Author Topic: Fixed HP1741A in a big shape but... I bet in a focus problem, and you?  (Read 65694 times)
Offline SAUL BRITTOTopic starter
  • Country: br
  • Don't Make a Mistake
« Last Edit: February 26, 2016, 06:45:04 pm by SAUL BRITTO »
Thank You, for all earth.
Offline Alex Eisenhut
  • Place text here.
Re: HP1741A in a big shappe but... I bet in a focus problem, and you?
« Reply #1 on: February 05, 2016, 01:43:05 am »
You are in storage mode, you might have some trapped charge. Turn down the beam intensity to 0 and press erase a few times. If that doesn't work, there is a deep erase button on the back, follow the procedure in the manual.
  • Country: br
« Reply #2 on: February 05, 2016, 01:56:54 am »
Thank you Alex Eisenhut. Actually this scope is still coming to me, this is a picture that the seller sent me.
I hope you are right.
Offline edpalmer42
« Reply #3 on: February 05, 2016, 02:59:33 am »
I have the 1744A - an enhanced version of the 1741A.  I didn't look it up, but I think the writing speed on mine is much faster.  I normally run with the Brightness control fully CCW, the intensity much lower than shown in the picture, and the focus at about 12:00.  With the Brightness control turned up, blooming like in the picture is guaranteed to happen.  You have to balance the Persistence, Brightness, and Intensity.  They all have to change as you change the Horizontal speed.  But when you get them right, it's glorious!  :)
  • Country: br
« Reply #4 on: February 05, 2016, 03:08:15 am »
What  really does that deep erase? Of course I suppose, but I would like a  clarification if possible.
« Last Edit: February 05, 2016, 03:13:29 am by SAUL BRITTO »
  • Country: br
« Reply #5 on: February 05, 2016, 03:36:03 am »
Deep erase on the manual looks like a dangerous process!!!
Offline coppice
« Reply #6 on: February 05, 2016, 03:51:50 am »
In what way was the 1744A enhanced? I remember it as cost reduced model with a limited display area.
Offline edpalmer42
« Reply #7 on: February 05, 2016, 04:05:24 am »
I don't know.  The 1744A doesn't have 'Deep Erase'.  The manual only talks about B-scan for erasing deeply stored images, but I've never used it.
Ed
Offline edpalmer42
« Reply #8 on: February 05, 2016, 04:11:44 am »
Oh, so now you're going to force me to look it up!  Fine!  Be like that!   :box:    :rant:    Just kidding!   ;)
They're both 100 MHz, but the 1741A has a storage writing speed of > 100 cm/us.  The 1744A has a storage writing speed of > 1800 cm/us.
  • Country: br
« Reply #9 on: February 05, 2016, 04:21:52 am »
Look at this:
  • Country: br
« Reply #10 on: February 05, 2016, 04:23:59 am »
But,guys:  can't be a focus defect, anyway?
« Last Edit: February 05, 2016, 04:46:25 am by SAUL BRITTO »
Offline edpalmer42
« Reply #11 on: February 05, 2016, 04:42:05 am »
But,guys: this can't be a focus deffect, anyway?
Yes, it could be defective.  Until you get it on your bench you won't know for sure.
  • Country: br
« Reply #12 on: February 05, 2016, 04:49:08 am »
I think will measure this voltage very well, I do have a high V RCA probe that measures 58kv,maybe this focus V is around 2kv?
« Last Edit: February 05, 2016, 04:51:55 am by SAUL BRITTO »
Offline edpalmer42
« Reply #13 on: February 05, 2016, 05:19:59 am »
Okay Saul, I'm onto you now!  You're just looking for an excuse to take it apart and poke at it.  Haven't you learned yet?  We don't need an excuse!  We do it 'cause we want to!  :-+  ;D
Just follow the procedure in the manual to measure the HV.  It tells you where, how, and what you should measure.
  • Country: br
« Reply #14 on: February 05, 2016, 05:26:00 am »
No I do not have it in my hands right now. I'm just trying to get an idea of what I purchased blindly. I have two HP1740A and already have a little experience with these scopes not so much with the hp1741a, for its storage resource
« Last Edit: February 07, 2016, 01:29:03 am by SAUL BRITTO »
Offline oldway
« Reply #15 on: February 05, 2016, 12:59:32 pm »
In my opinion, this scope has no fault, only wrong settings.
The guy who sell this scope does not know how to operate this scope.
See on brighter picture:
What's wrong ?
Delay time base selected instead of main time base.
Main time base not triggered because no input signal and automatic trigger is not selected.
Focus almost anticlockwise.
« Last Edit: February 05, 2016, 03:26:15 pm by oldway »
  • Country: br
« Reply #16 on: February 05, 2016, 04:59:22 pm »
oldway> I do not know if the seller is a technician or has any knowledge in this area
« Last Edit: February 06, 2016, 05:09:46 am by SAUL BRITTO »
Offline oldway
« Reply #17 on: February 05, 2016, 07:00:56 pm »
oldway> I do not know if the seller is a technician or have any knowledge in this area
Looking at the picture, he has not for sure  :-DD
  • Country: br
« Reply #18 on: February 05, 2016, 07:52:08 pm »
Good laughs  here! :-DD
  • Place text here.
« Reply #19 on: February 06, 2016, 02:52:03 am »
I agree with oldway, I remember seeing that kind of bloom when the settings are wrong. I think leaving a bloom like that in the CRT while it's OFF is not a good idea.
But if it survives shipping, it means at least everything is working.
Deep erase can be dangerous, it's only a last resort.
I no longer have the 1741, I gave it away years ago. But I remember it had a burned-in image from the lab it came from, it took a lot of erasing and fiddling to get the image to disappear.
If you press erase a few times and it doesn't go away, run it in auto erase with no input signal, put it in auto mode with no beam intensity and it will store nothing and erase the screen continuously.
This is also in the manual IIRC.
  • Country: br
« Reply #20 on: February 06, 2016, 05:13:16 am »
Thank you Alex.I believe now will look this scope with a bit more care
  • Country: br
« Reply #21 on: February 06, 2016, 05:16:53 am »
Guys please, be pacient with my "english" because I speak portuguese and to me it's beeing hard.
  • Country: br
« Reply #22 on: February 06, 2016, 05:20:58 am »
I already have some experience with this  HP scopes because, I have a few more 1740A that was defective and repaired myself and I think they are excellent.
« Last Edit: February 06, 2016, 05:26:27 am by SAUL BRITTO »
  • Country: br
« Reply #23 on: February 06, 2016, 05:23:10 am »
It cost me about $ 40 with shipping fee. It was a good price?
« Last Edit: February 06, 2016, 05:24:52 am by SAUL BRITTO »
  • Country: br
« Reply #24 on: February 11, 2016, 11:52:54 pm »
Hi friends. When I worked with tube tv, seen many triplers causing a similar defect. As the anode lost the ability to attract the electron beam causing, this type of screen. Tomorrow the courier will deliver the scope in my home.
« Last Edit: February 12, 2016, 12:04:57 am by SAUL BRITTO »
  • Place text here.
« Reply #25 on: February 12, 2016, 12:58:30 am »
Good luck! But before tearing everything apart, remember it's tricky to get working in storage mode; don't assume it's broken right away!
  • Country: br
« Reply #26 on: February 12, 2016, 02:40:54 am »
Hi Alex,how are you guy? Thank you for advise.
I'm not interested storage. I just want it running smoothly
« Last Edit: February 12, 2016, 02:48:54 am by SAUL BRITTO »
  • Country: br
« Reply #27 on: February 12, 2016, 02:43:28 am »
I do have a tektronix 2220 that have storage mode but ,I never used,my use of oscilloscope is very simple
  • Country: br
« Reply #28 on: February 12, 2016, 10:20:32 pm »
Hi All. The scope finally is in my bench. I did a small video,look:
https://youtu.be/SpsUYWx4-_0
« Last Edit: February 12, 2016, 10:25:55 pm by SAUL BRITTO »
Offline rf+tech
  • Real radios are olive drab
« Reply #29 on: February 13, 2016, 03:44:55 am »
Hello Saul,
First thing to try is turn OFF storage mode.
Turn the scope around, on the back panel is a pushbutton marked "Conv". This to the right of the black CRT cover.
Press this button to place the scope in Conventional mode.
The "Conv" light on the front panel should now be ON and the scope should display a trace that can be properly focused.
Here's the HP1741 manual: http://cp.literature.agilent.com/litweb/pdf/01741-90911.pdf.
Section 3 covers the operation of this analog storage scope, and Section 5-59 describes the storage adjustments procedures.
Without the knowledge necessary to use analog storage mode - and most importantly - how to properly adjust these "unconventional" scope controls, it is possible to end up looking for something that does not need to be repaired. The blooming present in the video appears "normal" to me for an old scope that has not had the storage adjustment procedure performed in many years.
I once owned an HP1727 275 MHz storage scope, getting the collimator and flood guns properly adjusted on these old HP analog storage scopes can be a challenge. Once adjusted, the illumination may still not be uniform across the screen. It all becomes a compromise.
RF+ Tech
RT-1133 AN/PRC-70  *  RT-794 AN/PRC-74  *  RT-841 AN/PRC-77  *  RT-524 AN/VRC-12  *  RT-834 AN/GRC-106  *  RT-F100
  • Country: br
« Reply #30 on: February 13, 2016, 04:20:59 am »
Wow! Thank you so mutch rf+tech.It was of a big help to me! I will do this and report what happened.
  • Country: br
« Reply #31 on: February 13, 2016, 04:30:34 am »
 rf+tech when I use conventional mode I get a black screen. No trace,no bright, nothing but,I found a transistor
Q17 on the vertical board that seems in short thre is a diode near to it that looks in short. There is a -12 Volts with -14,2V. You see a connection between these things?
« Last Edit: February 13, 2016, 04:37:25 am by SAUL BRITTO »
  • Country: br
« Reply #32 on: February 13, 2016, 04:57:39 am »
  • Country: br
« Reply #33 on: February 13, 2016, 05:02:34 am »
At this place has a 2n4035
« Last Edit: February 13, 2016, 05:05:30 am by SAUL BRITTO »
Offline rf+tech
« Reply #34 on: February 13, 2016, 01:21:32 pm »
Saul,
Where is this diode that may be shorted? I see no diodes in the attached schematic.
Please provide the voltages measured on each lead of Q17. One measurement by itself is meaningless.
Since Q17 and Q18 comprise a differential amplifier, it is better to evaluate both sides together.
Let's see all six measurements.
And have all power supply voltages been checked, before jumping into the vertical amplifier?
RF+ Tech
  • Country: br
« Reply #35 on: February 13, 2016, 03:54:48 pm »
Saul,
RF+ Tech
Q17 is: base -12.17V, collector and emmiter -14.12V. Q18 is working very very hot! I will to see all the PS, despite I did a preview yesterday night and, just found the 120V rail with 124V and, on the vertical board a -12V with -14,2V,That diode is fisicly near to  this transistor but on schematic no, it isn't in this image.
« Last Edit: February 13, 2016, 04:10:35 pm by SAUL BRITTO »
Offline rf+tech
« Reply #36 on: February 13, 2016, 04:54:38 pm »
Saul,
Something to make note of: the A17 schematic shows +/- 12 Volts coming from A3. The A3 schematic does not show where +/- 12 Volts comes from.
Looking at the interconnecting diagrams, sheet 1 of 3 shows +/- 12 Volts at A3 block but *nowhere* can I find the source for +/- 12 Volts.
Since you measure 14.2 Volts where the schematic indicates 12 Volts, and there is no 12 Volt supply to be found, the conclusion is an error in the manual. HP was very good about releasing errata/change notices in paper form. Without access to these for the 1741, we have to conclude from the evidence of your observations that 12 Volts is really 15 Volts. 14.2 Volts is only 5.3% low from 15 Volts and does not present a problem.
Please report what the -12 Volt point measures to see if this theory is further supported.
This illustrates the importance of conveying as much information as possible. With minimal information, we can only waste time guessing at things. Even if you think the power supply voltages are good, post them for our benefit. Everyone sees things a little differently.
As for Q18 running hot, consider that the A17 is DC coupled from the A3 - specifically from A3A1. And the output balance of A3A1 depends on many more inputs. It is possible that Q18 is hot because the real problem is in the A3.
By providing the DC voltage measurements on both Q17 and Q18 emitter, base and collector, we can better evaluate and advise on the problem.
