Most probably, U201-B is defect, i.e. the input transistors are broken.
These are bipolar super beta input stages, having only about 50 pA bias current.
I assume, that the + input, pin5, drives excess current (~35µA) into the node between Q201, pin1 and the resistor array, 28k75, pin4, as this voltage 2) should be about 9.08V, instead of 10.8V.
This also excludes the possibility, that the failure is somewhere in the output stage.
As + input 2) is lower than - input 3), the OpAmps output 14) should have zero volt output.
That also indicates, that this OpAmp is defect.
I think, that is exactly the same problem as recently.
Replacement of this OpAmp AD706 should fix the problem.
Frank