Author Topic: (FIXED) Keithley 225 (current source) repair  (Read 23265 times)
Offline motocoderTopic starter
  • Electrical Engineer
« on: May 22, 2015, 07:07:39 am »
I picked up a Keithley 225 off of eBay. I probably paid too much for it given it's in need of repair, but I thought it might turn out to be one of those units where the seller just didn't know how to operate it.
There are at least two things I've found wrong with it so far. The output was pegged on -130V no matter what I set the controls to. In addition, the 100 ohm precision resistor on the main current range switch is toast.
Both output driver transistors have failed with collector shorted to emitter. I guess this is why the output was stuck on -130V. So first question I have is where to source a suitable replacement. The part seems to be a TO-66 package, which really limits the selection. I've attached a picture of it. I believe the part number is RCA 60761. This does not seem to match the number in the Keithley manual, so either they swapped parts in later generations of the product, or someone has already had a go at replacing this part.  The manual says Mfr. Code = 80164, Mfr. Desig. = 40318, Keithley Part No. = 24560A. It also has an asterisk next to it which leads to a comment about it being a specially selected transistor (selected for good hFE?).
So my first question is how to go about selecting a suitable replacement transistor. I did a search on "RCA 60761", and a page came up listing NTE124 as a replacement. I have a distinct mistrust of NTE, as they seem to sell whatever crap fell off the boat, with minimal data sheet info and no guarantee of consistency or quality.
The manual for this beast is here:
http://exodus.poly.edu/~kurt/manuals/manuals/Keithley/KEI%20225%20Instruction.pdf
« Last Edit: June 07, 2015, 11:22:06 pm by motocoder »
Online johansen
Re: Keithley 225 (current source) repair
« Reply #1 on: May 22, 2015, 09:37:41 am »
they are likely matched hfe npn/pnp pairs
replacement should not require an exact match, but ymmv.
you've got the schematic.. what is there to complain about?  :D
« Last Edit: May 22, 2015, 09:39:17 am by johansen »
« Reply #2 on: May 22, 2015, 12:24:56 pm »
They are both NPN, and the manual says "specially selected, order individually". There is another part where it says to order in pairs. So my guess is they are just selected for good hFE.
No complaining here. I am actually quite happy with the condition of this thing. If I can get it working, it will be a beautiful addition to the lab.
My local electronics store stocks the NTE part, so will just give those a try.
« Last Edit: May 22, 2015, 01:46:15 pm by motocoder »
Offline Grapsus
« Reply #3 on: May 22, 2015, 09:48:28 pm »
The part you are looking for is "RCA 40318". I can't find the RCA datasheet, but there is one by "New Jersey Semiconductors":
http://download.njsemi.com/40318.pdf
All the characteristics seem to match the RCA part which is described here:
http://alltransistors.com/transistor.php?transistor=20627
There is even a french site selling the exact RCA part:
http://www.donberg.fr/descript/4/40318.htm
« Reply #4 on: May 23, 2015, 01:01:40 am »
Thanks, Graspus. I just noticed a bunch of toasted resistors on the range switch. I also see signs of someone that tried to make some repairs. However, in at least one case they used the wrong component value, so the whole thing is suspect and will need to be checked...
« Reply #5 on: May 24, 2015, 03:15:49 pm »
All the non-precision resistors on the range switch (R102, R103, R107, and R108) are toast, as well as precision resistor R185. Those are all easy to replace. However, the pot on the end of the switch assembly (R138) is also damaged. It has a mechanical coupling to the shaft connected to the center dial. No idea where to get a replacement that will fit properly.
The pot is from CTS corp, and browsing their data sheets, I think this might be work:
http://www.ctscorp.com/components/Datasheets/450.pdf
450TA20S501A1A1
However, that's now a custom part apparently. Their stock parts seem to use plastic shafts, but may work, so I've ordered one of those.
« Last Edit: May 24, 2015, 03:55:06 pm by motocoder »
« Reply #6 on: May 25, 2015, 05:08:19 am »
Looks like I'm typing to myself here, but I'll continue to update the thread in case it helps someone with a similar repair in the future.