The 120 Volt supply measuring 124 Volts is of little concern as this is only 3.2% high.
RF+ Tech
  • Country: br
« Reply #37 on: February 13, 2016, 05:07:21 pm »
That diode is connected to a -15V rail,but I can not find in the schematic,for now.
  • Country: br
« Reply #38 on: February 13, 2016, 05:21:43 pm »
RF+ Tech >> I do have two more boards like this one (Two scopes) and those boards has different measures with the DMM in diode scale, the junction colletor emmiter of Q17 is 0.600v on the other boards and on the sick scope is full short 0.0V on diode scale. That one diode measures 0.2V and on the other board is 0.500V on diode scale.
Q17 and Q18 has each one a equal diode near it,they are connected to a -15V rail,but on the A3 vertical board I just found a transistor in this rail,at the eschematic. If you prefer I can make an image of exactly place on the board.
« Last Edit: February 13, 2016, 05:54:29 pm by SAUL BRITTO »
  • Country: br
« Reply #39 on: February 13, 2016, 05:30:42 pm »
Here is the image.Sorry for this webcam.
« Last Edit: February 13, 2016, 05:32:54 pm by SAUL BRITTO »
  • Country: br
« Reply #40 on: February 13, 2016, 05:41:59 pm »
The pins 5 and 6 of this connector are -12 and +12V rails.Near to Q17 has a test point connected to -12V that's measuring -14,2V.
  • Country: br
« Reply #41 on: February 13, 2016, 06:20:25 pm »
If you see on the schematic Q17 is polarized with 12.6V on emmiter and 0.6 on collector but,here is measuring -14,2V base and emmiter and Q18 on the schema is PNP on the board is NPN wow!!
« Last Edit: February 13, 2016, 06:30:24 pm by SAUL BRITTO »
Offline rf+tech
« Reply #42 on: February 13, 2016, 06:47:44 pm »
Very good information Saul.
The diodes are not likely to be VR2 and VR3, on the right side of the A17 schematic, based on their connection to -15 Volts. Do the other two boards look identical to this defective board, in this area? The fact that HP revised this differential amplifier and added these diodes, together with +/-14.2 Volts on the "+/-12" Volt rails adds confirmation that these are indeed +/- 15 Volt rails.
The diode test on Q17 certainly is conclusive. The diode test on this unknown diode may be influenced by the shorted Q17. Can you see if this diode connects in some way to Q17, perhaps through a low value resistor? It would be good to unsolder one leg of the suspect diode and try the diode test again.
I don't see the mystery Q13 (on the A3 parts layout page) on the A3 schematic, either. So far, this manual has not been very helpful. More searching has not turned up different scans of the manual.
Just now saw your new post - it would appear that someone has "repaired" this board. I "think" some people do this just to help us sharpen out troubleshooting skills. :box:
For a quick fix, a 2N3906 should work. The bandwidth of the amplifier may suffer, but at least we will know more.
RF+ Tech
  • Country: br
« Reply #43 on: February 13, 2016, 06:53:10 pm »
I removed the Q17 it's in full short and the short on the diode was your fault,diode is good.
  • Country: br
« Reply #44 on: February 13, 2016, 06:59:11 pm »
That -12V seems to have origin in any way in Q17 because when I removed Q17 did a measure on that -12V and it is 0.2V  now. On the others boards are absolutly equal all you find in one find on the other and,Q18 is a NPN transistor on all that.
« Last Edit: February 13, 2016, 07:01:36 pm by SAUL BRITTO »
  • Country: br
« Reply #45 on: February 13, 2016, 07:14:24 pm »
I have an original one Q17 but, I think: what cold be te cause of this short in Q17? What do you think about?
« Last Edit: February 13, 2016, 07:24:48 pm by SAUL BRITTO »
  • Country: br
« Reply #46 on: February 13, 2016, 08:08:28 pm »
Replaced the Q17 and some things changed your behavior,that big IC on the vertical board seems working normally,
Beause it was working cold and now is in a temperature as the other scopes. Just don't have a trace,maybe I have a tube polarization problem or in lacking of HV.
  • Country: br
« Reply #47 on: February 13, 2016, 08:42:17 pm »
I found a tantalum cap in short on the vertical board,I will replace and see what happens
Offline rf+tech
« Reply #48 on: February 13, 2016, 09:59:50 pm »
That is too strange to find NPN transistors in place of PNP, especially in a differential amplifier where both sides mirror each other. Have you compared the E-B and B-C diode drops to a known PNP or NPN to confirm this?
On another forum, I found a marked up A5 schematic for the HP1740 vertical amplifier. It is very similar to the A17 and someone has recorded many voltage measurements that may be of help:
HP 1740A Vertical Mod -0001: http://www.electro-tech-online.com/attachments/hp-1740a-vertical-mod-0001-pdf.44405/
Measuring the voltage on the vertical deflection plates will show if the amplifier can be balanced when turning the vertical position control, or if there is still another problem to locate. Knowing that the trace is not off the top or bottom is more important than jumping to the high voltage power supply.
You are making good progress. :-+
RF+ Tech
  • Country: br
« Reply #49 on: February 13, 2016, 10:06:28 pm »
rf+tech>> I am almost sure that the tripler is bad,Has a 58kv probe here and the way is to measure this.
I did measure the HV on the output of tripler and got about 2.8kV I think it'snt right...
« Last Edit: March 04, 2016, 04:55:45 pm by SAUL BRITTO »
  • Country: br
« Reply #50 on: February 13, 2016, 10:11:04 pm »
Sure. I have a 1740A unit,do you know if the tripler of  "40A" is the same as 41A?
« Last Edit: February 15, 2016, 10:51:11 am by SAUL BRITTO »
  • Country: br
« Reply #51 on: February 13, 2016, 10:26:37 pm »
On the schematic I see anything that looks like 5kV and, can't see very well because  the copy is not so good.
  • Country: br
« Reply #52 on: February 13, 2016, 10:41:52 pm »
Ok confirmed the HV is +5kV but just js generates 2.6 or 2.8 kV post tripler.
  • Country: br
« Reply #53 on: February 13, 2016, 11:00:46 pm »
You are surely right about the transistors but, it's incredible,look http://alltransistors.com/transistor.php?transistor=2128 this is the transistor 2n2536 (Q18) how do you see it is a PNP Tr but when a measure it just seems a NPN Tr black probe on the base dont works, when I invert the probes RED on the base it works as a NPN Tr wow!
 And this in two Boards There is a one I know: It works well.
« Last Edit: February 15, 2016, 10:52:07 am by SAUL BRITTO »
  • Country: br
« Reply #54 on: February 13, 2016, 11:48:13 pm »
rf+tech>> I am almost shure that the tripler is bad,Has a 58kv probe here and the way is to measure this.
Ok.I measured the DC level in the plates and saw 70V and 36V on the channel that has 70V changing the potentiometer varies between 70 and 69V and 36V which has nothing happens
Offline rf+tech
« Reply #55 on: February 14, 2016, 12:00:44 am »
Saul,
The HV adjustment procedure is in paragraph 5-15. This says the HV at A15TP4 is to be -2250 Volts, +/- 50 Volts.
The specifications given on page 1-5 indicate the CRT post accelerator voltage is approximately 7.5 kV.
Somehow I have missed the change notice section in the manual. There is a backdating schematic for the A15 HVPS on page 7-3 that shows approximately 5 kV from the tripler. I agree that 2.8 kVolts is not good.
If the horizontal and vertical plate voltages show the beam can be centered on-screen, I think there would be some sign of a dim trace.
A17 has many changes shown on page 7-2. There are new A17 schematics and parts placement drawings included in figures 7-6, 7-7 and 7-8. The photo you made with Q17 and Q18 circled does not match up with either parts placement. Note that the parts placement shows Q17 and Q18 are further in from the edge of the board and next to A1 which is square.
The transistors in your photo are from a different place, they are next to a 14 pin IC that is rectangular, and much closer to the edge of the board. The transistors you have marked as Q17 and Q18 are not correct. That would explain finding NPN where PNP is expected and explain the presence of the diodes, that are not shown in the schematic.
All of these differences are much too confusing. Every time I look closer at this manual, more differences are seen and this makes it more difficult to be of help.
2N2536 is a germanium transistor. The new question of the minute is why HP would use germanium at this late time in semiconductor history?
RF+ Tech
Offline rf+tech
« Reply #56 on: February 14, 2016, 12:05:28 am »
Since we just cross-posted,
So the differential amplifier cannot be balanced. We would expect both plates to become equal voltage near the center of the potentiometer travel. And this tells us that the beam is deflected heavily off-screen.
RF+ Tech
  • Country: br
« Reply #57 on: February 14, 2016, 12:09:49 am »
I just see any sgnal of trace on the non-conventional mode,on the conventional just a black screen.
To tell the truth is not a trace only a shadow of something moving on the screen.
« Last Edit: February 14, 2016, 12:11:24 am by SAUL BRITTO »
  • Country: br
« Reply #58 on: February 14, 2016, 12:15:04 am »
Really a just see a shadow of a trace out of screen.
  • Country: br
« Reply #59 on: February 14, 2016, 12:17:58 am »
But how can a PNP Tr acts like a NPN? sex change?  :-DD
« Last Edit: February 14, 2016, 12:30:07 am by SAUL BRITTO »
  • Country: br
« Reply #60 on: February 14, 2016, 12:33:22 am »
I will remove it from the board to see what it is really
Offline rf+tech
« Reply #61 on: February 14, 2016, 12:48:56 am »
Of course, the transistor will be an NPN transistor.
From everything I have seen in the parts placement drawings, compared to your photo, you are looking at the wrong transistors.
Whatever they are, they are not Q17 and Q18. They are some other transistors. Study the drawings and look carefully at the surrounding components. Can you now see what I see?
If you do not agree, then I must be looking at the wrong board. Is it not A17 we are discussing?
RF+ Tech
  • Country: br
« Reply #62 on: February 14, 2016, 01:03:09 am »
Oh it is true in part, Q17 is really 17 in the drawings.
Q18 that I show in the photo (Sorry) is actually Q19, I apologize to my friend, my eyes already are giving signs of aging
« Last Edit: February 14, 2016, 02:22:38 am by SAUL BRITTO »
  • Country: br
« Reply #63 on: February 14, 2016, 01:11:38 am »
In the drawings I see Q19 but on the schematic no.
  • Country: br
« Reply #64 on: February 14, 2016, 01:30:11 am »
Q17 was that metallic I replaced and Q18 is a small transistor as a BC548 in your way.
Offline rf+tech
« Reply #65 on: February 14, 2016, 02:22:54 am »
Saul,
No problem, my eyes are likely much worse. Sometimes it is easy to get lost between schematics and boards. Some of what I work on has no schematics available and still I manage to repair everything.
An observation about Q19, it is shown in figure 7-8 "Schematic 16 Replacement." Look just to the right of U6.
RF+ Tech
  • Country: br
« Reply #66 on: February 14, 2016, 02:38:43 am »
Thank you rf+tech for your help and your time and patience.I am back to electronic after a long time out.
I am a bit oxidized, rusty!
« Last Edit: February 14, 2016, 07:08:12 am by SAUL BRITTO »
  • Country: br
« Reply #67 on: February 14, 2016, 03:31:10 am »
Yes  rf+tech I found Q19 on the page 117 near to U2 main trig amp.
  • Country: br
« Reply #68 on: February 14, 2016, 05:29:03 am »
rf+tech I really think that the tripler is lowering the HV circuit, is there any danger if disconnect it's ground pin?
« Last Edit: February 14, 2016, 06:49:31 am by SAUL BRITTO »
Offline rf+tech
« Reply #69 on: February 14, 2016, 02:44:26 pm »
Saul,
Disconnecting the ground pin on a HV tripler sounds like a very good way to inflict more damage on this scope. What is your expectation in taking such a risk?
Even if you manage to get the HV up to spec, there will be nothing to see on the screen.
Why?
Because the vertical deflection amplifier is forcing the beam off screen. Fix this first.
Does the vertical imbalance exist at the output of A3A1?
If so, have you disconnected the A3 preamp output from the A17 input and checked the A3 output to see if the imbalance is still present?
If so, have you checked to see if some input to the A3A1 is forcing the output out of balance?
If the A3A1 is defective and a replacement is not economical, why bother fixing the HV?
What good is a scope with working HV if the vertical amp is beyond repair and cannot bring the beam on screen?
