Author Topic: HP E3615A power supply OVP light wont go away  (Read 1486 times)
Offline CaTxTopic starter
« on: April 19, 2018, 07:30:09 pm »
My HP E3615A power supply has an OVP light that wont go away.  Does anyone know what went wrong and how to fix it?  Thanks.
Offline MadTux
Re: HP E3615A power supply OVP light wont go away
« Reply #1 on: April 19, 2018, 10:06:41 pm »
Does clockwise turning the  the OVP pot in the middle of the front panel solve the problem?  ;D ;D
Offline PedroDaGr8
  • A sociable geek chemist
« Reply #2 on: April 19, 2018, 10:17:39 pm »
It might not just be the OVP light, if it is triggering the instant that the device powers on, then you need to look at the turn-on overshoot voltage control circuit as well. The service manual has a lot of good data and schematics, plus test points to observe and diagnose. Including a table of tests to perform and diagnosis to make from said test. Way better and easier than me going through it here.
The very existence of flamethrowers proves that some time, somewhere, someone said to themselves, "You know, I want to set those people over there on fire, but I'm just not close enough to get the job done." -George Carlin
« Reply #3 on: April 19, 2018, 11:02:03 pm »
Nope, turning the OVP Adjust wont help.  I am new, so can someone give me pointer.  Following the service manual now, but it seems like a maze.
« Reply #4 on: April 19, 2018, 11:13:26 pm »
R102 = 146.8 Ohms, in specs
R103 = 1.4xx KOhms, and drifting even when powered down
Am I close to catching the culprit?  What is happening with R103?
« Reply #5 on: April 19, 2018, 11:43:27 pm »
Alright, following Table A-7.
Short U19 pin 4 to TP6 --> OVP indicator OFF --> U20 defective or C57 shorted
Offline PedroDaGr8
« Reply #6 on: April 20, 2018, 04:47:39 pm »
If the indicator went off, that is good news. U20 is an LM393 comparator and C57 is a 50V 47uF electrolytic capacitor. BOTH of those are super cheap to the point that shipping will be WAY more expensive than the part.