high pitch noise from a tek 422 oscilloscope
« on: September 08, 2017, 09:04:39 am »
Hi,

I got an old tek 422 oscilloscope from ebay.  It's from the 60s or early 70s.  I am guessing it's been in storage for decades.  When its power on, there is a high pitch noise.  Its like the squeaky sound a mouse mades.  Is it coming from the transformer?  bad filter caps?  What would cause this?

Please note this scope is completely sealed; there is no fan.  No dust bunny in this machine!  Aside from the common problem with dirty potentiometers and switches, the scope mostly works!  I really like the aesthetic of this scope.
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #1 on: September 08, 2017, 09:24:47 am »
Likely bad filter capacitors at the power supply.
 

Offline voltz

  • Frequent Contributor
  • **
  • Posts: 267
  • Country: gb
Re: high pitch noise from a tek 422 oscilloscope
« Reply #2 on: September 08, 2017, 09:54:30 am »
Would check the EHT transformer. Its basically part of a DC-DC converter driven by a high frequency clock. If the transformer starts to break down or is just loose internally, you will hear a high pitched squeal. Hopefully its not breaking down, your trace looks bright enough and i assume there are no bad smells from the scope like burning plastic. May just need a slight tighten of a screw or a dab of silicon.
Would also check the schematic for any other DC-DC converters in there. Same thing applies to those, if any. Not got the diagram in front of me so just general advice here.
 
The following users thanked this post: crazyhog

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #3 on: September 08, 2017, 02:04:40 pm »
See: TekWiki 422 for pictures and schematics.

As voltz said, the squealing is in the high voltage section (EHT aka Flyback transformer) that supplies the anode of the CRT.  It's the only area that runs at the frequency you describe.  Follow the the large grey wire and cap in the pictures that's hooked to the side of the CRT.  Per the schematic, it's running at 4.9kV.

Open the unit up and see if you can localize the noise.  You can check for an ozone smell then and perhaps run the scope in a no light environment looking for any corona glow on that HV lead.  You could also have a condensation issue from its time in storage.

T
« Last Edit: September 08, 2017, 02:07:49 pm by Toasty »
veritas odium parit
 
The following users thanked this post: crazyhog

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #4 on: September 09, 2017, 09:37:22 am »
Thanks everyone for your help.  Today the scope blew it's fuse.  I ended up troubleshooting that and haven't got around to follow up on the EHT. 

Two electrolytic caps blew their top and failed short.  To my surprise, I found the failed caps are hot glued to the circuit board!  I don't believe Tektronix would use hot glue in the manufacturing of this beautiful machine!  As such, I am probably following the previous owner's foot steps.  Anyway, after replacing the capacitors, the scope is working again. 

Unfortunately, the high pitch noise is back too.  I will dig into the EHT.
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #5 on: September 09, 2017, 09:51:33 am »
Then you should look for further electrolytic caps, not just 2.
Don't wait for it to show smoke signs, ESR all those, it's old enough to give it a new life with all new caps   :D
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #6 on: September 09, 2017, 10:14:02 am »
Yup, I've checked all the caps.
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #7 on: September 09, 2017, 10:20:55 am »
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #8 on: September 09, 2017, 04:16:51 pm »
I've confirmed the noise problem is not the EHT driving the CRT (no noise under the high voltage shield of pic 1).  The noise is from the power supply module (pic 2).  Good thing since I don't know know how to fix a failing EHT transformer :D.

There is an AC only Tek 422 model and a AC/DC Tek 422 with internal battery pack model.  I have the later and I suspect the power supply maybe using SMPS circuit to deal with DC power source.  Will read the manual and dig around more.   I hope the noise in the power supply is not inherent in the SMPS tech of the 1960s.  I like a silent scope.

While digging around, I cleaned up a really dirty pot use to control intensity (see pic 3).  This pot is not like the others.  The factory used a green stuff in making this pot.  The green stuff has coated the wiper and the track and has dried and hardened over the years.  The green stuff is non conductive!  No idea why this green stuff is needed for the intensity control.  Anyway IPA worked well cleaning and restoring the pot. 
 

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #9 on: September 09, 2017, 04:49:27 pm »
The "green stuff" is a lubricant and insulator that keeps any wear and possible arcing that might occur on the pot's wiper to a bare minimum.  It also provides some mechanical resistance so the pot feels "smooth" when adjusting it. 

