Hi all,
I allowed the unit to operate for about 4 hours, and is supplying a voltage in the "correct" range more or less to 3-4 places after the decimal. Ive found myself in the position of not having a volt meter with sufficient resolution to view the output voltage accurately. I think a voltmeter upgrade is in my future.
This morning I discovered a cap on the current limiter has leaked electrolyte during the 4 hours it was operating. This is disturbing to me because it tested fine on my component tester. I think I may replace more electrolytic caps to avoid future issues. For starters I plan to replace C1-5 on the HV mother board, C1, C4, C6, and C10 on the current limiter, C2 on the series pass, and C2 on the Aux Power board.
Any thoughts on this plan? I would not want to disturb the accuracy or linearity potential, I dont think any of these caps would effect this correct?
Good luck finding a high resolution VM at a reasonable price.
Replacing caps on any of the removable boards is pretty straightforward once you get the cap. I found that axial caps are a dying breed.
Replacing caps on the HV MB is very difficult. Fortunately I only had one (C7 which parallels the interlock K1) which measure bad. Getting to the big 3, C1,2,3 is very difficult. I found that the 'paper' tube was not the actual capacitor but a cover. I my unit the actual caps filled about 2/3rds the height of the tube.
My first round of replacements allowed setting 0 volts output on all ranges and dropped the power input from 40 to about 30 watts. Prior to replacement I couldn't get below a few mV at 0 setting.
As near as I can tell, linearization doesn't depend on any caps.
My 335D has been on or in standby continuously for almost a year with no problems except the decimal LEDs and null meter 0 setting on the 1mV and under ranges. (Optical chopper aging I think)
I originally set the 335D reference with a new DM3068 that was still in cal. The three meters on the top (2 HP3478As and an HP3456A) are as received from eBay sellers with far out of date cal stickers from well known labs associated with either the gov't or big manufacturers. I fiddled with the HP3456A on the bottom.
I've had the 335D up to 1100V a couple times to check a couple VTVM meters with 1500V ranges. Most of the 335D use is at 10V or less. I did use the 100V range a couple times to match neon bulbs for the optical chopper in the null meter and to check a couple zeners.
PS: I have modified the bench for better thermal management, i.e., I added some space above the 335D for better cooling.