So, the monkey that owned this thing previously replaced the 1/4A fuse with a 2A fuse, which explains why the range switch resistors were destroyed. The resistors that were destroyed were:
R185 - This is a 100 ohm 0.1% resistor that is used for the second highest current range.
R107, R108, R102, R103 - These appear to be used, along with Q103 and Q106, for a range-specific gross current limit.
For the gross current limit, I believe it operates as follows: the voltage across R102 or R103 (depending on which range is selected) is Vbe of Q103. So as the current through R102 increases, eventually Q103 starts to conduct. This has the effect of pulling down the base of driver transistor Q102, which makes Q101 conduct less, dropping the output voltage. This limits the current to ~150mA or ~33mA, depending on which range is selected. There is a similar mechanism for R107/108 and Q106 (although I'm not sure why R108 is a different value there).  Because Q101 and Q105 failed in a short condition, this current limit was not effective, and large amounts of current flowed. Because previous owner-monkey replaced fuse with 2A version, those resistors cooked until they eventually failed. I guess he switched it around to different ranges to make sure he cooked all 4 resistors.
So I replaced those resistors, but also decided to check the transistors (Q102 and Q104) that drive the output transistors, and the current limit transistors (Q103 and Q106). The driver transistors look OK, but the current limit transistors were both bad one shorted, one open circuit).  Q103 is a 2N3565, and Q106 is a 2N3638. These both seem to be unobtanium, so I replaced them with 2N3904 and 2N3906 which I selected to have similar beta.
Checking the output voltage now, it's not pegged anymore, but it's still not working correctly. For example, it does not respond to changes in the output current selector switches. Also, switching between positive and negative output polarity does nothing; it is always outputting a positive voltage. It does seem to respond a little to the "compliance voltage" (voltage limit) pot, but even there it's not operating as  you would expect.
I suspect the op-amp that or something else in the feedback loop is fried. Will check that tomorrow.
« Last Edit: June 03, 2015, 03:54:17 am by motocoder »
« Reply #7 on: May 25, 2015, 05:35:19 am »
I checked QA201 output (voltage regulator). It need a little trimming, but it's outputting 24V. However, QA202 is not working correctly. This IC is an Op-Amp that derives a virtual GND half-way between +12V and -12V from the regulated output of QA201.
Offline cncjerry
« Reply #8 on: May 26, 2015, 12:37:24 am »
If you pull the plug on top of page 24203e that connects those transistors you should be able to check them in situ with a dmm diode tester.
« Reply #9 on: May 30, 2015, 11:51:28 pm »
Hi cncjerry -
I am not sure what you mean about "page 24203e". However, I do have some good progress to report on the repair.
First off, I found a bunch of bad parts. It looks like some very high voltage or over-current situation happened on one or both of the output channels. I found a bunch more bad parts. And I'm not speculating when I say "bad". I mean they measured bad in-circuit, I removed them and they still measured bad.
So far this is the tally:
Resistors burned in half: R107, R108, R102, R103, R185,
Resistors, no visible damage but resistance way off: R138 (pot, just damaged on one end of the range), R101 (marked 12K, measured 20K), R120 (marked 10K, measured 17k).
Transistors: Q103, Q104, Q106, Q115, Q116
Also I was wrong when I said 24V regulator (QA201) was OK - that and the op-amp (QA-202) being used to generate the "virtual ground" for the +/- 12V supply were both bad.
This last issue was what really confounded me on the repair. I would measure various voltages in the circuit, but everything was way off. There would be no logical change in voltage at the appropriate test point as I adjusted the KV range switches. This turns out to be because the buffer op-amp I used to replace QA-202 did not have enough current drive. I had convinced myself there was another bad component dumping too much current into this net and causing the problem, but after carefully reviewing the schematic, and doing some testing out of circuit, I came to the conclusion that the op-amp just didn't have enough drive.
