Hi
I have an Agilent Infi Vision MSO-X 4054 A oscilloscope which hangs up during startup. Following the troubleshooting guide, I found that the backlight supply and the auto probe voltages are absent. Further probing shows that the NCP3064 switching regulator looks dead. Even after replacing the IC the problem persists. THere is no signal at timing capacitor pin 3. VCC is available on switch collector pin 8. Switch Emitter pin 7 is low. On /Off pin is also low.
Has anyone faced / solved this problem Any help is appreciated.
Regards