Jumping to conclusions, jumping from one problem to another is not only poor troubleshooting technique, it makes far greater difficulty for people to follow and be of assistance.
Good troubleshooting technique stays focused on one problem at a time, makes lots of measurements from which to draw logical conclusions, checks and double-checks anything that appears to defy expectations then splits circuits in half to further isolate individual stages.
RF+ Tech
  • Country: br
« Reply #70 on: February 14, 2016, 04:43:20 pm »
I was thinking about short on the tripler. But I did it and have not changed anything, the voltage did not rise If there is a problem in tripler will be open diod.
« Last Edit: February 14, 2016, 04:52:28 pm by SAUL BRITTO »
  • Country: br
« Reply #71 on: February 14, 2016, 04:55:47 pm »
Saul,
Why?
RF+ Tech
Ok I will make these tests.
  • Country: br
« Reply #72 on: February 14, 2016, 05:01:08 pm »
I bought this scope very cheap just for remove your vertical skills,I have an 1740A with a problem in vertical skill
And was this my intention.But, I thought it was worth trying to fix the 1741A,and in fact I did not even look right if it is fully compatible skills. :-BROKE
  • Country: br
« Reply #73 on: February 14, 2016, 05:03:37 pm »
Really you're right about the sequence analysis
Offline rf+tech
« Reply #74 on: February 14, 2016, 06:27:05 pm »
Sual,
Maybe what we learn about the 1741 vertical problem will help repair the 1740 vertical problem.
I have been scrolling around in the 1741 manual and see there is a second schematic for A3 located five pages past the first schematic. This is figure 8-21 on page 8-29 and shows where the +/- 12 Volt regulators are located.
Perhaps you have been referring to Q17 and Q18 on A3, while I have been referring to Q17 and Q18 on A17. This has added to the confusion of finding an NPN in place of a PNP. The physical placement of Q17 and Q18 on the A3, close to U2 seems to match your photo. So yes, we have not been talking about the same board.
On the principle that measurements made with incorrect supply voltages are themselves incorrect and lead to incorrect conclusions, let's see why there is +/- 14.2 Volts where there should be +/- 12 Volts.
Looking at the A3 schematic, R123 and R124 bias the base of Q20 at +12.1 Volts. Q20 emitter and Q19 base should be about 0.6 Volts higher, +12.7 Volts. The emitter of Q19 should be about 0.6 Volts lower than the base, +12.1 Volts.
The -12 Volt regulator is a mirror of the +12 Volt regulator.
Let's see why both of these are not working correctly, before going ahead with the vertical deflection problem.
RF+ Tech
  • Country: br
« Reply #75 on: February 14, 2016, 08:30:12 pm »
Is about this your comment?When you say: Does the vertical imbalance exist at the output of A3A1?
« Last Edit: February 14, 2016, 08:38:18 pm by SAUL BRITTO »
  • Country: br
« Reply #76 on: February 14, 2016, 08:33:46 pm »
In CR29 and CR30 I found a low resistence,where is marked 12V I get a small variation changing Position.In one channel is around 0.3 V on the other almost nothing it's stoped in 12,1V
  • Country: br
« Reply #77 on: February 14, 2016, 09:08:31 pm »
Yes on the anode of diode CR27  I get 14.97V  and on the collector of Q19 14.2V. 14.97-0.7 of diode= 14.2 ok?
 On the base 12.7 and emmiter 12.01. 12.7-0.6 (base emmiter)= 12.01. Correct?
This is that I get here.
  • Country: br
« Reply #78 on: February 14, 2016, 09:10:25 pm »
If you are not so experienced I think it would here turn a Babylonian tower :-DD
  • Country: br
« Reply #79 on: February 14, 2016, 09:19:22 pm »
You remember,that one in short? Was Q17, +12 -12 are ok.I replaced Q17 with an  original one.
« Last Edit: February 15, 2016, 01:45:47 am by SAUL BRITTO »
Offline rf+tech
« Reply #80 on: February 14, 2016, 09:47:13 pm »
I don't understand about these voltage measurements - from where to where?
The back to back diodes prevent the *differential* voltage from exceeding about 1.2 Volts - as measured directly across the two outputs. When this preamp is working properly, with a meter connected across the two outputs, the vertical position control should swing this voltage from +0.6 Volts, through zero Volts and to -0.6 Volts. By adjusting the vertical position control for zero volts - balanced output - the beam will be almost exactly centered.
If the series combination of R64, R65 and R66 in parallel with the diodes is close only several hundred Ohms, the diode test reading will not be reliable.
You remeber,that one in short? Was Q17, +12 -12 are ok.I replaced Q17 for an  original one.
This is where we were Babel-ing on about Q17 in two different places. I was looking at the A17 and you were looking at the A3.  :blah:
I can Babel with the best, I do this professionally in my own business - and repair things too.
RF+ Tech
  • Country: br
« Reply #81 on: February 14, 2016, 10:00:59 pm »
Measuring in CR 29 and CR30 Vertical output, I get a variation of -403mV to +500mV,in both channels.
« Last Edit: February 14, 2016, 10:11:31 pm by SAUL BRITTO »
Offline rf+tech
« Reply #82 on: February 14, 2016, 11:47:30 pm »
That is good to see. Set this to close to zero Volts, reconnect the A3 output to the A17 input and let's measure the voltage from the collector of Q17 to the collector of Q18 on the A17. It should be close to zero Volts here and move positive/negative with the vertical position control. Set the vertical position back to zero Volts.
Then proceed to measure from the cathode of VR2 to the cathode of VR3. The same positive/negative swing should be measured here, but with much greater voltage range. Again, set the vertical position control for zero Volts.
Each point where the voltage can be set to zero, the two halves of the differential amplifier will be balanced. When the output across the deflection plates can be set to zero Volts, we will know the beam will be very close to the center of the screen.
Let's see how this goes.
  • Country: br
« Reply #83 on: February 15, 2016, 12:25:57 am »
rf+tech,when I put a probe in X1 to channel B Using CAL 1V at .1V scale I get around 0.6 Volts pp, in other scope, after delay line on the input of that power vertical circuit,when I put position full left it shows 0.2V at the other scope and when I put full right it goes to 0.2V too.
Offline rf+tech
« Reply #84 on: February 15, 2016, 12:45:55 am »
Saul,
For now, it is not important about the difference between the two scopes. Our concern is only with no signal, will the difference adjust to zero?.
Set the input coupling to GND and continue with the procedure above.
Once we know that the differential output can be balanced to zero Volts all the way to the deflection plates, then we can evaluate this difference between the two scopes.
RF+ Tech
  • Country: br
« Reply #85 on: February 15, 2016, 12:52:40 am »
Are that diodes  a kind of 1n4148? it shows 13.8 to -36.4. I did all that tests and very near to 0V, i can but, it's very critical I just can 0.03V,ok?
« Last Edit: February 15, 2016, 01:36:25 am by SAUL BRITTO »
  • Country: br
« Reply #86 on: February 15, 2016, 12:56:44 am »
rf+tech> Ok, it was the same way with GND connected,I get 0V and the other voltages above.
« Last Edit: February 15, 2016, 01:18:11 am by SAUL BRITTO »
  • Country: br
« Reply #87 on: February 15, 2016, 01:10:57 am »
Guys please sorry for my crazy english!
Offline rf+tech
« Reply #88 on: February 15, 2016, 01:45:40 am »
Saul, your English is not a problem. My problem is guessing about missing details to be certain we are talking about the same thing, the same place and the same type of measurement.
If that diodes are as 1n4148 it shows 13.8 to -36.4.
Guessing that 13.8 to -36.4 is from one diode to ground?
So the difference across the deflection plates can be set to zero, then the entire vertical amplifier is working okay, except maybe the gain may be wrong.
it's very critical it was 0.03,ok?
0.03 Volts is close enough to zero and that is very good.
So you are ready for the next problem, yes?
  • Country: br
« Reply #89 on: February 15, 2016, 01:50:13 am »
OH sorry it was to measure cathode to ground? I did it cathode to cathode,but I can do it to ground.Please confirm that I did it wrong.
  • Country: br
« Reply #90 on: February 15, 2016, 01:56:36 am »
Measuring to ground it is 43.9 to 70 Volts in one diode. The other was 32.7 to 50.0 Volts
  • Country: br
« Reply #91 on: February 15, 2016, 02:03:32 am »
If I made it wrong I interpreted you text in bad way,when you said: That is good to see. Set this to close to zero Volts, reconnect the A3 output to the A17 input and let's measure the voltage from the collector of Q17 to the collector of Q18 on the A17. It should be close to zero Volts here and move positive/negative with the vertical position control. Set the vertical position back to zero Volts.
"let's measure the voltage from the collector of Q17 "to the collector of Q18 on the A17."
It was to measure the difference between the two collectors,the two probes on the collectors?
« Last Edit: February 15, 2016, 02:09:30 am by SAUL BRITTO »
Offline rf+tech
« Reply #92 on: February 15, 2016, 02:13:24 am »
Thank you for clarifying. Both types of measurement are valid.
Maybe my use of the word differential is part of the problem. Push-pull may be easier understand. The deflection plates push and pull the beam, depending on the voltage across the deflection plates and the polarity of one plate to the other plate. The actual voltage from one plate to ground is of lesser importance.
Here is a thought experiment: with only one meter, measure the voltage from one deflection plate to ground. Next measure the other plate to ground. Then adjust the vertical center control, measure one plate voltage to ground, then the other, back and forth until both plates are at the same voltage. Now measure the voltage across the two plates and it should be close to zero Volts. This is the condition required for the beam to be in the middle of the screen.
See how much easier it is to simply place the meter across both sides of a push-pull amplifier and adjust the vertical center control for zero Volts? This is why I am saying the voltage to ground is not important. What is most important is the difference between the two sides.
Okay, I am very slow to write my thoughts.
Yes, this is the correct way. Please, if there is anything you do not understand, ask. I do not want you to feel frustrated.
  • Country: br
« Reply #93 on: February 15, 2016, 02:23:45 am »
Thank you rf+tech! When you says: adjust de vertical center,you want to say that adjust ,on the board, not the position,right?
« Last Edit: February 15, 2016, 02:26:32 am by SAUL BRITTO »
  • Country: br
« Reply #94 on: February 15, 2016, 02:31:37 am »
I see on the board polarity balance... This is on the shield cover with drawings.
« Last Edit: February 15, 2016, 02:34:51 am by SAUL BRITTO »
Offline rf+tech
« Reply #95 on: February 15, 2016, 02:34:30 am »
When you say adjust de vertical center,you want to say that adjust on the board, not the position,right?
No, adjust the "vertical position" control on the front panel. Think of how the vertical position control works on a good scope. The beam moves up or down with the vertical position control.
The adjustments on the board should not be altered yet. This may add more confusion.
Please allow me more time to write.
  • Country: br
« Reply #96 on: February 15, 2016, 02:36:29 am »
Oh my friend you have all the time of the word! Relax. Maybe I am the nervous here ! :-DD
« Last Edit: February 15, 2016, 02:38:26 am by SAUL BRITTO »
  • Country: br
« Reply #97 on: February 15, 2016, 02:50:43 am »
rf+tech > near to zero I can, around 0.053mV.
  • Country: br
« Reply #98 on: February 15, 2016, 02:53:27 am »
I am measuring with a keysight DMM but has a fluke 179 here...
  • Country: br
« Reply #99 on: February 15, 2016, 02:58:30 am »
Ok I did the work. I can measure 0.0V on the two channels but with de position a bit to left, around 20%.
« Last Edit: February 15, 2016, 03:06:34 am by SAUL BRITTO »
Offline rf+tech
« Reply #100 on: February 15, 2016, 03:08:34 am »
I can measure 0.0V on the twa channels but with de poisition a bit to left around 20%.
Great!  :-+
No concern about the 20%. When the alignment procedure in the manual is performed, this error will be corrected.
If it will help to understand, the deflection plates have a sensitivity expressed in Volts per centimeter. A guess, maybe 10 Volts deflects the beam by 1 cm. One Volt will deflect the beam 1 mm. 0.1 Volt will deflect the beam 0.1 mm. 0.053 Volts will deflect the beam .053 mm. Can you see that small?
  • Country: br
« Reply #101 on: February 15, 2016, 03:29:11 am »
Good! But I don't know if you remember: this scope don't have a visible trace.I said to you: it has just 2.8 kV when it was to have at least 5kV.