I would use a high quality silicone grease available today.  You want something that's more of a paste.  The intensity control is in a 1400V circuit, with the focus control next in line.  See the schematic "CRT Circuit" in the link I posted.

Since you've narrowed the noise to the power supply, it's probably time for recapping the electrolytics!  :D

You can't depend on value checks alone, so get your ESR meter out an have at it. 

Is the battery pack connected?  If it is and it's bad/dead/shorted, that could be causing the power supply to squeal too.

... Good thing since I don't know know how to fix a failing EHT transformer :D

There's usually no "fix" to those transformers.  However, fixing the rectifier/multiplier can be quite easy.  Just HV caps & diodes.

T
 
« Last Edit: September 09, 2017, 04:51:34 pm by Toasty »
veritas odium parit
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #10 on: September 11, 2017, 11:51:43 pm »
Thanks Toasty for your help.

The noise is coming from a toroidal transformer (big metal can in pic 1).  The Tek AC/DC model has a power supply with a boost converter to handle a wide DC power source (11.5v to 35v).  The boost converter has a 7khz oscillator that drives the toroidal transformer.  Even when the scope is power by AC, the power supply would rectify the AC to DC and then feed the DC into the boost converter. 

My only idea for quieting the toroidal transformer is to fill the the metal can with silicon grease.  Hopefully the grease will lower the pitch of the noise.  It's a bummer.


« Last Edit: September 12, 2017, 12:05:49 am by crazyhog »
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #11 on: September 11, 2017, 11:59:58 pm »

My only idea for quieting the toroidal transformer is to fill the the metal can with silicon grease.  Hopefully the grease will lower the pitch of the noise.

Better to use Shellac instead of grease as a norm practice.
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #12 on: September 12, 2017, 12:08:46 am »
Thanks Armadillo!
 

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #13 on: September 12, 2017, 12:28:11 am »
Loosen the screws top & bottom that mount the coil and try moving it about a bit while running. See if that quiets it down.  Push on the wires with a plastic tool.  You might just find a sweet spot.  You can secure it with some silicone sealant and let if fully cure.

While shellac is the correct choice, the best method is to submerse the coil in a sealed container and apply a vacuum.  That gets the shellac into the nooks & crannies in the windings.

I'd still check those caps.

T
veritas odium parit
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #14 on: September 12, 2017, 07:30:43 pm »
Thanks Toasty. 

I share your concerns about the caps.  The first thing I did after finding the 2 exploded caps was I checked all other caps for both capacitance and ESR.  I did take one shortcut by doing mostly in circuit measurements.  I look for capacitance that is inline with the spec which give me confidence the rest of the circuit didn't influence the in circuit measurements. 
 

Offline xwarp

  • Frequent Contributor
  • **
  • Posts: 367
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #15 on: September 13, 2017, 09:49:29 am »
Just curious, is the NiCad pack still in there?

I think I have an AC supply somewhere around, if you are interested, I'll look for it.
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #16 on: September 13, 2017, 08:47:45 pm »
Hi Xwarp,

The nicad were still in the scope when I got it.  There were 20 D cells and were quite heavy.  They were all dead and some had corrosion.  The first thing I did was removing the battery.  After I get this scope working to my satisfaction, I plan to convert it for li-ion battery.  I really like the idea of having this as my portable scope that can do floating measurements!  The tactile feel and aesthetic of this old timey machine is a joy forever! 

Thank you for your offer for the AC power supply.  I don't need it at this time.

« Last Edit: September 13, 2017, 08:55:11 pm by crazyhog »
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23018
  • Country: gb
Re: high pitch noise from a tek 422 oscilloscope
« Reply #17 on: September 13, 2017, 10:58:46 pm »
Just popping in to say thanks for the decent photos. Lovely bit of engineering that scope. I hope you win the battle :)
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #18 on: September 17, 2017, 03:23:57 am »
Pic 1: Tektronix wants you to keep your dirty paws off the donut!

Notice the can holding the transformer is soldered  all around the entire edge!  It is also a giant heat sink.  All of my soldering equipments were defeated.  I am sure my propane torch can force it open.  Unfortunately, it will be a pyrrhic victory since the can's content will be cooked too!

I thought about giving up.  But I am doing this for fun and to settle seems to miss the point!  Therefore, brute force it is!  A large metal file and half an hour of elbow grease later, I am in like flynn (pic 2)! 

The transformer is coated in a soft rubberly material.  Let's see if this can be improved...



