I've recently renewed my hobby of building electronic stuff, and I talked myself into buying an old oscilloscope, a Tektronics 453. For somebody with my skill level, it's more than good enough for me. Still, I've been putting it through its paces and I noticed something, and I wonder if any old hands will know what to make of it.
I get crisp readings when I set the scope to >20mv/div, but when I get lower than that, the scope loses resolution - that is, instead of a clean line, I get a fuzzy blur, more or less centered where the real reading should be, that grows proportionally with the volts/div setting. That is, at 5mv/div, the cloud of readings is a full div. At 10mv/div, it's a half a div, and at 20mv/div, it's a quarter (meaning that it's pretty much accurate enough at that level). Focus is set correctly - e.g. if I set the oscilloscope to GND, the line is crisp.
If I clip the probe and ground together, I get a clean line. If I simply clip a 1K resistor between the probe and its ground, I get a clean line. If I clip a 25K resistor, the fuzz starts to come back and at 1Meg, it's fully there, so I don't believe it's noise in my power supply that's doing this.
I haven't come across this in reading the manual (that I got from the Boat Anchor Manual Archive). There's talk about "Astigmatism Adjustment", but from what I read elsewhere, that doesn't seem like what's going on here.
Honestly, I won't cry if I just have to accept this as a limitation of the equipment. What do you want for a $60 'scope? Almost everything I do is such that 20mv/div is more than enough anyway. But if it's a fixable thing, I'd be glad to know it.
Thanks!