Hi!
If the current display changes to "0" or similar TP307 & TP311 are short-circuited, the fault lies in the current-sense amplifier between the output-current shunt and the test-points TP307/TP311, whereas if short-circuiting these two TPs together makes no difference to the "OVERFLOW" indications, the fault is between the TPs and the ADC/microprocessor circuitry.
It seems like there's an open-circuit somewhere in the current monitoring loop after TP307/TP311 causing the ADC/uC to overflow!
Unfortunately I can't offer much more assistance beyond this, as it really needs someone with one of these units to start reverse-engineering the main PCB between TP307/TP311 and the ADC/microprocessor & LCD circuitry - we need to identify what circuitry is between TP307/TP311 and the ADC, and also the current-amplifier circuits between the current-shunt and TP307/TP311!
Don't forget tho', there is a full component-level parts-list given by Keithley in the service manual, my recommendation being to print out those pages of the parts-list covering the "300" block of components, and then start testing all the resistors and capacitors from R301/C301 upwards for faults against the stated values in the parts list - these will cover a lot of the current-measuring circuitry.
If you can't find anything amiss with the Rs and Cs, repeat with the diodes and discrete-transistors in the "300" block of components.
You can use a Chinese Tester or a DMM with a Capacitance Test range for these checks, then anything wildly out can be removed for an out-of-circuit test!
I don't have, and can't afford to buy, one of these units at the moment, so can I pass this thread over to someone who has a 2303 and is willing to consider a rev. project please?
My apologies!
Chris Williams