« Last Edit: September 17, 2017, 10:51:39 am by crazyhog »
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #19 on: September 17, 2017, 10:37:40 am »
I plan to submerge the transformer in a potting compound (i.e flood the metal can).  Other folks has already suggested electrical resin. 

Thinking it over, I realized there are many other options that I can experiment with for the specific purposes of sound dampening.  I am currently considering hot glue and silicon chalk(i.e the stuff used to seal bathroom tiles).  I like them because they are soft which should work well for sound dampening.  Removing the material is also very easy, which means I can do over if I am not satisfied.

My concern with resin is that I had to get it into all the crevices of the transformer in order to stop the coil from vibrating.  I don't have a practical way to do this.  Secondly, there is no do over if the results are not satisfactory(i.e cannot remove hardened resn).  Lastly, resin when cured will harden.  Therefore it may not be as good at sound dampening as a soft material.

Am I right to think soft potting compound is better for sound dampening? 

Any other potting material I should consider?
« Last Edit: September 17, 2017, 10:59:30 am by crazyhog »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28366
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: high pitch noise from a tek 422 oscilloscope
« Reply #20 on: September 17, 2017, 11:24:25 am »
Two pot encapsulation epoxy. Electrician suppliers will have something that doesn't cure fully hard and if warmed will flow nicely into crevices. Ensure the curing time is not too quick as to let any bubbles escape. Some gentle rocking before cure will help them find their way out of the windings.

The one I have is probably some rebrand, Plasti-Cast from Fraser Brown & Strathmore Ltd., Wellington, NZ
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #21 on: September 17, 2017, 12:20:44 pm »
Would it be too troublesome to hook up to the ac source and confirm where and how exactly the noise emanate from, since now it is more readily visible from the open transformer, and then conclude your approach.?    :)
 

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #22 on: September 17, 2017, 02:31:31 pm »
I find it hard to believe any noise would be coming from that coil.  :o

It is well covered with, as you said, a soft rubbery coating.  If there is noise coming from it, it must have been like that from the factory.  The only conceivable place would be from bad factory encapsulation where that compound didn't get into all the places it should have.  Adding anything over top of what's there, quite honestly, won't make it any quieter.  If pushing on the attached wiring didn't make -any- difference, I seriously doubt anything you could add will.  You'd have to completely strip that compound off (without damaging the enamel insulation) and vacuum it down submerged in a new batch. 

I'm just wondering if there is something wonky in the circuit that's causing the noise.  Or, you may simply have a noisy toroid that will never be quiet.

You may be stuck living with it, or adding some insulating and/or sound dampening material exterior to the can.

Can you do what Armadillo asked?

T

veritas odium parit
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #23 on: September 17, 2017, 04:39:35 pm »
The sound is definitely not the 60 hz AC.  I have a isolation transformer on my bench and I know the 60hz hum well.  Here is a 7khz tone on youtube and the noise from the power supply sounded like that.

https://youtu.be/-vAqiFrG9cg

To test the transformer in isolation, I would need a power source that oscillates at 7khz and sources about 1 amps.  Note that the oscillation is not sine, it's PWM.  I would need to rig something up from scratch. Hmm, I could use my function generator to switch a large power transitor to power the transformer.  I'll need to look into this idea more.  Although filling the whole can with silicon chalk is easier.

Stepping back, what other noise makers can there be in this power supply?  I've review the parts list and checked the likely candidates.  Here is the manual I found http://exodus.poly.edu/~kurt/mmm/tek-422.pdf.  Note that we need the AC/DC power supply and not the AC only supply.

Regarding the white stuff on the can lid and wall, it looks thick but its foam.  No sound dampening.  The rubbery coating is only on the winding and its about the thickness of a finger nail.  It just cannot provide much sound dampening, if any.  I won't bother removing this surface coating; there will be crevices within the windings that I will never get at.  At 7khz, the frequency is high enough that I am hopeful that dampening the noise is do able.

 

« Last Edit: September 17, 2017, 04:58:37 pm by crazyhog »
 

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #24 on: September 17, 2017, 05:28:32 pm »
According to the PSU schematic I pointed to in Post 4, the frequency sounds about right.  The reference trace they show is ~125µ sec which comes out to 8kHz.

http://w140.com/tekwiki/images/3/3d/Tek-422_lvps2.png

I believe Armadillo was asking if you could reconnect it back in the circuit with it open, not build something.  If you do that, can you more effectively isolate the noise?  Just touching components would change the frequency or intensity.

If you're absolutely convinced it's the transformer, perhaps some of that "tool grip" rubberized coating?  Once you reassemble it, you could dip the whole coil, casing and all...?