  • Country: br
« Reply #102 on: February 15, 2016, 03:32:06 am »
I don't know right now who is the culprit, maybe fllyback or tripler ou another component in HV or LVPS.Despite I did measure all the voltages on the LVPS and all that was correct.unless a mistake my. :bullshit:
  • Country: br
« Reply #103 on: February 15, 2016, 03:34:19 am »
I did see that some rectifier bridges,are very hot.I dont know the regimen of those rectfiers.
Offline rf+tech
« Reply #104 on: February 15, 2016, 03:41:58 am »
Yes, I remember the tripler may be defective. The manual says the CRT cathode voltage should be -2295 Volts. Can you measure this safely?
If this voltage is correct, then the tripler is defective. But do not worry about it, I have an idea.
What CR number on the power supply schematic?
  • Country: br
« Reply #105 on: February 15, 2016, 04:03:57 am »
I measured The cathode V and,it was around -1850V.What is CR? Is the part number of CRT?
  • Country: br
« Reply #106 on: February 15, 2016, 04:19:45 am »
I do have a RCA 56kV probe. I will measure  the cathode again.
« Last Edit: February 15, 2016, 04:34:49 am by SAUL BRITTO »
  • Country: br
« Reply #107 on: February 15, 2016, 04:29:39 am »
rf+tech> It shows me exactly -1832V.
  • Country: br
« Reply #108 on: February 15, 2016, 04:57:44 am »
Ah, ok! CR= rectifiers. I think most.
Offline rf+tech
« Reply #109 on: February 15, 2016, 04:59:14 am »
CR1, CR2, CR3, CR4, CR5, CR6, CR7 are the bridge rectifiers on the power supply board schematic.
What is the CR number of the hot rectifier bridges on the power supply board?
This cathode voltage is too low. Measure the +120 Volts on the HV power supply. And check that the "Deep Erase" switch on the rear panel is set to "Normal."
About the -1832 Volts, here is what I think. Look at the HV power supply schematic. See where the tap on the HV transformer secondary (pin 7) is less turns than the end of the secondary that becomes -2295 Volts with CR3?
If the +120 Volts is good, and the tripler is defective, the tripler may be loading the HV transformer and pulling the CRT cathode voltage down. Disconnect the lead to the tripler, from this pin 7 of the HV transformer. Measure the CRT cathode voltage again. Is it still -1832 Volts or is it now close to -2295 Volts?
  • Country: br
« Reply #110 on: February 15, 2016, 05:42:28 am »
rf+tech is normal the UNG 15V with 24V? I know that is unregulated but...As I don't know very well this circuit...
  • Country: br
« Reply #111 on: February 15, 2016, 05:49:35 am »
rf+tech the 120V measuring on the botton of board on the connector that goes to HV is with 123V  the pin oh tripler I will do it later,because I'm almost dead tired! :-DD
  • Country: br
« Reply #112 on: February 15, 2016, 05:51:12 am »
Here in Brazil: 2:50 until later. Thank you for all . have a good day.
« Last Edit: February 15, 2016, 01:39:32 pm by SAUL BRITTO »
  • Country: br
« Reply #113 on: February 15, 2016, 04:42:10 pm »
rf+tech disconnecting the cable post-tripler and the GND will not be the same than disconnecting that one of pin 7?
Offline rf+tech
« Reply #114 on: February 15, 2016, 05:58:12 pm »
Yes, it is the same. It is not safe. There may be an arc that damages other components or yourself.
Only if the wire to the tripler from the HV transformer is one piece without a solder or twist lock connection, should floating the tripler be attempted and only with great caution and very good insulation. In this case, be very very careful.
Here is a saying to think about:
There are old electricians, and there are bold electricians.
But there are no old, bold electricians.
  • Country: br
« Reply #115 on: February 15, 2016, 09:45:55 pm »
 :-DD Oh yes it's true!! But I wil do it as you said in another text,the correct way.I Preffer be an old not a bold.
« Last Edit: February 15, 2016, 10:00:37 pm by SAUL BRITTO »
  • Country: br
« Reply #116 on: February 15, 2016, 10:43:10 pm »
rf+tech already managed to repair a tripler? It's possible to use another trilper as one tektronix in lacking of an original?
  • Country: br
« Reply #117 on: February 15, 2016, 10:45:31 pm »
I know a guy here in my country that made it times ago.  He  fixed the tripler,was a philips one.
« Last Edit: February 15, 2016, 10:51:19 pm by SAUL BRITTO »
  • Country: br
Offline rf+tech
« Reply #119 on: February 15, 2016, 11:21:34 pm »
Saul,
Yes, it is possible to build one yourself. Here is an application note by Vishay on just this subject:
http://www.vishay.com/docs/88842/anusingr.pdf
Yes, it should be possible use almost any other scope tripler and if you are careful, you may be able to excavate and repair the HP tripler.
There are many possibilities. First, tell me about the test results.
RF+ Tech
  • Country: br
« Reply #120 on: February 15, 2016, 11:30:59 pm »
rf+tech I was seeing a french video abouT this scope (hp) and it was fenomenal for your time.
Ok, soon I tell you the result.
« Last Edit: February 15, 2016, 11:33:30 pm by SAUL BRITTO »
  • Country: br
« Reply #121 on: February 15, 2016, 11:53:48 pm »
rf+tech beeing it a tripler, that multiplies x3 then, the flyback generates around 1600V?
Sorry for this idiot question but,is just to be sure about a small question: "sparks," in case of disconnection of pin 7.
« Last Edit: February 15, 2016, 11:55:41 pm by SAUL BRITTO »
Offline rf+tech
« Reply #122 on: February 16, 2016, 12:21:07 am »
Yes, when the HV circuit is able to supply 2295 Volts to the CRT cathode.
Can the wire to the tripler be disconnected at the HV power supply board? Or must the wire be disconnected from the tripler?
  • Country: br
« Reply #123 on: February 16, 2016, 01:04:57 am »
it must to be disconnected from the board,has a pin on the botton of the board connected to the flyback with a kind of thermo-retractil insulator.
« Last Edit: February 16, 2016, 01:09:18 am by SAUL BRITTO »
  • Country: br
« Reply #124 on: February 16, 2016, 01:15:34 am »
I just can disconnect the tripler on that connection of anode of tube and the ground pin as I said before.
  • Country: br
Re: HP1741A in a big shape but... I bet in a focus problem, and you?
« Reply #125 on: February 16, 2016, 01:43:10 am »
courageously I disconnected the cable from the anode and the earth and it continues measuring -1832V.
« Last Edit: February 17, 2016, 01:33:11 am by SAUL BRITTO »
Offline rf+tech
« Reply #126 on: February 16, 2016, 02:06:34 am »
This is interesting. My best guess is someone tried to "fix" the original problem shown in your video by adjusting the HV.
Always, always mark any control before turning, so it can be put back in the original place.
Now it is right to try HV ADJ R38, to see if the HV can be increased to -2295 Volts. If it only goes less, tell me what the measurement is.
Of course, reconnect the tripler first.
  • Country: br
« Reply #127 on: February 16, 2016, 02:16:48 am »
Online rf+tech really,all the  electrolityc capacitors was replaced in all the PS. Power oscillator is with strange behavior on DMM.
I will back with more.
« Last Edit: February 16, 2016, 02:20:30 am by SAUL BRITTO »
  • Country: br
« Reply #128 on: February 16, 2016, 02:37:30 am »
On the HV ADJ I can -1922 maximum on the cathode test point.
« Last Edit: February 16, 2016, 02:45:45 am by SAUL BRITTO »
Offline rf+tech
« Reply #129 on: February 16, 2016, 02:47:31 am »
Okay, I would put this back to -1832 Volts.
How dirty is the pc board around CR3, C6, R11, C7, R12 and all components that connect to the cathode circuit? Could there be leakage around these parts?
Look closely at R13, 30 MOhm resistor. If there is electrical leakage across R13, the HV will be low and not adjust. Better to remove R13 and measure it.
  • Country: br
« Reply #130 on: February 16, 2016, 02:51:16 am »
I found a resistor of 560 Ohms measuring 598 Ohms,just need to know what is it's connection. I don't know ifit is R18 or R19 on the pin 6 of IC U1.Or that one connected to CR3 it seems R11.
« Last Edit: February 16, 2016, 02:55:18 am by SAUL BRITTO »
  • Country: br
« Reply #131 on: February 16, 2016, 03:06:44 am »
 The board is very clean in all areas.
  • Country: br
« Reply #132 on: February 16, 2016, 03:15:16 am »
R13 is  29.99M
« Last Edit: February 16, 2016, 03:46:49 am by SAUL BRITTO »
Offline rf+tech
« Reply #133 on: February 16, 2016, 03:36:46 am »
Good information on R13 and very clean pc board.
The voltage on U1 pin 6 may tell something.
  • Country: br
« Reply #134 on: February 16, 2016, 03:56:56 am »
R21 1M on pin 6 is  more than 20% altered,it measures 1.260M
« Last Edit: February 16, 2016, 05:26:43 am by SAUL BRITTO »
  • Country: br
« Reply #135 on: February 16, 2016, 04:24:23 am »
 rf+tech at the pin  6 of U1 i have -2.6V.
« Last Edit: February 16, 2016, 05:43:38 am by SAUL BRITTO »
Offline rf+tech
« Reply #136 on: February 16, 2016, 02:46:59 pm »
This is good and shows that the feedback loop is regulating properly.
But, this tells me the RCA HV probe calibration may be not be good.
Since U1 pin 2 voltage is zero (returned to ground via R24), the voltage at U1 pin 3 must also be zero.
Therefore, the two currents through the resistor dividers must also sum to zero.
-1832 Volts / 29.3 MOhms = -62.5 uA, derived from the -HV
If the curent through R13 is correct, then the current through R36, R38 and R37 must also be the same magnitude, but opposite polarity. So we solve to find the total resistance required for +62.5 uA from +15 Volts:
15 Volts / 62.5 uA = 240 kOhms
The total resistance of R36, R38 and R37:
182 kOhms + 10 kOhms + 9.09 kOhms = 201 kOhms, with R38 at maximum
201 kOhms - 5 kOhms = 196 kOhms R38 mid control
Now we calculate the current from +15 Volts:
+15 Volts / 196 kOhms = +76.5 uA
This must be balanced by the same current of the opposite polarity from -HV:
-76.5 uA * 29.3 MOhms = -2242 Volts
This shows that the -HV must really be -2242 Volts and shows that the RCA HV probe cannot be trusted.
  • Country: br
« Reply #137 on: February 16, 2016, 04:37:46 pm »
rf+tech Ok. That probe is really bad then.Despite I've measured some other voltages with it and was ok,voltages of 220V or below. In this case what do you think that can be the defect?
« Last Edit: February 16, 2016, 04:40:03 pm by SAUL BRITTO »
  • Country: br
« Reply #138 on: February 16, 2016, 04:49:16 pm »
I was thinking about an overload caused by any things as tube,can it be possible?
  • Country: br
« Reply #139 on: February 16, 2016, 04:51:52 pm »
But with this news I am a bit confuse.If the cathode V is right maybe the anode V is  right too.
  • Country: br
« Reply #140 on: February 16, 2016, 04:59:19 pm »
rf+tech I do have here some VTVM in good condictions  it can measure 1500V and it's input impedance is 11M in DC skill.  Can I double or duplicate the impedance with another resistor of 11M in series? I think there is a way to resolve this.Of coarse: My hand the probe will not see  :-DD
« Last Edit: February 16, 2016, 05:09:43 pm by SAUL BRITTO »
  • Country: br
« Reply #141 on: February 16, 2016, 05:42:56 pm »
I did this question before I don't know if you remember: It's normal the 15V Unreg  be 24V? It's 24V on the PS board and HV board.
Offline rf+tech
« Reply #142 on: February 16, 2016, 11:20:33 pm »
Saul,
I'll try to answer some things.
Depending on your mains voltage, if it is higher than normal, then yes, the unregulated 15 will be high. This really does not affect the HV output, U1 compensates for any deviation on the HV output, or the input, by adjusting the DC voltage to the base of the power oscillator transistor. Since U1 is the regulator, the 15 volts to the power oscillator does not need to be regulated.
More math: The HV probe error is 2242 - 1832 = 410 Volts low.
410 / 2242 = 18.3%
2800 * 18.3% = 512 Volts
2800 + 512 = 3312 Volts from the tripler
3312 / 2 = 1656 Volts into the tripler
1656 * 3 = 4968 Volts out of the tripler.
