This is the power supply for a Fluke 8050 with the battery - 01 option. The batteries are new, and it seems to work, except the readings are all about 40% too high. One of the battery leads broke off, the manual says don't run it without batteries, and that's about when the problem started. Tracking it down, the VR1 reference is too low, causing the deintegrate to take longer, causing the high readings. Further checking finds that all the power supply voltages are low. I've marked the voltages in red that I'm seeing now.
Advice?
The zener VR2 has about 3.6v across it, but out of circuit it's ok at 5.6v.
Am I correct assuming Q5 and Q6 must be ok, otherwise I'd get no output voltages?
My problem is I'm busted and can only sit or stand for,, 5 minutes at a time. I run down, make a measurement, and then consider it while lying on my back, so I need to plan carefully for each test.
Thanks for suggestions.