T


veritas odium parit
 
The following users thanked this post: Armadillo

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #25 on: September 17, 2017, 11:50:21 pm »
The disassembly was quite complicated. To put things back into a functioning state will be similarly complex.  I don't want to just put the unmodified transformer back in.

Anyway, $4 worth of silicon caulk is in.  I'll give it a few days to cure and will know.
 

Offline Toasty

  • Regular Contributor
  • *
  • Posts: 98
  • Country: us
  • Component Level Repair Tech
Re: high pitch noise from a tek 422 oscilloscope
« Reply #26 on: September 18, 2017, 12:25:12 am »
How are you going to reattach the rest of the housing with that much caulk?  It is shielding for the transformer...

T
veritas odium parit
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #27 on: September 18, 2017, 02:07:43 am »
Here is a pic with the lid and clamp.  Excess caulk are squeezed out for a perfect fill.  Hopefully the caulk doesn't shrink too much after it's cured. :o
 

Offline Armadillo

  • Super Contributor
  • ***
  • Posts: 1725
  • Country: 00
Re: high pitch noise from a tek 422 oscilloscope
« Reply #28 on: September 18, 2017, 02:55:13 am »
The sound is definitely not the 60 hz AC.  I have a isolation transformer on my bench and I know the 60hz hum well.  Here is a 7khz tone on youtube and the noise from the power supply sounded like that.

https://youtu.be/-vAqiFrG9cg



Good thing, I have a natural 7K filter installed, can't hear a thing.     :palm:
 

Offline voltz

  • Frequent Contributor
  • **
  • Posts: 267
  • Country: gb
Re: high pitch noise from a tek 422 oscilloscope
« Reply #29 on: September 18, 2017, 08:55:44 am »
Not wanting to cause any alarm here, but some silicon sealants react with copper badly. Silicon sealant with high acetoxy will rot copper and turn it green. Check your silicon and see if it contains acetoxy. It will have a strong smell of vinegar if it does.

Its still possibly ok, because the copper wire has enamel covering of course. But years ago i made some traps for a dipole and filled the trap tubes with silicon sealant. Nice job i thought... What i did not know at that time was the silicon had a high acetoxy content. Its an acid released during the curing process. Once its cured it stops reacting. Anyway, my traps stopped working after just 1 week. I took them apart and they were totally green inside. And if i remember correctly, that was even with enameled copper wire.

So just check what you have. Neutral sealants without acetoxy or low content (no vinegar smell) would be fine. Just thought this was worth mentioning possibly for others reading this post and about to silicon any copper wires or parts.
« Last Edit: September 18, 2017, 09:07:16 am by voltz »
 
The following users thanked this post: crazyhog

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #30 on: September 18, 2017, 02:57:04 pm »
 :palm:

The silicon caulk label doesn't list its ingredients.  It has a distinct smell that I can't identify.  The label says low odor.  Most of the transformer is coated at the factory with a white rubbery coat. But the lead wires are only covered with enamel.

I got that pot committed feeling. 
« Last Edit: September 18, 2017, 03:03:59 pm by crazyhog »
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #31 on: September 24, 2017, 12:43:44 am »
Everything is back together and working(pic 1).  The high pitch noise is much quieter but not completely silent.   From 3 feet away, I cannot hear it.  However when I am right up against the scope I can still hear a faint  high pitch noise.  The noise frequency has dampened to around 4khz.  Whereas the AC only Tek 422 is completely silent.

So far no problem with the silicon caulk; fingers crossed.  While waiting for the silicon chaulk to cure, I desolder and check every electrolytic caps in the power supply.  I replaced 4 more with high ESR.  Surprisingly they all still had good capacitance.  They were part of other capacitor banks.  Their higher ESR fractionally increased the capacitor bank's overall ESR.  The power supply is definitely happier now.  The AC current draw has been reduced by about 20%.
« Last Edit: September 25, 2017, 07:45:16 pm by crazyhog »
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #32 on: September 24, 2017, 08:24:48 am »
Recently, I spotted an AC only Tek 422 at a local electronic recycler selling for a song.  Of course I had to rescue it!  Amazingly this machine, older than me, fires right up.  The main problem, as typical, was dirty switches.  Although I still need to check all the capacitors :)

Here are some pics of the AC only 422 and AC/DC 422 scope and a couple of my favorite old timey multimeters. 

