Author Topic: Fluke 883A Differential Voltmeter: Teardown Pics + Mods  (Read 10728 times)

0 Members and 1 Guest are viewing this topic.

Offline FlyingHackerTopic starter

  • Frequent Contributor
  • **
  • Posts: 807
  • Country: us
  • You're Doing it Wrong
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #25 on: May 22, 2016, 04:25:19 am »
Thanks for the info SeanB. With Sulfuric acid I may just replace them with aluminum electrolytics. But I have everything working really nicely now. So perhaps I will just inspect it every few months.
--73
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16276
  • Country: za
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #26 on: May 22, 2016, 05:07:07 pm »
Less than 5 drops inside them, so not to worry. We used to use the dead ones as impromptu fireworks, just by applying 28VDC in reverse to them. Ones that were totally dead did nothing, but the dying ones would go off like a cracker. Mount in the vice and aim it, and you could get a good whack on somebody across the room.
 

Offline FlyingHackerTopic starter

  • Frequent Contributor
  • **
  • Posts: 807
  • Country: us
  • You're Doing it Wrong
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #27 on: May 24, 2016, 05:04:55 pm »
I wasn't so much worried about the acid hurting me, as I was worried about the acid hurting the PCB and other components if they leak. :scared:
--73
 

Offline crissterrian

  • Newbie
  • Posts: 1
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #28 on: December 01, 2017, 07:36:28 pm »
Enyone have the BOM for the fluke 883a and schematics?
 

Offline Martin.M

  • Frequent Contributor
  • **
  • Posts: 956
  • Country: de
  • in Tek we trust
    • vintage Tek collection
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #29 on: October 12, 2018, 10:23:17 am »
I have buyed 883A today
greetings
Martin
 

Offline KK6IL

  • Regular Contributor
  • *
  • Posts: 55
  • Country: us
Re: Fluke 883A Differential Voltmeter (Teardown Pics)
« Reply #30 on: June 06, 2019, 02:59:26 am »
One thing that drives me nuts working on old meters (and new meters I guess) are the crazy requirements for AC calibration. You need things like a 1000V AC 400Hz source, and a 20KHz source (I think).

Is there any reasonably priced way to get these signals? I am always on the lookup for a Fluke calibrator, but they usually go for at least $500, which is way too much to spend on something I would only use to repair old stuff.

Old thread, but I'll answer anyway.  Back about 1968, I was working at Wavetek, when they were making differential voltmeters, trying to compete with Fluke. Initially for AC calibration, we were using a audio oscillator, into a Macintosh audio power amp into a transformer to step up the voltage.  We used a Holt thermal transfer to determine what the output voltage was.  Eventually replaced the kludge with an Optomation AC calibrator, which was a 4 foot tall rack mount oscillator, 6 digit ratio transformer and power amp.  Still required the Holt to determine AC voltage.

So there are ways to fake it with non-standard signal sources. Not too hard to get the high voltage.
And unless you need to measure high v, high freq with accuracy, skip calibrating the 1KV range or calibrate at a lower voltage.  Something like the AnEng AN870 for $30 can measure AC true RMS, 0.3% accuracy up to 3 kHz.  A scope could be used to insure the signal source is flat with frequency.  A scope with differential plug in, 7603/7A13, 545/W, etc could measure the signal directly with accuracy.  Peak voltage X 0.707 = RMS.  I'm using the 7A13 to measure AC signal source to verify my new HP 3456 is reasonably accurate on 1 and 10 V AC ranges.  DC measurement of the comparision voltage terminal on the 7A13 verifies it's accuracy.
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16604
  • Country: us
  • DavidH
Re: Fluke 883A Differential Voltmeter: Teardown Pics + Mods
« Reply #31 on: June 06, 2019, 05:35:39 pm »
For a precision high voltage AC measurement for calibration purposes, I would try sampling at the high voltage reducing the problem to a high voltage DC measurement.  (1)

Sampling also works well to make RMS (and peak, and noise) voltage measurements at RF and low microwave frequencies because samplers can have a very predictable frequency response curve which can be verified by measuring the first null in their response.

(1) I have a truly marvelous design for a high voltage sampler which this post is too small to contain.
 


Share me