I've ordered something from DigiKey that should work better, but being impatient, I decided to just whip up a little circuit of my own. I used a 741 op-amp (it's what I had laying around) and a couple of transistors in a push-pull configuration to create a beefier virtual ground. After wiring this in, I can see a very steady ground node for the +/-12V supply - it changes less than a uV as I change the range selector and output selector switches. And wonder of wonders, I can actually see the KV divider working now.
So, I am waiting on a replacement for Q104, and then I will put that in, reconnect the +/- 130V rectifier and give the whole thing another try. My only real concern at this point is that all the transistor replacements will not turn out to really be a good match for the parts they replaced, and there will be some problems created because of that.
« Last Edit: May 31, 2015, 01:24:02 am by motocoder »
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #10 on: May 31, 2015, 12:16:02 pm »
Great job.
Some photos of internals would juice up this thread even more  :-+
« Reply #11 on: June 03, 2015, 03:04:55 am »
Great job.
Ok, I'll add some photos to this post.
I am starting to get a bit frustrated with this thing. I keep finding more bad parts. The latest is variable resistor R121. The manual says this pot is the adjustment for the bias current through transistors Q101 and Q102. However, after replacing the pot, I measured the voltage across the three diodes that are part of that circuit (D108, D109, and D110). The voltage across these is so low, that it's hard to imagine that this circuit is doing anything at all currently, and this voltage, nor that across the pull-up resistor (R120) does not change at all when I adjust R121. I don't really understand this part of the circuit, so hard to say what is going on there.
I spent a good deal of time looking at the very first differential stage, which uses a matched JFET pair (Q118A and Q118B). I took some measurements of D/G/S voltages for these JFETs as I adjusted the current value and range values. It appears that this circuit is completely biased over to one side, with the JFETs so saturated that they don't do much. This would explain why the output of the unit is always a positive voltage, regardless of which polarity I set the output selector to. One of the pics below shows the voltage measurements I took on that JFET - all voltages were made with respect to the 12V ground (this ground effectively "floats" on top of the voltage output by the series driver transistors, before the output current sensing resistors that connect to the output). I tried adjusting the "Amplifier Zero", which is a pot connected to the source of Q118A/B, but it has no effect at all. This isn't too surprising, because I think something else in the high gain stage is badly skewed to one side. I didn't capture it in the initial measurements, but I also took some measurements with the output polarity reversed. I could see the polarity of the gate voltage of Q118B reverse, but because of the aforementioned problem, this has no effect on the output voltage.
Finally, I have some questions for anyone who owns one of these. The manual has a diagram (Figure 16) with some points to use to measure various voltages during calibration. However, it's really hard to tell were the attachment points are for the "Ref or Zero" measurements. In Figure 16, it just points at the potentiometer, and at some unlabeled point on the range selector switch. Can someone clarify exactly where these attachment points are?
Range selector resistors burned in half (bottom left side of photo, just to the right of the black wires)
Top view of main circuit board
Measurements I took on Q118A/Q118B - first differential amplifier stage
« Last Edit: June 03, 2015, 03:12:53 am by motocoder »
« Reply #12 on: June 03, 2015, 03:51:54 am »
Ok, looks like something is wrong in the "transient absorber" section (Q401, Q402, R402 - R405). I measured the voltages on the transistors and the voltage on the emitter of Q402 is not what it should be (-3.30V instead of the expected -1.04V). This circuit is connected to the gate of Q118A, and so might be the source of the imbalance problem.
« Last Edit: June 03, 2015, 04:02:49 am by motocoder »
« Reply #13 on: June 03, 2015, 06:06:41 am »
Bad idea to remove the "Transient Absorber" transistors and power on. Now I'm going to need a new JFET pair...
« Reply #14 on: June 04, 2015, 04:56:18 am »
Big breakthrough this evening. I replaced the part I burned out with my dumb idea to remove the "transient absorber" transistors. I also replaced those two transistors since it wasn't possible to salvage them (leads too short).
