Hi everyone!
I just got a used Keithley 2000, datecodes indicate end of 1997. On first try it powered up and seemed to work fine. I then brought it into my "lab" and it didn't want to start anymore: When I switch it on, it beeps once, all display segments are lit and that's it. It simply stays like this forever, no reaction to any button presses either.
- The inside looks ok, no obvious damages, blown caps or similar stuff. Just one bodge-wire on the main processor.
- All the supply rails and the reference are fine (although +-15V is at the limit of the spec with +-15V83).
- The oscillator for the main processor works.
- I poked all the memory chips with a scope. Interestingly the 24LC16B IIC EEPROM (U136) is never accessed.
- There is continuous activity on the bus for the RAM and OTP EPROMs, even after the display locks up.
I have the user/repair/calibration manuals and some schematics, but they don't really help for this problem.
What do you guys recommend to troubleshoot/repair this thing? Although I have no idea when the meter was last calibrated, I'd like to retain the calibration values if at all possible.
EDIT: Possibly good news, I can get it to fire up. Just have to put it in the freezer for 10min
So there is some kind of temperature dependent problem. My guess would be that some timing is out of spec at higher temperatures leading to faulty data in the self check at startup. Any hints to which component could be the likely cause?
Regards,
Michael