4956 Volts is "close enough" to 5 kV, yes?
Looks like the tripler is only a doubler, it is defective.
Defective may be a bit strong. 18.3% error is excessive. The probe reading ok at 220 V may indicate the high value resistor inside has some voltage-dependent non-linearity. If this is true then even the 18.3% figure may in fact need to be increased with higher voltages, such as at 2.8 kV. Post a photo of the probe and how it is connected to your meter. I am guessing this probe is a color TV HV probe and it is being used with a DVM? If this is true, the probe is intended to work into an 11 MOhm VTVM, Using this probe with a 10 MOhm DVM introduces an error immediately. But I am guessing because I do not know any of the important details to make a good conclusion.
I have experience working with feedback resistances from 2 GOhms (2*10^9) to 2 TOhms (2*10^12). This is an art all of its own, just as with RF and microwaves. There may be some hope for your probe, but I need to see it first and know the details about it.
RF+ Tech
  • Country: br
« Reply #143 on: February 17, 2016, 01:28:54 am »
Hi rf+tech Good night. Thank you. Yes looks like the probe was not so good seems it was a bad contact and really I WAS USING A DMM! wow! But I has some VTVM like the heathkit IM-10 RCA-Voltohmyst etc. The probe is an old RCA WG-297
« Last Edit: February 17, 2016, 01:44:59 am by SAUL BRITTO »
  • Country: br
« Reply #144 on: February 17, 2016, 02:11:46 am »
rf+tech can I measure with a VTVM directly with the HV probe,without that 11M one?
  • Country: br
« Reply #145 on: February 17, 2016, 02:40:31 am »
rf+tech Using a VTVM on the post tripler I can measure 5kV now
  • Country: br
« Reply #146 on: February 17, 2016, 02:46:38 am »
Including I can hear a noise of HV coming from tripler when I touch with the HV probe inside the connector even when the scope is off.
  • Country: br
« Reply #147 on: February 17, 2016, 03:12:53 am »
Where is marked I can measure exactly 5kV,can this be a gridd not polarized or tha HV post-tripler cable with some kind of problem? Did you see any thing like this? Times ago I remember not exactly if in same way a problem on HV connection of anode of a tv tube, the contact was oxidized.
Offline rf+tech
« Reply #148 on: February 17, 2016, 03:30:34 am »
Saul,
rf+tech
My memory is fuzzy, the Heathkit and RCA VTVMs were 10 MOhm at the front panel connector and have an additional 1 MOhm resistor in the probe body. The over-all input resistance is 11 Megohms.
The RCA WV-98A manual states:
DC voltages as high as 50,000 Volts can be measured directly by the WV-98A when the WG-289 High-Voltage Probe is used. The probe uses the WG-206 Multiplier Resistor, having a value of 1090 megohms, to present an overall voltmeter input resistance of 1100 megohms.
Well, this is good news, indeed.  :-+ And you have learned that improper results and incorrect conclusions occur when using HV probes with the wrong meters.
You may ask yourself "how did I know your mistake?"  ::)  :-DD
There may not be a bleeder resistor present in the tripler to discharge the output filter capacitor. When the probe tip approaches the connector, the air will ionize and a small sizzle noise will be heard. Leaving the probe in contact should show the voltage falling, as the probe discharges the output capacitor. Perfectly normal.
I think what you are saying is that the grid may be in cutoff, so there is no beam from the cathode. This is what I believe to be the *next* problem. I do not think oxidation of the HV anode is the problem. Really, how can a little oxidation stop 5000 Volts?
Next, the cathode and grid voltages need to be measured, to see if the CRT is cutoff. This should be done with the HV probe and VTVM because both are at high voltages.
RF+ Tech
  • Country: br
« Reply #149 on: February 17, 2016, 03:36:53 am »
rf+tech I can read -2450V on the TP5 of gridd,is it ok? I do not see at the schematic the indication of voltage for that.
« Last Edit: February 17, 2016, 04:00:45 am by SAUL BRITTO »
  • Country: br
« Reply #150 on: February 17, 2016, 04:04:57 am »
You may ask yourself "how did I know your mistake?" >> The masters are like poets, they know and can all the things.
rf+tech you are a big professional and I hope be a good disciple.
« Last Edit: February 17, 2016, 05:06:01 pm by SAUL BRITTO »
Offline rf+tech
« Reply #151 on: February 17, 2016, 05:10:17 am »
Saul,
By itself, this measurement has little meaning. Both measurements must be considered together. What is the cathode voltage?
Thank you for the good compliments. I am glad to count you as a follower.  ;)
RF+ Tech
  • Country: br
« Reply #152 on: February 17, 2016, 05:19:54 am »
rf+tech I can read 2.4kV on the cathode. Seems I have a difference of -50 Volts gridd to cathode.
« Last Edit: February 17, 2016, 05:38:49 am by SAUL BRITTO »
Offline rf+tech
« Reply #153 on: February 17, 2016, 09:06:07 pm »
Saul,
The -50 Volts grid to the cathode appears to be normal cutoff. The next circuit to check is the gate amplifier, to see if the gate pulse is missing.
First, it is very important from now on to set front and rear panel controls exactly as specified in Section V, paragraph 5-12. Do not change any of these control settings, except as the manual instructs in a specific procedure.
Many measurements are affected by the front and rear panel controls. Any measurements made with controls not set as instructed can be considered invalid, cause confusion and waste time. Read paragraphs 8-71 through 8-76 carefully.
The Gate Amplifier circuit should be checked as described in paragraph 5-19. This requires another working oscilloscope and 10:1 probe. The 10:1 probe must be properly compensated using the built-in 1 kHz signal source.
Follow steps in 5-20a and 5-20b. At step 5-20c, do not make any adjustment at this time, only measure the peak amplitude of the gate signal. Tell me the measurement, or if the gate signal is not present. Continue with step 5-20f and 5-20g.
If the gate signal is missing, paragraph 8-86 offers some tips about failure.
RF+ Tech
  • Country: br
« Reply #154 on: February 17, 2016, 09:26:10 pm »
rf+tech I was right now looking the gate circuit it has a signal that goes to HV board and seems that varies with the sweep frequency,when is low it is low too. Ok I know what is the procedure to accomplish the measures.
  • Country: br
« Reply #155 on: February 17, 2016, 10:05:53 pm »
rf+tech CONV disengaged means convetional mode off?
  • Country: br
« Reply #156 on: February 17, 2016, 10:14:01 pm »
rf+tech it's ready!
Offline rf+tech
« Reply #157 on: February 17, 2016, 10:19:07 pm »
Saul,
You mean the scope is now working in conventional mode, with a trace?
I was about to post more troubleshooting steps...
RF+ Tech
  • Country: br
« Reply #158 on: February 17, 2016, 10:45:18 pm »
No trace.I said ready for troubleshootings. I just need you explain a small thing to me: where or what is VOLTS/DIV vernier  (Channel A and B) CAL detent (Fully Cw) I am with a small doubt about this. I just see one knob as vernier and don't understand  where it is with Channel A and B. Vernier appears twice have to be adjusted to CALL, call detent (Fully cw) and CAL so I did CAL
« Last Edit: February 17, 2016, 11:00:21 pm by SAUL BRITTO »
  • Country: br
« Reply #159 on: February 17, 2016, 11:08:49 pm »
I already has another scope on the hand.
Offline rf+tech
« Reply #160 on: February 17, 2016, 11:23:44 pm »
No trace.
Aw, I was ready to post my beer icon!
Photo showing verniers (if I did this right):
Okay, the manual describes:
A protection circuit consisting of A15Q4, A15U2, A15Q5 is incorporated in the cathode circuit to prevent accidental burning of the CRT during deep erase operations, while turning the instrument on and off, and during quick power dropouts.
Check that Q4 does not stay on continuously, by measuring pin 1 of U2. If the voltage is about -13 Volts, then the LED in U2 is turned on, the output of U2 will turn on Q5 and the CRT will be cutoff.
Be very careful identifiying pin 1 of U2, the other side of U2 is -HV cathode potential.
Capacitor C18 should be checked. What about the filament in the CRT? Does the neck feel warm?
RF+ Tech
  • Country: br
« Reply #161 on: February 17, 2016, 11:38:37 pm »
Ok. I have to put the knobs CAL on,where that red led lights?
  • Country: br
« Reply #162 on: February 17, 2016, 11:42:11 pm »
U2 is the 6 pins IC that seems a optocoupler or is that in gate board? That in gate board appear as U1,maybe it's second section be U2...
« Last Edit: February 17, 2016, 11:55:56 pm by SAUL BRITTO »
  • Country: br
« Reply #163 on: February 17, 2016, 11:46:09 pm »
rf+tech if CAL detent has to be Fully CW then it has to be off,rigth?
Sorry for those small questions it's because I want to do it right.
« Last Edit: February 18, 2016, 12:17:51 am by SAUL BRITTO »
Offline rf+tech
« Reply #164 on: February 18, 2016, 12:00:20 am »
Yes, the red lights should be off. When the CAL knobs are turned full CW, the Volts/Div is CALibrated.
When the CAL knob is turned CCW, the Volts/Div is uncalibrated and the red light is on to remind.
U2 is the 6 pins IC that seems a optocoupler?
Yes, this is correct.
Ah, gafanhoto, você está se tornando sábio como eu.  ;)
(Do you know the movie 'Kung Fu"?)
  • Country: br
« Reply #165 on: February 18, 2016, 12:03:45 am »
If U2 is that 6 pins,pin 1 is -10.2V
  • Country: br
« Reply #166 on: February 18, 2016, 12:06:55 am »
Hey man this about gafanhoto I did not see but,will tell you later about the cat and the jaguar.
« Last Edit: February 18, 2016, 12:14:01 am by SAUL BRITTO »
  • Country: br
« Reply #167 on: February 18, 2016, 12:25:25 am »
I already see the filament and it lights and feel a bit hot that shield over it.
Offline rf+tech
« Reply #168 on: February 18, 2016, 01:01:45 am »
This says that the LED inside U2 may be turned on when it should be off. To verify another way, measure the voltage drop across R45, 270 Ohms. The current through R45 and the LED will be Volts divided by 270, maybe 10 mA. But there should not be any voltage drop, no current and the LED inside U2 should be off.
The voltage on C18 should be very close to -15 Volts. This will keep Q4 turned off. If C18 is leaky/shorted, Q4 may be turned on, then U2 will disable the trace. If Q4 is shorted, U2 will disable the trace.
  • Country: br
« Reply #169 on: February 18, 2016, 01:16:10 am »
I am trying to find C18 but can't. I see it on the schematic but on the boar looks like it isn't there,do you know it's location?
« Last Edit: February 18, 2016, 01:20:30 am by SAUL BRITTO »
  • Country: br
« Reply #170 on: February 18, 2016, 01:24:02 am »
I found that 270 Ohms resistor altered to 320 ohms.
Offline rf+tech
« Reply #171 on: February 18, 2016, 01:26:21 am »
Figure 8-17, page 8-20 chart says C18 is at grid "E4"
That is a problem, but not *the* problem.
Offline SAUL BRITTOTopic starter
  • Country: br
  • Don't Make a Mistake
« Reply #172 on: February 18, 2016, 01:28:57 am »
Seems the resistor 2700 ohms on the normal position do not works, right? It is altered to 3.200 ohms too.
Thank You, for all earth.
  • Country: br
Re: HP1741A in a big shappe but... I bet in a focus problem, and you?
« Reply #173 on: February 18, 2016, 01:35:25 am »
At my mind C18 was a electrolityc CAP but, it is more smart than a gafanhoto,was under any cables and is another kind of cap seems a mica cap.
  • Country: br
« Reply #174 on: February 18, 2016, 01:42:06 am »
Across R45 I read around 4.2 mV
  • Country: br
« Reply #175 on: February 18, 2016, 01:53:30 am »
Wow measuring the R45 I see 6.2 mV one lead to the other.Is this you want?
  • Country: br
« Reply #176 on: February 18, 2016, 02:01:17 am »
6.2mV/320 Ohms (270) will this led work with this current?
  • Country: br
« Reply #177 on: February 18, 2016, 02:05:59 am »
Is it 19uA or, am I confused?
Offline rf+tech
« Reply #178 on: February 18, 2016, 02:09:13 am »
Is this really 3200 Ohms (3.2 kOhms)? Maybe it is time to check your meter with some 1% precision resistors?