Then I started walking through each transistor in the high gain stage. I discovered that the base of Q115 was pegged out at -12V, clearly not correct. The problem turned out to be Q116, which was shorted between collector and emitter. I replaced this and Q117 with a pair of ZTX550's, which I selected to have fairly close beta. Not sure those are a perfect match, but they can sustain more current than the original part, which seemed to be the primary reason it was selected.
Now I can see the output voltage change as I turn the selection dial. I've got more to check to confirm whether or not it's all working, and eventually need to get my questions about the calibration test points answered so I can do a calibration, but this is good progress.
The following users thanked this post: Spike
« Reply #15 on: June 04, 2015, 01:10:05 pm »
Well, was all working!
But then while I was trying calibrate it, R105 went up in smoke. I figured those current-limiting transistors had gone out. Checked and sure enough, Q103 was bad. Replaced Q103 and Q106 with beefier ZTX450/550 parts, put it all back together, and it burned through another R105. Quick check of output transistors and I think they might have gone too Output transistors are still good.
Not sure what is causing this.
« Last Edit: June 04, 2015, 01:35:52 pm by motocoder »
Offline cncjerry
« Reply #16 on: June 05, 2015, 07:21:33 am »
I cant see how the base resistor r105 would burn out unless q105 is shorted.   also, wrt  my earlier note, there is a plug with the reference 'see dwg 24203e...' on it.  if you pull that plug you can diode check all those transistors possibly from the plug itself.
one other thing on r105, did you scope it after changing the transistors to another type?  Maybe it is oscillating.  Current sources are known to have problems with certain reactive loads.
« Reply #17 on: June 05, 2015, 12:28:53 pm »
I pulled Q105 out of the socket, and it is fine (not shorted). I checked Q101 as well while I was there. What's interesting is that the negative polarity is not working at all anymore. It did work before the R105 problem. I don't have a thermal camera, but using one of those little infrared thermometers, it appears to me that Q104 is heating up quite a bit. Q104 has been replaced with a 2N5416S, whereas Q102 (the symmetrical/complimentary transistor on the positive side) is still the original part (MM3003). Maybe the overall gain between the two polarities is too mismatched now?
Regarding your comment about plug dwg 24203e - there is no plug anywhere on the board. What you're referring to is just how they've split the schematic into two parts. There is no physical connector between the two parts.
« Reply #18 on: June 05, 2015, 01:29:47 pm »
Something is definitely fishy here. I measures the voltage across R105 and it was ~55V. THen just a few minutes ago, before connecting my scope to it, I measured again, and it's just a few 10's of millivolts. And there is a fair amount of noise on the Vbe of Q101 (~60mV), and to a lesser extent Q105, so there is some oscillation looks like.
Also, I can definitely see that Q104 is heating up, even though when I measure Vbe across it, it is just a few tenths of a volt, and Vce is the full 130V. Maybe enough leakage current there that with the high Vce value it generates some heat.
« Last Edit: June 05, 2015, 08:24:57 pm by motocoder »
Offline cncjerry
« Reply #19 on: June 05, 2015, 10:34:07 pm »
did r105 pop again?  Sometimes the burn out and don't flame.  I don't remember the value, but do the power math at 55v to be sure it isn't too much which would point you somewhere.  when I had a resistor popping in my keithley 228 VI source, one cool product by the way, I put 1/4w in to make it easier to blow.
I'm going to go look at the schematic again.
Offline cncjerry
« Reply #20 on: June 05, 2015, 10:45:08 pm »
So r105 is 180ohms.  At 55V my ohm's law says it is handling 16.8W  (55^2)/180 = 16.8W.  So that means the bias is wrong on Q105 which is determined by Q104, those diodes, R104, the cap I can't see and what looks like a ladder network of some type.  I assume you checked Q104 and those diodes as well as the cap.  It could be just a gain issue with the transistor you put in.  But Q105 probably won't last long with all that base current.  Without digging in to far, I would go back and do some ohm checks.  If it get tricky to figure out, you can throw the circuit into LTSpice and try to figure out what would cause the 55V across r105.
