Hi Everyone, my first post.
Thanks to Dave for the great videos and all contributors on this board,
it is great reading, and clearly a great wealth of knowledge here.
I would like some help with a cheapie scope purchased on Ebay that had
an EHT fault, since repaired but now I have no intensity control.
Both A&B traces are there and intensity varies with the timebase control but
the Intensity control does nothing.
So far, all the .01 2KV caps and the 2 22Meg resistors have been
replaced and a couple of burnt carbonised spots on the board have been
scratched out which got the EHT working and the scope traces back.
My issue now is the intensity control has no effect.
Focus works, Astig, Symetry works. Trace rotation works. The main Power
supply voltages are correct, however I don't have the gear to measure
the EHT.
I have 74.8V (Max clockwise) and 122.5V (Max anticlockwise) at the
junction of C424/D409 when I turn the Intensity control but the voltage
between the Cathode and Grid of the CRT only changes by .5V, and
depending on the timebase setting this voltage is anywhere between 3.6
and 21 volts.
I am measuring this across the 2 22Meg resistors R422 and R423. What is
weird to me is that the 3 neons are lit constantly at the same
brightness... always... 3.6 - 21 volts
having the CRT plug connected or disconnected makes no difference.
The manual is here
https://elektrotanya.com/protek_p-3502c_user_sm.pdf/download.htmlThanks in advance, any help appreciated.
Grant