C18 ... seems a mica cap.
22 uF in mica will be about as large as the entire HV board. The parts list has the letters "TA" which indicates this is a Tantalum capacitor.
Okay, there is no current flowing through the LED in U2.
.0042 Volts / 270 Ohms = 15.5 uA. The LED should be off.
.0062 Volts / 320 Ohms = 19.3 uA.
The LED would need maybe 10 mA to turn on U2.
Move the Normal/Deep Erase to Deep Erase. This will turn on Q4 and the LED in U2. Now measure the voltage across R45.
With the scope turned off, some tests on the HV side of U2. Check zener diode VR3 (zener diodes like to fail shorted), diode-test Q5 for shorted and diode test U2 pins 5 to 6 for shorted.
Oh, yes, verify sweep at the deflection plates. Using another scope, check both horizontal deflection plates for the sweep ramp.
  • Country: br
« Reply #179 on: February 18, 2016, 02:18:52 am »
The voltage over C18 is -11.2 V
  • Country: br
« Reply #180 on: February 18, 2016, 02:20:47 am »
 :-DD no I said 320 Ohms (270).
  • Country: br
« Reply #181 on: February 18, 2016, 02:28:48 am »
With deep erase on I read 0.2mV.
Offline rf+tech
« Reply #182 on: February 18, 2016, 03:06:09 am »
This looks okay. With very low current, VR4 will be less than 11 Volts. 11.2 Volts minus 0.6 Volts for Q4 base-emitter = 10.6 Volts across VR4. I did not look closely at this.
Yes, but read again what you wrote earlier:
2700 Ohms implies R43
270 Ohms implies R45.
Who is  :-//
So please explain "3.200 Ohms."
3.2 Ohms is 2696.8 Ohms less than 2700 Ohms.
I have to guess is this really Ohms or kOhms? I think this is 3200 Ohms (3.2 kOhms).
If your meter measures 270 Ohms as 320 Ohms, 15.6% higher, and measures 2700 Ohms as 3200 Ohms, also 15.6% higher --- then your meter may not be accurate. The meter needs to be checked with known good 1% resistors. It would be very unusual for all resistors in this HP scope to be 15.6% high.  :rant:
Hmm. This is only 625 uA, I don't think U2 can turn on with this low current.
What part number appears on U2? The parts list shows 1990-0607.
What are the emitter, base and collector voltages on Q4, in normal and deep erase?
  • Country: br
« Reply #183 on: February 18, 2016, 03:19:07 am »
I measured with three meters and was all near basicly the same thing.The resitor of 2700 Omhs is 3.000 Right now
Maybe because I removed U2 and replaced it the pins was very dirty,I don't know...
 I did look if was a short in collector and emmiter of U2 but, no.
« Last Edit: February 18, 2016, 03:38:41 am by SAUL BRITTO »
  • Country: br
« Reply #184 on: February 18, 2016, 03:27:18 am »
The part number is 199006077812
« Last Edit: February 18, 2016, 03:37:09 am by SAUL BRITTO »
  • Country: br
« Reply #185 on: February 18, 2016, 03:34:45 am »
Normal
Base= -11.23V
Emmiter= -10.63V
Collector= -10.64V
Deep erase
Base= -0.6
Collector= -13.37
Emitter= -3.7
« Last Edit: February 18, 2016, 03:45:59 am by SAUL BRITTO »
  • Country: br
« Reply #186 on: February 18, 2016, 03:52:34 am »
Man! I think this resistor of 68K don't smell good...
  • Country: br
« Reply #187 on: February 18, 2016, 04:07:24 am »
The 68k is 76.6k Omhs
Offline rf+tech
« Reply #188 on: February 18, 2016, 04:49:19 am »
The voltages on Q4 show that my interpretation of Q4 on/off is reversed.
These old HP schematics where transistors are drawn both "right side up" and "up side down" always make my head hurt.
The most common convention is to place the transistor terminal, emitter of PNP and collector of NPN at the top where positive voltage is also at the top. Negative voltages and transistor terminals connected to negative voltages are drawn lower. Here is an example where PNP transistors are drawn two different ways in the same schematic. Crazy. This just makes it harder to understand the circuit.  |O
After drawing this circuit on paper the "proper" way, it is obvious at a glance how Q4 works. When the scope is off, the -15 Volt supply is at zero. C18 discharges through R44. When switch S1 is in "normal" and the scope is first turned on, C18 looks like a short and holds Q4 off. R44 charges C18 with current toward -15 Volts. When the voltage on C18 exceeds VR4 plus the base-emitter drop of Q4, about -11.2 Volts, Q4 turns on.
When "deep erase" connects R45 to ground through switch S1, C18 is discharged and Q4 is turned off.
In "normal" with Q4 turned on, current through R45 turns on the LED in U2 and the output transistor of U2 also turns on. This action turns on Q5 and shorts VR3.
VR3 causes the CRT cathode to be 100 Volts more positive than the grid, the CRT is more deeply cutoff. This protects the CRT from burning when the scope is turned on and turned off, or when electric power is lost momentarily and comes back in very quickly.
After a short delay by C18 and R44, the protection is released.
Based on all the voltage measurements, I think the problem is the LED inside U2 has failed open circuit. The output transistor inside U2 cannot turn on and the extra 100 Volts keeps the CRT deeply cutoff.
The single most valuable measurement is the voltage drop across R45 (270 Ohms). Using a 4N25 datasheet as a reference, 10 mA of LED forward current is used to describe the transfer characteristics. Data found for U2 part number 1990-0607 is too minimal to be 100% conclusive, but 625 uA seems not sufficient to turn on U2.
4N25 is not suitable for replacing U2. A good parametric search is necessary to ensure a replacement capable of withstanding 120 volts across the emitter-collector of the output and 4 kV across the input side to the output side.
You indicated that U2 is in a socket, do you know how to test U2 out of the circuit?
  • Country: br
« Reply #189 on: February 18, 2016, 04:51:07 am »
rf+tech I measured some resistors of 1% 1M,100K,10K and the fluke is ok.
  • Country: br
« Reply #190 on: February 18, 2016, 05:01:40 am »
This troubleshooting is better than fix the scope  :-DD. Yes I know how to test this LED.
Despite I have another CI of a HP1740A just don't know if it's the same.
« Last Edit: February 18, 2016, 05:12:46 am by SAUL BRITTO »
  • Country: br
« Reply #191 on: February 18, 2016, 05:10:59 am »
OOps the 1740A do not have this CI. Fail.
  • Country: br
« Reply #192 on: February 18, 2016, 05:15:15 am »
I have a PS of a tektronix TDS320 tha has two CI Like this one,just don't know the electric feature.
  • Country: br
« Reply #193 on: February 18, 2016, 06:11:30 am »
rf+tech how many volts I can use in this IC?i see -15V on the cathode,can I use +15 on the anode, on inverted situation? i jusT did put 0.6 V.Looks like it is with inttermitent work but,I am not sure.What is the right way to test that IC?
« Last Edit: February 18, 2016, 06:24:47 am by SAUL BRITTO »
Offline rf+tech
  • Real radios are olive drab
« Reply #194 on: February 18, 2016, 06:36:33 am »
Saul,
Use the IC part number and search for a datasheet and post the link here.
Yes, inverted is okay, but the LED does not see 15 Volts in the circuit. The voltage is 15 - 11 = 4 Volts, with 270 Ohms in series. Measure the voltage drop across 270 Ohms, exactly as in the circuit.
A good test to show that Q4 is working correctly would be to carefully push the leads of red 20 mA LED into IC socket pins 1 and 2, with the proper polarity. The LED should turn on about 1.5 seconds after the scope is turned on. The actual current through R45 (270 Ohms) and the LED will be a function of Q4 gain and Q4 base current supplied by R44 (68 kOhms). Measure the voltage drop across R45 with the red LED for reference.
Continue tomorrow,
RF+ Tech
RF+ Tech
  • Country: br
« Reply #195 on: February 18, 2016, 06:40:56 am »
I used  breadboard
  • Country: br
« Reply #196 on: February 18, 2016, 07:24:07 am »
  • Country: br
« Reply #197 on: February 18, 2016, 07:54:48 am »
Well, I tested this IC of datasheet and now I have 2.45V across R45 and maybe a current of 9mA.
« Last Edit: February 18, 2016, 07:58:11 am by SAUL BRITTO »
Offline rf+tech
« Reply #198 on: February 18, 2016, 03:03:28 pm »
Very good Saul, this is the kind data needed.
The MOC8101 minimum breakdown voltage (BVceo) is only 30 Volts. U2 is expected to withstand 100 Volts, the MOC8101 output may have already failed.
Please put original U2 on a breadboard with 270 Ohms in series and supply 4 Volts.
What is the voltage across 270 Ohms?
Good engineering decisions depend on good data, not guesses.
RT-1133 AN/PRC-70  *  RT-794 AN/PRC-74  *  RT-841 AN/PRC-77  *  RT-524 AN/VRC-12  *  RT-834 AN/GRC-106  *  RT-F100
  • Country: br
« Reply #199 on: February 18, 2016, 06:15:59 pm »
Look: contrary to logic I put MOC81 on the oscilloscope and got 2.7V acros R45 and -14 on pin 1 of U1 but the picture remains the same (Just for quick test).
« Last Edit: February 18, 2016, 06:34:14 pm by SAUL BRITTO »
  • Country: br
« Reply #200 on: February 18, 2016, 07:14:19 pm »
Replacing that R45 270 Ohms that was 320 in fact, I can measure  across it 2.60V and 9.6mA almost that 10mA
You said before.
But no with the original U2,ok?
« Last Edit: February 19, 2016, 08:03:37 pm by SAUL BRITTO »
Offline rf+tech
« Reply #201 on: February 18, 2016, 10:29:05 pm »
Okay, that was a good test and proves the LED in U2 has failed.
Even I would try the MOC8101 if that was all I had, just to see if Q4 would turn on the LED. Knowing the breakdown voltage of MOC8101 is too low, I would expect the output side to fail. Sometimes it is good to make a sacrifice to the silicon gods so they may look favorably upon your efforts and grant you success. (Did I really say that?)  :-DD
Here are some preferred part numbers to replace U2:
H11D1 (H11D2, H11D3,vH11D4)
MOC8204
H11G1 (H11G2, H11G3, H11G4)
TLP372
PC725
These all have BVceo of 200 Volts or higher on the output side and around 5000 Vrms isolation input to output. The first two are bi-polar output and the last three are darlington output.
  • Country: br
« Reply #202 on: February 18, 2016, 10:51:54 pm »
rf+tech the original U1 has the same behavior on t beadboard and scope and the other too, despite different.
« Last Edit: February 18, 2016, 10:56:52 pm by SAUL BRITTO »
  • Country: br
« Reply #203 on: February 18, 2016, 11:00:14 pm »
Explain a thing to me please: This scope just works if U1 activates Q4,or can we make any adaptation for that?
  • Country: br
« Reply #204 on: February 18, 2016, 11:03:27 pm »
H11D11 I found in a site here in Brazil 2,50 $
  • Country: br
« Reply #205 on: February 18, 2016, 11:17:46 pm »
MOC8204
TLP372
PC725
rf+tech, any one of these will work?
  • Country: br
« Reply #206 on: February 18, 2016, 11:30:01 pm »
Found all that vey cheap.but,we can't do more troubleshootings with that no original?
Offline rf+tech
« Reply #207 on: February 18, 2016, 11:43:38 pm »
Saul,
I don't understand something and need some help. I will try to say what I think you are saying:
U2 (U1 is an op-amp) in the scope, only some mV measured across R45 and on the breadboard, similar mV across 270 Ohms.
MOC8101 in the scope, 2.6 Volts across R45 (9.6 mA) and on breadboard, similar Volts across 270 Ohms.
Or is it:
U2 and MOC8101 in the scope only measure some mV across R45.
U2 and MOC8101 on the  breadboard both measure about 2.6 Volts across 270 Ohms.
Yes.
Q4 activates U2. U2 activates Q5. Q5 deactivates VR3. This safety circuit is to prevent burning the CRT and permanently damaging it. You are so close to making this scope work, why take a foolish risk and defeat this feature?
I think it is time to take a break from this until the replacement parts arrive. Since the parts are very cheap, it would be good to buy 2 or 3 spares.
Maybe it is time to put some of what you have learned toward work on the 1740?
