Author Topic: Fluke 123 scopemeter Repair  (Read 1635 times)

0 Members and 1 Guest are viewing this topic.

Offline RaymondTopic starter

  • Newbie
  • Posts: 2
Fluke 123 scopemeter Repair
« on: June 02, 2017, 04:59:03 am »
Hello there..
I have a scopemeter 123 who is acting not okey.

Wen i select continuous mode it start to buzz wen i put the probe a + com togehter it stops chiming it does not measuring cap mode and ohms.

Wen i put the test leads in the wall from channel a and the com port i get the right readings 227v and showing a 50hz trace. Besides that nothing seems to work. Either i do not hear the clicking of the relays path ways

I dont know what the do with this little boy but they seems to blew it good..

Things i did check.
Thermal protection ptc's
The protection circuit. The diodes and the fusable resistors


The reference circuit its showing the right value.
From the service manuals.
I hope there is some with a kind of this experiences who can give a tip where to look for i really lost in this tiny parts who calls (smd) :-DD

Thank you so much for your time
Take care
Raymond
From Holland




 

Offline Nivlek

  • Newbie
  • Posts: 1
  • Country: gb
Re: Fluke 123 scopemeter Repair
« Reply #1 on: August 23, 2023, 06:15:18 pm »
I know that this is an old post , but I have a Fluke 123 with a similar issue . Did you ever find the problem ?
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf