, will replace with proper HRC fuse once repaired.
- 4 input big resistors: each around 450 ohms.
- PTC1 670 Ohms.
- PTC2 1K94.
- SPG1, SPG2 (MOVs?): 11 Megs
- Fusible resistor R57: 1 Ohm (matches markings).
- Also checked every diode on both boards, all in the 0.6 volts ballpark.
Thanks!