I've been thinking about a 225 to complement my 228.
« Reply #21 on: June 06, 2015, 12:37:29 am »
It's actually a little toasted, but I've measured it a few times with the ohmmeter and it's still measuring 179 ohms.
« Reply #22 on: June 06, 2015, 01:08:31 am »
I can't reproduce the 55V across R105. Maybe I just misread a "mV" on the meter instead of "V". Probably not, though, given it burned out already once.
I've been staring at the schematic for a couple of weeks now, and I think I have most parts of it figured out. I think the diodes you're referring to are D108, D109, and D110. These, along with R121 bias the voltage across Q104 and Q102 such that they are separated by a couple of volts. This last stage is basically a push-pull amplifier, so they're biased apart a little so there isn't a dead zone near zero. A lot of times people use a "Vbe multiplier" to do this, but here they just used 3 diodes and a potentiometer (an inferior choice, IMHO). The bias point difference, however, means both complementary stages will be conducting some current at 0 output. The current can go from +130V, through Q101, out through R102 or R103 (depending on range switch setting), and then back through R107 or R108, and then through Q105 to -130V. All of that can happen without changing anything in the feedback loop, so without some limiting mechanism, it could easily run away. That's what Q103 and Q106 are for. Using Q103 as an example, and assume range switch has seletected R102. Q103's base is on one side of R102, and its emitter is on the other. R102 is 3.9 ohms. Q103 will be conducting pretty good when it's Vbe is around 0.6V. That happens when the current through R102 is 0.6 / 3.9  = 153mA.
Now I think the beta of Q101, the replacement I put in, was around 48. So if ~153mA is flowing through Q101 emitter, that means Ib is around 153/(1+48) = 3.2mA. Let's assume most of this current is coming from Q102. At 0 current output, it's dropping about 130V minus a couple of Vbe drops from collector to emitter on this transmitter. So that's about 0.4W being dissipated. So I did the calculation for the wrong side :), but let's assume they're similar for the negative side. I replaced Q104 over there with a 2N5416S, which looks to have a slightly lower power dissipation rating. Maybe it's just overheating?
But either Q104 is now burned out, or there is some other part that went out because the negative polarity is not working at all - even when I start with everything cold. It does seem to work on positive polarity just fine, although as I turn up the current, Q104 seems to be heating up faster.
I did check all the transistors and diodes in this part of the circuit with the diode check function on my meter (in circuit), and they all showed up fine. However, I've had that mislead me on a couple of parts already. I hate to desolder anything, though, as the traces on this board are in rough shape after all the rework...
« Last Edit: June 06, 2015, 03:13:21 am by motocoder »
« Reply #23 on: June 06, 2015, 03:44:35 am »
Yes, Q104 was bad. My transistor tester said it was a "common cathode diode network", so it still looked like valid PN junctions, but just wasn't operating like a transistor. This is why I didn't detect the issue in circuit.
I've replaced it with a NTE397, which NTE lists as an equivalent part for the MM3003. I am going to go very slowly and make sure nothing is overheating this time.
« Reply #24 on: June 06, 2015, 06:41:03 am »
Ok, it's *mostly* working, but have a few problems:
- I am not able to set the bias for transistors Q101 and Q102 as described in the manual, table 6.2d on page 19. I was able to set this before, so something is definitely still not right. The bias voltage is way too high, which means there is too much current flowing at equilibrium. I think the problem might be that the gain for the two polarities is too imbalanced.
- There is definitely some oscillation that was not there before when it was briefly working perfectly.
- I can't set the positive output voltage past 30V. I can set the negative polarity, but just not positive. I think this is related to the gain problem as well.
- I noticed that the package for the power transistor I am using (NTE124) says maximum Ic is 0.1A. Now the NTE data sheet says 1A, but I'm not sure I trust that. I did find a source for the Central Semiconductor parts, mentioned earlier on the thread, that are a very close match. I may try those.