I started with checking the transistors of the output stage and found out that Q521, Q523 and Q525 had a short between drain and source or collector and emitter respectively.
The good news is, that after removing these three components, the power supply stoped breaking down and the 2400 finally started up with the analog connected.
I am not shure, how the unit should behave without these components. I am currently measureing 240V on the output, even with the output swiched off. I belive, this is not correct (even with half of the output stage removed) and I have to look for further damaged components.
I try to understand the (partial) schematics of the output stage shown in the service manual. Does someone know actually, where the output is connected to this circuit?
the partial schematic in the service manual is not error-free, but it gives you a picture
the output of the power stage is the triangle marked "F". it is then routed through the current sensing resistor bank (the blue ones like R451, R452, and so on), then passes a semiconductor disconnect switch (made up by Q415, Q414, Q412 and a PV optocoupler), going through the disconnector relay K206 and the FRONT/REAR selector relays K207, K208, to finally arrive at FRONT or REAR outputs. so, when setting the instrument to OUTPUT OFF, there should always be approx. zero volts across the output terminals.
note that the bipolar supply of the push-pull driver OPAMP U500 is critical for proper operation. the supply is quite "whimpy". so, if the opamps's supply voltages are less than expected (V+ and V- referred to
TP501 OCOM TP500 FCOM!!), then you may want to replace U500, and expect a hole in your pocket. i can only assume that VR500 and VR501 may be part of the supply circuitry for that particular opamp, because that is exactly the case for my 2410.
your 2400 is slightly different in that respect, but TiN 's website
https://xdevs.com/fix/kei2400/ may give more clarity in this case.
EDIT: forgot to mention, make sure that your instrument is in 2-WIRE mode, and _not_ in 4-WIRE mode. if you operate it in 4-WIRE mode, without having the sense inputs connected, the ouput voltage can indeed go up to high voltages, no matter what the output voltage was programmed for:
try [CONFIG], [V], SENSE-MODE, [ENTER], 2-WIRE, [ENTER], [EXIT]
EDIT: corrected OCOM and FCOM