RF+ Tech
  • Country: br
« Reply #208 on: February 18, 2016, 11:51:01 pm »
Or is it:
Sorry. Some times I use the brasilian way of thinking and confuse you. Look: the Original U2 on the breadboard work inttermitent and the other IC, works well: on the bradboard or scope.
« Last Edit: February 18, 2016, 11:52:43 pm by SAUL BRITTO »
  • Country: br
« Reply #209 on: February 18, 2016, 11:54:08 pm »
MOC8204
TLP372
PC725 what do you prefer?
  • Country: br
« Reply #210 on: February 18, 2016, 11:57:27 pm »
I found here H11D1 (H11D2, H11D3,vH11D4)
MOC8204
Offline rf+tech
« Reply #211 on: February 19, 2016, 12:27:59 am »
Ah yes, of course.
I did take two years of Español in high school and found it interesting that English is backward in sentence structure from many other languages. To write in that manner is better for you.
H11D1, H11D2, H11D3 and H11D4 family same, difference little. All okay.
what do you prefer?
TLP372. Phototransistor base not external, other parts base is exposed on pin 6. Exposed base can pick up interference. Remove pin 6 okay.
1 - Optocoupler LED wears out with time.
2 - Bond wire die attach inside intermittent?
When original U2 working on breadboard, voltage across 270 Ohms = 0.6 Volts, or higher?
  • Country: br
« Reply #212 on: February 19, 2016, 12:30:05 am »
rf+tech put the beer on the freezer.I have good news :box:
  • Country: br
« Reply #213 on: February 19, 2016, 12:34:24 am »
I picked an other IC that was on the TDS320 PS board and... I have a signal on the screen when push beam finder
a not so good signal but...
« Last Edit: February 19, 2016, 12:48:48 am by SAUL BRITTO »
Offline rf+tech
« Reply #214 on: February 19, 2016, 12:36:22 am »
Photo, or it isn't true  :popcorn:
  • Country: br
« Reply #215 on: February 19, 2016, 12:40:45 am »
Yes it's true hahahah! Just a pale signal,when i push beam finder it shows a green screen,before was nothing "nada"!!! Then that other optocoupler of TDS320 wasn't good,don't ask me why.
  • Country: br
« Reply #216 on: February 19, 2016, 12:44:43 am »
Intensity now works in a pale way but yes.
  • Country: br
« Reply #217 on: February 19, 2016, 12:46:55 am »
And now I has a optocoupler tester  :scared: :clap:
« Last Edit: February 19, 2016, 12:50:50 am by SAUL BRITTO »
Offline rf+tech
« Reply #218 on: February 19, 2016, 12:50:52 am »
Really, a photo can provide clues. Show me what you see.
  • Country: br
« Reply #219 on: February 19, 2016, 12:53:06 am »
Did you know that when we brasilians speak the spanish don't understand nothing,but when they speaks we understand 90%  :-DD
  • Country: br
« Reply #220 on: February 19, 2016, 12:55:30 am »
I sse a pale green screen with no trace really just when push beam finder I see better.With no beam finder I see but very very pale.
« Last Edit: February 19, 2016, 02:40:51 am by SAUL BRITTO »
  • Country: br
« Reply #221 on: February 19, 2016, 01:03:51 am »
I did lights off and saw a trace out of screen there is sweep I can see very well. I wll try some images.
« Last Edit: February 19, 2016, 01:08:51 am by SAUL BRITTO »
  • Country: br
« Reply #222 on: February 19, 2016, 01:21:31 am »
Sorry it's is a so pale that my cam can't make a video or photo. Is a vertical trace going left to right.
  • Country: br
« Reply #223 on: February 19, 2016, 01:58:37 am »
rf+tech I got a way: https://youtu.be/WfvYR9KzVis
  • Country: br
« Reply #224 on: February 19, 2016, 02:08:28 am »
rf+tech I found a way to show you better,there's really a trace when a push beam finder and adjust position to the maximum.Wait I will make another video.
« Last Edit: February 19, 2016, 02:43:07 am by SAUL BRITTO »
  • Country: br
« Reply #225 on: February 19, 2016, 02:31:35 am »
rf+tech look at this: https://youtu.be/sE2myNBwYzw
Offline rf+tech
« Reply #226 on: February 19, 2016, 02:52:06 am »
Yes, the beam is off the top. The pale glow is electron scatter from internal reflections.
So we must go back some pages in this thread to check again vertical amplifier balance.
Set Chan A on (blue button A under Display), 20 Volts/Div, input Gnd.
Set Chan B off (blue button B under Display)
Set Chan A POSN pointing up.
Measure voltage across input or output of delay line (not delay line to ground.)
Set Chan A POSN full left (trace toward bottom)
Measure voltage across input or output of delay line.
Set Chan A POSN full right (trace toward top)
What are the three measurements?
  • Country: br
« Reply #227 on: February 19, 2016, 02:58:04 am »
I did remember trig view end did push the key and appear a nice trace.it's all working,focus,intensity etc.
It still has a vertical problem.
  • Country: br
« Reply #228 on: February 19, 2016, 03:20:00 am »
0k. The V is 0.335V
« Last Edit: February 19, 2016, 03:24:20 am by SAUL BRITTO »
  • Country: br
« Reply #229 on: February 19, 2016, 03:27:00 am »
V= 0.532V
  • Country: br
« Reply #230 on: February 19, 2016, 03:34:49 am »
Full right= 0.350.If iverted the probes all the voltages are inverted = minus
Offline rf+tech
« Reply #231 on: February 19, 2016, 03:52:48 am »
Tell me if I understand:
Up = 0.335 V
Left = 0.532 V
Right = 0.350 V
I *think" it should be:
Up = 0V
Left = -0.35V
Right = +0.35 V
Voltage should become minus left, plus right, equal amount. This is hard to describe:
Left -0.35 ... -0.3 ... -0.25 ... -0.2 ... -0.15 ... -1.0 ... -0.05 ... 0 ... +0.05 ... +0.1 ... +0.15 ... +0.2 ... +0.25 ... +0.3 ... +0.35 Right
Probes must be same at each measurement.
  • Country: br
« Reply #232 on: February 19, 2016, 03:55:26 am »
Ok I will do it again;Wait.
  • Country: br
« Reply #233 on: February 19, 2016, 04:02:24 am »
UP= .-342
Full left= .-335
Full Right= .-334
  • Country: br
« Reply #234 on: February 19, 2016, 04:07:39 am »
Please forget it trig view was on! I will be back with  news wait.
« Last Edit: February 19, 2016, 04:17:13 am by SAUL BRITTO »
  • Country: br
« Reply #235 on: February 19, 2016, 04:14:24 am »
UP= 0.-288
Full left= 0.-531
Full right= 0.350
And I can zero it a bit to right,around 15 or 20%.
Offline rf+tech
« Reply #236 on: February 19, 2016, 04:26:47 am »
Set Chan A POSN UP
Adjust R58 for 0.0 V, measured same way.
Full left -V should now be very close to full right +V
Offline rf+tech
« Reply #237 on: February 19, 2016, 04:32:55 am »
Set Chan A off, Chan B on
Set 20 Volts/Div, input Gnd.
Measure at delay line, same as for Chan A
Set Chan B POSN UP
Adjust R50 for 0.0 V, measured same way.
Chan B POSN full left -V should now be very close to full right +V
  • Country: br
« Reply #238 on: February 19, 2016, 05:14:43 am »
Ok I began by channel two.I still don't found the adust of channel 1-A sorry my eyes are day after day worse!
Channel 2 I can adjust to 0.000 well on the mid course of position knob.,but on the end of the trimpot of adjust.
« Last Edit: February 19, 2016, 05:21:46 am by SAUL BRITTO »
  • Country: br
« Reply #239 on: February 19, 2016, 05:24:21 am »
Ok column D  line 3,correct?
  • Country: br
« Reply #240 on: February 19, 2016, 05:31:40 am »
rf+tech I can't zero the channel A like the B.On the trimpot of adjuste the  more I can is -120mV, it just be zero with
the "position" a bit to right,maybe 3% but...
« Last Edit: February 19, 2016, 05:39:15 am by SAUL BRITTO »
Offline rf+tech
« Reply #241 on: February 19, 2016, 05:34:54 am »
Our eyes are fine, it's the manual poor quality.
R58 really is R38??
Red boxes I think are correct adjustments.
Check resistors, may be +15.6% like on HV board?
  • Country: br
« Reply #242 on: February 19, 2016, 05:42:07 am »
If 58 is 38 then, I adjusted on the wrong trimpot?
Offline rf+tech
« Reply #243 on: February 19, 2016, 06:02:09 am »
Really cannot read pot numbers in drawing, all look the same.
Check resistance markings. 2000 = Gain, 200 = POSN.
  • Country: br
« Reply #244 on: February 19, 2016, 06:03:46 am »
Ok my friend I think you are tired like me! When we are rested,will think better and do better actions.  :-DD
Thank you for all ! Hasta la vista. Até a próxima batalha!
as says the Terminator: I will be back! :-+
« Last Edit: February 19, 2016, 06:06:05 am by SAUL BRITTO »
  • Country: br
« Reply #245 on: February 19, 2016, 06:11:40 am »
Just a small question: Why do not the trace  appear,is for this vertical problem? When I push trig view and beam finder I see a good trace.
« Last Edit: February 19, 2016, 06:13:30 am by SAUL BRITTO »
Offline rf+tech
« Reply #246 on: February 19, 2016, 02:52:46 pm »
Yes, there is a problem with the vertical amplifier. First image below tells how beam finder works.
For now, set Chan B on, Chan A off
Set POSN UP
Second image shows where to measure across path through vertical amplifier, from delay line to deflection plates. All points should be close to zero Volts. Somewhere there will be a big difference. This pushes beam off top of screen. Beam Find reduces the difference (gain).
  • Country: br
« Reply #247 on: February 19, 2016, 04:03:35 pm »
rf+tech these measures has to be accomplished ground refferenced?
Offline rf+tech
« Reply #248 on: February 19, 2016, 04:10:15 pm »
Not ground reference, only side to side same as last night.
  • Country: br
« Reply #249 on: February 19, 2016, 05:38:58 pm »
Ok.
  • Country: br
« Reply #250 on: February 19, 2016, 08:38:15 pm »
Hi rf+tech good afternoon here in Brazil and all of good in USA for you.
Well guy looks like I have an open transistor I don't know exactly what,it's Q17 or Q18 on the V amp.I will look well because I need to be sure.
« Last Edit: February 19, 2016, 08:41:41 pm by SAUL BRITTO »
  • Country: br
« Reply #251 on: February 19, 2016, 08:48:53 pm »
Is a PNP 824-3-354 it's HP ID
Offline rf+tech
« Reply #252 on: February 19, 2016, 09:43:13 pm »
Q17 and Q18 parts list show 1853-0354.
Replacements, best first:
MRF545 Ft 1400 MHz
2N4405 Ft 600 MHz
2N5160 Ft 400 MHz
MRF525 Ft 400 MHz
2N2905 Ft 200 MHz
Important parameter is Ft for best frequency response. MRF545 may be too good. 2N2905 is very common part, good for quick test, but scope bandwidth will be reduced less than 100 MHz. Replace Q17 and Q18 to keep symmetry. Mismatched transistors here should be avoided.
Tell me what can be found in Brasil, I have MRF545 and 2N5160. (Really, I have too many RF transistors.)
  • Country: br
« Reply #253 on: February 19, 2016, 09:46:17 pm »
Any friend knows where to find a page to part numbers of HP?
  • Country: br
« Reply #254 on: February 19, 2016, 09:49:48 pm »
rf+tech can you say me the link to this place where you got these datasheets? (if not professionally interesting I'll understand)
Another question: it will better replace the two Tr?
« Last Edit: February 19, 2016, 10:00:51 pm by SAUL BRITTO »
  • Country: br
« Reply #255 on: February 19, 2016, 09:51:34 pm »
I do have another HP1740A that seems will be a donator for now... :-//
  • Country: br
« Reply #256 on: February 19, 2016, 09:53:11 pm »
MRF545 Ft 1400 MHz
2N4405 Ft 600 MHz
2N5160 Ft 400 MHz
MRF525 Ft 400 MHz
2N2905 Ft 200 MHz
Thank you for this rf+tech.
