I checked the voltages on the CRO pins;
Cathode pin 2, -1760 volts.
Control grid (1st grid) pin 3, -1120 volts
Focus grid (2nd grid) pin 4, -1760 volts.
Those are almost certainly wrong. Since the resistors are high impedance, it is possible that the meter is loading the chain and affecting the indirect measurement of the key value, V
gk. It would be better if you measured V
gk directly. Ob warning: be careful with those voltages! Keep one hand stuffed in your back pocket!
Based on my debugging a Tek465 that was acting as a torch[1], I would expect to be in the range 40-150V, depending on the brightness. The V
gk waveform should be an exact copy of the LT z-axis brightness waveform, translated to the cathode voltage. Of course V
gk is difficult to measure without an HV probe, but the LT variant is easier
Typically problems with brightness are related to the DC restorer circuit, and people usually just replace the caps and diodes rather than debug specific items.
Of course, if you don't want to debug that scope, working Tek 475/2445B are available at reasonable prices; PM me if interested - I come to London intermittently.
[1]
https://entertaininghacks.wordpress.com/2016/03/09/rescuing-a-broken-tektronix-465-crt/