Both meters are quite special in that they're hybrids.  They're the bastards of VTVM and VOM!  They are battery powered even though they're part VTVM!  It's crazy!  The world is a more colorful place because they existed!  Hahah!

« Last Edit: September 24, 2017, 08:32:41 am by crazyhog »
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23018
  • Country: gb
Re: high pitch noise from a tek 422 oscilloscope
« Reply #33 on: September 24, 2017, 08:43:16 am »
Never seen those meters before. Rather interesting :)
 

Offline voltz

  • Frequent Contributor
  • **
  • Posts: 267
  • Country: gb
Re: high pitch noise from a tek 422 oscilloscope
« Reply #34 on: September 24, 2017, 08:53:57 am »
Nice. Looking good. :)
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #35 on: September 24, 2017, 09:16:18 am »
The left meter is a Simpson 260 with a Model 651 DC VTVM adapter attached at the base (like Voltron, the defender of the Universe!).  The right meter is a Triplett Model 631 VTVM-VOM.  Both meter can operate in plain VOM mode or in VTVM mode for measuring DC voltage. 

Both came to me in really sad state.  There were short circuits, burnt resistors, and wrong wiring.  It was quite obvious the previous owners tried to repair their meters.  However, the sloppiness of their effort were very clearly evident and that sloppiness pretty much doomed them from the start. 

That sloppiness also sometimes lead to a happy outcome!  For example, the previous owner wired the Triplett 631 meter movement terminals wrong!  As a result no current can ever get into the meter movement and therefore protected it!  The meter movements, if fried, are the only component that can't be repaired.

I'd like to also give a shoutout to Alan Wolke on youtube.  I've learned a lot from his videos.  Here for example is a video on analog meter movement that I studied before I repaired these hybrid bastards.  Allan is excellent at explaining idea simply and clearly.




« Last Edit: September 24, 2017, 09:58:59 am by crazyhog »
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #36 on: September 24, 2017, 10:00:55 am »
Here is a picture of the Tek 422 with the front cover.  The shell, including the cover is made of aluminum with the thickness of a coin!  The whole Tek 422 exudes quality.  No mucking about!

Here is a pic of my favorite VOM: the Simpson 269.  This meter used to cost more than 2x the much more popular Simpson 260. The area of the meter face is 2x that of the Simpson 260.  The meter movement has 100,000 ohm/volt vs the 260's 20,000 ohm/volt.  At the 8 volt range, the impedance is already 800,000 ohm.  Its impedance is close enough to the 1M ohm of a oscilloscope with a 1x probe!  All this from a passive VOM!

There are 3 Simpson 269 series.  The best and rarest is series 3.  It comes with Taut Band meter movement that eliminates needle movement friction.  No need to tap on the meter face to unstick the needle to get an repeatable reading.  The other 2 series doesn't have taut band.  However, series 3 is very rare; I hardly ever see it on ebay.

Anyway, the Simpson 260 is hugely more popular and the Simpson company probably sold 100 to 1 the 260 vs the 269.  You can get a rough estimate by counting how many 260s are for sale on ebay vs the 269.  However, the Simpson 269 easily trounce the Simpson 260 in almost every way.  The Simpson 260 and Triplett 630 fanboys are missing out! :=\
« Last Edit: September 25, 2017, 07:54:41 pm by crazyhog »
 
The following users thanked this post: edavid

Offline The Guy

  • Regular Contributor
  • *
  • Posts: 169
Re: high pitch noise from a tek 422 oscilloscope
« Reply #37 on: September 24, 2017, 06:45:05 pm »
I have one of them scopes... Beautiful machine. I agree.

Good luck gettin it back to life.
 

Offline crazyhogTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: high pitch noise from a tek 422 oscilloscope
« Reply #38 on: September 25, 2017, 08:01:25 pm »
Hah, that guy with the belated well wishes!  :=\   I'll take it anyway :). As I reported in previous comments above, both of my Tek 422 scopes are back in service doing my bidding. >:D
« Last Edit: September 25, 2017, 08:05:45 pm by crazyhog »
 

Offline lordvader88

  • Frequent Contributor
  • **
  • !
  • Posts: 935
  • Country: ca
Re: high pitch noise from a tek 422 oscilloscope
« Reply #39 on: September 25, 2017, 10:54:54 pm »
Just popping in to say thanks for the decent photos. Lovely bit of engineering that scope. I hope you win the battle :)
Oh I never realized it runs off batteries too. I have a working Tek422 on the way soon, can't wait.
 