Offline rf+tech
« Reply #257 on: February 19, 2016, 10:10:57 pm »
See attached file: 300-hpxref.pdf
Offline rf+tech
« Reply #258 on: February 19, 2016, 10:19:24 pm »
MRF545: http://pdf.datasheetcatalog.com/datasheets2/89/89884_1.pdf
2N4405: http://pdf.datasheetcatalog.com/datasheet/central/2N4405.pdf
2N5160: http://www.datasheetarchive.com/dlmain/Datasheets-8/DSA-154832.pdf
MRF525: (same sheet as 2N5160 above)
2N2905: http://pdf.datasheetcatalog.com/datasheet/philips/2N2905.pdf
Definitivamente, é melhor para preservar a simetria em todos os sentidos entre ambos os lados do amplificador.
  • Country: br
« Reply #259 on: February 19, 2016, 10:28:24 pm »
Something say me that you do have RF Tr... :-DD
Thank you rf+tech for your kindness :clap:
  • Country: br
« Reply #260 on: February 19, 2016, 10:32:21 pm »
I'm enjoying your Portuguese :clap:
Estou gostando do seu português!
« Last Edit: February 19, 2016, 10:34:10 pm by SAUL BRITTO »
  • Country: br
« Reply #261 on: February 19, 2016, 10:37:08 pm »
I am trying to find here in Brasil,if I don't  find it I will tell you about.
  • Country: br
Offline rf+tech
« Reply #263 on: February 20, 2016, 12:50:11 am »
Muito interessante pesquisa, mas eu não sou capaz de encontrar os detalhes sobre as diferenças entre o transistor 3D NAMITEC eo transistor 3D da Intel?
Certamente transistor 3d de NAMITEC é muito superior à Intel, porque a Intel não é brasileiro, não é?  ;)
  • Country: br
« Reply #264 on: February 20, 2016, 02:46:18 am »
rf+tech Eu nem sabia da existência desse tipo de Tr,nem intel nem namitec,Só achei este artígo interessante.O importante é que eles façam  :popcorn:
  • Country: br
« Reply #265 on: February 20, 2016, 02:50:41 am »
Aí no texto diz que o Brasil está em nível de "igualdade" com outros países.Se está em igualdade quer dizer somos apenas mais um no páreo.
Offline rf+tech
« Reply #266 on: February 20, 2016, 03:26:21 am »
Sim, isso eu entendido a partir do artigo. Buscando mais informações sobre NAMITEC e transistor 3D, eu encontrei este relatório 2013-2014. Muito bom ver Brasil contribuindo avanços em muitas áreas da tecnologia.
Também interessante como o Google Translate reorganiza palavras para construir a estrutura sentance adequada. Eu deveria ter pensado para tentar este muito mais cedo.
Você tem sido capaz de encontrar um transistor em 1740 para corrigir o 1741?
  • Country: br
« Reply #267 on: February 20, 2016, 06:29:17 pm »
rf+tech só não use ponto nem vírgula (dot or comma) que o google tradutor trabalha melhor. E o transístor eu tenho um par no 1740A. Só não sei se são iguais.
« Last Edit: February 20, 2016, 06:38:44 pm by SAUL BRITTO »
Offline rf+tech
« Reply #268 on: February 20, 2016, 06:50:21 pm »
Saul,
Ao igual que eu acho que você quer dizer um par combinado. isso não é necessário. É apenas necessário utilizar dois transístores com o mesmo número de peça.
RF+ Tech
  • Country: br
« Reply #269 on: February 20, 2016, 07:46:59 pm »
rf+tech Eu só estou com dó do 1740A mas,não tem problema depois eu reponho os Tr.
Este Tr está com um comportamento estranho,quando a máquina está fria ele mostra uma resistência muito alta nas junções base emissor e base coletor.Depois, quando liga a máquina logo após um tempo, eu desligo e ele está aberto.
« Last Edit: February 20, 2016, 08:09:34 pm by SAUL BRITTO »
  • Country: br
« Reply #270 on: February 20, 2016, 08:06:17 pm »
Se você preferir podemos conversar em inglês ou,está afim de treinar seu português? :box:
Offline rf+tech
« Reply #271 on: February 20, 2016, 08:15:28 pm »
Saul,
Q1 e Q3 no mesmo circuito do 1740 como no 1741 têm diferentes números de peça. As tensões e circundantes viés resistências são os mesmos em cada escopo. A única diferença real é o plástico pacote de TO-92 no 1740 e TO-39 do metal no 1741.
Com os transistores de cuidados pode ser devolvido ao 1740 sem danos.
É uma boa coisa que eu olhei a lista 1.740 peças e veja esta pequena diferença. Não deve haver mais possibilidades de substituições PNP.
De minha mais excelente memória, TO-18 2N2907 metal e TO-92 PN2907 plástico são eletricamente equivalente a 2N2905. Também TO-92 2N3906 plástico é muito semelhante.
2N3906 Ft 250 MHz, ligeiramente melhor do que 2N2905, 2N2907.
Então você tem mais opções. Vou olhar para mais números de peça com ainda maior Ft.
Para preservar a largura de banda de 100 MHz requer transistores de substituição com Ft igual ou melhor do que os transistores originais. Não podemos saber Ft dos transistores originais por isso devemos contar com palpite e experiência prática.
Devo aposentar mais cedo esta noite para despertar às 03:00. I ir com um amigo e dirigir 265 km para uma hamfest (rádio amador reunião de troca). Talvez eu vou ter sorte e encontrar melhores transistores?
Traduz Google é bom para mim. Vejo muitas palavras portuguesas que reconheço a partir Español para que eu possa, por vezes, peça coisas juntos.
RF+ Tech
Offline rf+tech
« Reply #272 on: February 20, 2016, 09:23:57 pm »
Ok eu tenho mais alguns números de peça. Ordem é melhor primeiro.
NTE395 Ft 2300 MHz
2N4957 Ft 1600
NTE161 Ft 800 MHz
MPS-H81 Ft 600 MHz
NTE395 e 2N4957 tem TO-72 caixa de metal com quatro pistas. A quarta vantagem conecta internamente para o caso de metal. Este chumbo é ligado à terra na maioria das aplicações. Com Ft modo muito elevado, é costume colocar um pequeno anel de ferrite sobre a liderança base. Isso elimina uma tendência natural de oscilação em frequncies rádio muito elevados.
Minha melhor recomendação para a substituição permanente é NTE161 ou MPS-H81.
Para fins de teste nenhum dos outros transistores PNP com baixo Ft pode ser utilizado até mais transistores adequados podem ser adquiridos.
  • Country: br
« Reply #273 on: February 20, 2016, 09:28:58 pm »
Você está indo bem com seu português,afinal,aprender nunca é demais. E sim,temos realmente um transistor aberto,removi a placa e o retirei da placa e ele está aberto com certeza. Eu queria na verdade colocar dois transístores definitivamente e não ter que remover mais uma vez a placa do circuito,entende?
  • Country: br
« Reply #274 on: February 20, 2016, 09:34:13 pm »
eU ACHO QUE 2N3906 eu tenho por aqui em algum canto,talvez numa sucata.Eu vou examinar.
  • Country: br
« Reply #275 on: February 20, 2016, 09:54:25 pm »
Eu tenho um velho tektronix 475 200MHZ que está morto,certamente deve haver algum transistor equivalente em polarização e frequencia.
  • Country: br
« Reply #276 on: February 20, 2016, 10:30:10 pm »
Encontrei uma dezena de 2n3906 em um velho receptor de satélites para tv.
Offline rf+tech
« Reply #277 on: February 20, 2016, 10:35:12 pm »
Ainda melhor que você tem a Tektronix 475. O manual 475 mostra o número da peça SAB4113 para os transistores usados no mesmo local como no 1741.
Há oito locais no 475, onde este transistor podem ser encontrados. Quatro no canal A e quatro no canal B.
  • Country: br
« Reply #278 on: February 20, 2016, 10:40:17 pm »
 :scared: Um amigo desses vale mais que ouro! You are a gold friend! Epero que você faça bons negócios na feira do tróca tróca.
Offline rf+tech
« Reply #279 on: February 20, 2016, 10:56:10 pm »
Você não pode se surpreender ao saber que tenho vindo a reparar eletrônica de todas as formas profissionalmente há muitos anos. O que pode ser uma surpresa é que eu não tenho nenhuma educação formal em eletrônica. I abandonou a universidade antes de um ano.
O que eu tenho é uma habilidade incrível para solução de problemas lógico e compreensão prática da teoria. A maior parte do meu negócio é o reparo de placas em máquinas-ferramentas CNC. I fazer isso sem esquemas, porque os fabricantes não torná-los disponíveis. É um jogo de criança para mim para encontrar a falha e repará-lo. placas de circuito, mesmo gravemente queimado que são caros para substituir ter sido dada nova vida com minhas habilidades.
Obrigado Saulo para a excelente complemento. É um prazer para ajudar com este desafio.
  • Country: br
« Reply #280 on: February 20, 2016, 11:34:19 pm »
Estive conversando com um amigo sobre sua habilidade e ele me disse que isso é um dom e,você acaba de confirmar.Algumas pessoas nascem pré-dispostas. Isso em português chama-se dom (Natural gift).
Estou muito orgulhoso em ter um amigo como você.
  • Country: br
« Reply #281 on: February 20, 2016, 11:39:10 pm »
Eu tenho um dom igual ao seu em música,aprendi tocar,compor e arranjar sozinho,sou capaz de criar uma melodia em minutos! Sei bem do que voce fala. :-DD
  • Country: br
« Reply #282 on: February 21, 2016, 01:09:20 am »
rf+tech the anterior owner has replaced R3 383 Ohms for a 331 Ohms it's connected to +48V on the V amp board. Think you this arrangement  can be a problem?
  • Country: br
« Reply #283 on: February 21, 2016, 01:10:26 am »
It looks like be a 2 watts resistor
Offline rf+tech
« Reply #284 on: February 21, 2016, 01:35:17 am »
Sim, 330 Ohms pode ser muito baixa. A melhor escolha seria 390 Ohms. Algo que pode não ser evidente no primeiro olhar porque todos estes valores de resistência ímpares. De meus anos na indústria é bem sabido que as tolerâncias de componentes terão frequentemente adicionar juntos criando problemas com a reprodutibilidade circuito. A melhor maneira de evitar este problema é para apertar as tolerâncias de componentes. HP faz exatamente isso usando 1% resistências em muitos lugares para minimizar a variação e garantir a cada um executa circuito como o designer pretendia.
Outro método de fabricação envolve a seleção transistor para controlar firmemente determinadas especificações que são fundamentais para o bom desempenho de um circuito. Esses transistores são marcados com um número de peça HP que não tem referência cruzada. Um exemplo dessa prática é Q17 e Q18.
Espero voltar para casa tarde de domingo.
  • Country: br
« Reply #285 on: February 21, 2016, 02:08:32 am »
Well... Beer on the hand! We have two traces! :box:
  • Country: br
« Reply #286 on: February 21, 2016, 02:10:19 am »
I think will need some adjusts but... It's alive.I  put two tektronix Tr.
« Last Edit: February 21, 2016, 02:17:42 am by SAUL BRITTO »
  • Country: br
« Reply #287 on: February 21, 2016, 02:20:45 am »
The traces just comes at the center of the screen with  "position" totally to left.
  • Country: br
« Reply #288 on: February 21, 2016, 03:15:14 am »
Look at that rf+tech:
  • Country: br
« Reply #289 on: February 21, 2016, 03:41:56 am »
  • Country: br
« Reply #290 on: February 21, 2016, 04:02:03 am »
 rf+tech,you will be back,sure.
  • Country: br
« Reply #291 on: February 22, 2016, 02:56:18 am »
rf+tech how was your hamfest?
  • Country: br
« Reply #292 on: February 22, 2016, 09:25:51 pm »
rf+tech the scope is working but I did see that it measures not so right, where is  1Vpp it measures 0.9 or 0.8 V, can it be that transitors?
« Last Edit: February 23, 2016, 01:35:47 pm by SAUL BRITTO »
Offline rf+tech
« Reply #293 on: February 23, 2016, 02:10:24 am »
Saul,
Please check your messages.
RF+ Tech
  • Country: br
« Reply #294 on: February 23, 2016, 02:32:50 am »
Saul,
RF+ Tech
Ok my friend,thank you.
  • Country: br
« Reply #295 on: February 24, 2016, 01:10:36 am »
I am back here to say THANK YOU for all the forum people and in particular to the friend "rf+tech"
 for all your kindness and patience. My scope is ok!!!