I've repaired a Voice recorder switch (something like
https://www.fishpond.co.nz/Health/AbleNet-10002100-Bigmack-Communicator/0736846869142The fault was a standard microswtich - rated 1/2A 125VDC, 1/4A 250V.
The fault was that the device would play back, but it wouldn't record. The cause was that the NO contact would not stay closed while the switch was depressed. The switch is grounding a pull-up resistor on the microcontroller, so only a few mA is being switched.
Another complicating issue as that the return spring on the switch was being used as the 'lifter' for the external switch plate - when I replaced the switch I had to include some open-cell foam to give my weaker microswitch some added oomph.
My questions:
- Does using a mains-rated switch to switch minuscule currents and voltages case it to fail early (ie. do they need the higher voltages to keep the contacts clean?)
- If a microswitch is not the correct thing to use, what is the correct thing to use?
- Should there be mechanical isolation from the switch pad to the switch - so if somebody bashes the switch the full force is not directly transferred to the switch?
- is using the switch as a spring 'bad form'?
- Where can I get some microswitches with greater lifting power?
edit: fix typos