After the last post, I tried quite a few things. Let me list them somewhat chronologically:
I did detailed review under microscope of guard traces and also checked continuity of them and possible shorts between them and microchips - nothing suspicious.
Then I did measure a Drain on Q214 for AC component with oscilloscope. I placed scope ground on Common test post that sticks up next to voltage reference under ADC. I see some low frequency and high frequency components, but I am not sure if this is significant, or I am just picking up some unrelated spikes. See attached.
I let meter to warm up for 3.5 hours and remeasured voltages at att transistors involved. See the Transistors1.jpg attached. Interestingly the meter drifted closer to be in spec, yet some voltages are almost double of previous measurements of good 2002 meter. Is it possible that from 1994 to 1997 they changed not only some microchips to their equivalents but also changed operating modes?
Today, being out better options, I replaced U234 (DG411) and two diodes next two it. When I removed U234 I noticed that R277 was pushed to a side and down (not by me) and basically laying on some traces and vias under it. I measured about 10 to 20MOhms between either lead of the resistor and its metal body. I did not like that for a 750k resistor. I used a small screwdriver to push the resistor higher. Afterwards the resistance between leads and the body became greater than 100M. I thoroughly cleaned the area around and reinstalled fresh U234 and diodes.
After U234 change and this last round of cleaning around U234, performance of the meter improved quite a bit in both Ohms and Amps.
Another positive is that it now passes all self tests, even when cold! If comparing 1 hour warm up readings, it used to read 98.6k and now is reading 99.95. The phantom current on 2A range after 1hr used to be -950uA. After initial cleaning it went down to -337uA and now I get -23uA.
After 2 hours the meter is very close to be fully functioning. I get 99.985k that is no longer improving, +6uA reading on 2A range.
I am still not sure if I am dealing with bad component/components or dirty board. When I got the meter the board was dirty to the point that I was initially removing mice excrement and dead insects from it.
The temperature dependence of performance is quite puzzling. I am considering putting a gloved finger on microchips one by one, trying to find one that reacts to temperature the most. Also I can use cold canned air to cool different areas of the board.
What else can I try? How can I distinguish components trouble from board trouble?