Howdy everyone, so a few days ago I found a nice arb/function gen for the lab I work in at the surplus property office. It was an HP 33120A 15 MHz arb/function generator, and it is in quite good condition, at least outwardly so. A cursory inspection of the internals looked pretty good - the unit was clean and looked like it had been taken care of pretty well (see below):
I hooked it up and turned it on, and connected the output to my SDS1104X-E ("upgraded" to 200MHz bandwidth) to test the output from the main output BNC terminal. What I got was a flat waveform with a slight DC offset and no oscillitory behavior whatsoever. I verified this over the entire frequency band of the instrument (1 Hz to 15 MHz), and on all channels of my scope:
I first tried to remove the offset, but found that the offset was set to 0 VDC (upper left panel in image below). I figured maybe the amplitude was set too low and was unable to be viewed against this unremovable offset without the waveform being off screen, so I set the frequency to 15 MHz and adjusted the amplitude from its lowest setting to its highest setting, yielding no change in the waveform except for a higher DC offset that corresponded to the 6 attenuation ranges:
Now that I was certain this behavior was not user error, I took a look at the main board more closely in order to find a possible culprit for this behavior. When I glanced over the main output amplifier section, I saw something that made me sad: someone had let the magic smoke out, and in a big way! Four of the resistors in the main output amplifier section had exploded! (
Here's a link to the service manual where I found the schematic)
This is clearly the culprit behind the missing output, although it is heartening to see that the attenuation relays still work. Hopefully this has not damaged any of the downstream power amplifiers, but I am certainly not optimistic about this, and I am at a loss as to what could have caused this level of destruction.
The sync output functions flawlessly, and I tested it over the whole range of frequencies with a 50 Ohm terminator. The rise time appears to be beyond the ability of my scope to measure:
Hopefully this can be fixed, I am thinking I will just solder in replacement resistors and try the output again. Any advice would be greatly appreciated.