After re-reading your reply I realize that is exactly what you meant. When left unplugged the internal NiCAD battery dies causing the unit to do perform some kind of auto cal. Based off this I think i'll take my chances again. Thanks!
Yes, sorry if that was unclear.
Incidentally, my 8643A came to me in "doesn't power up state", with the PSU in tick-tick mode. The first fault was easily isolated to a module by disconnecting them one by one. Turned out a SMD tantalum capacitor was making like a short on the + or -15V rail. After that she came alive but threw calibration errors after power up. The displayed error was quite accurate and descriptive, as it was complaining of a low 5V rail. These instruments are wired for quite extensive self-testing, and the 5V rail was indeed *just* out of spec.
Re-capping the PSU brought everything back into spec (as near as I or the instrument itself can measure).
Artek Manuals have the CLIP, in an excellent scan, although didn't cover the switch-mode PSUs on mine. I believe the 8643A is the only one of the family that has a switch-mode PSU, the others are all linear.