I have on Tek 2440 have problem , when running in 5 10 min have gain error and runing selftest fail . I open case and force cooling CCD and DAC module by other fan, everything ok when running 1 hour or more, selftest Pass.
Inside case not have must dusty so heatsink very clear. I checking power rail it's all OK , voltage value and voltage rip OK.
So i runing coldstart to carlibration Tektronix 2440 . Runing ATTEN and TRIGER pass but CTE not pass.
NVRam in 2440 using static cmos HM622556LP-12 , and external Lithium battery ( Checking 3.7V OK - Battery not dry )
So I attach picture and video clip , when runing CTE calibration , display flicker and allway report overload in 50R input ATTEN . I Don't know why CTE Calib auto switch to 50R ATTEN input
All information reference you can see in topic
https://www.eevblog.com/forum/metrology/tek2440-mini-teardown-nvram-replacement-and-calibration/I concern register memory in NVRam wave form capture have error.
Many thank so much!
Edit : Display Flicker only when runing CTE Calibration mode ( You can see in video clip )