TLDR; 1399 = VARISTOR,910, ±+-10%,1.0MA, V910LSX1399
I asked modemhead to check out his early generation Fluke 179 (rev 7) to see if his MOVs had the same markings. He was kind enough to include a picture (attached).
After some searching, the H in Harris.
As some of you know, the 170 predecessor was the Fluke 70 series. Then I found this thread
https://www.electronicspoint.com/threads/where-to-get-fluke-spares-surge-protectors.136603/to confirm it was Harris.
This led me to open my Fluke 75 II. The printing on the MOVs were hard to see, but I can clearly see H and 1399.
The service manual for the Fluke 75 II states VARISTOR,910, ±+-10%,1.0MA, V910LSX1399
http://assets.fluke.com/manuals/77______smeng0100.pdfAdditionally, the service manual for the Fluke 77 III has the same specifications. So it looks like the MOV specs were carried over from the II to the III to the early revision/generation 170 series.
That is why I asked for pictures and the revision number. All the above helped me piece together the above.