I'm still trying to figure out the telephone stuff:
- DBM BRDG
- METER SHORT
- TEL SET
- DBM 900ohm TERM
DBM BRDG = dBm Bridged. The meter is calibrated on the assumption that it's measuring across a circuit that is already terminated. It isn't explicitly stated, but the circuit impedance is probably 900 ohms. This is the standard impedance for a subscriber line.
METER SHORT is just what it says. It puts a short across the meter. This helps reduce the sometimes violent motion of the meter pointer when the meter is being transported. Sensitive meters can be damaged when they're being tossed around in the back of an installer's truck. Hold the unit in your hands and twist it from side to side along the needle's axis. Do it with the switch on METER SHORT and on any other position. The motion of the needle should be much less on METER SHORT.
TEL SET = Telephone Set. Connect a test phone to the Tel Set terminals. Set the meter to TEL SET. This connects the input terminals to the TEL SET terminals. Dial the test number, then turn the dial to the 150 MA position to measure the current in the line. Based on the known line characteristics you can get an idea if the line has some bad connections. You can also see if the current is unstable which also indicates problems on the line. Line current should be greater than 23 ma. If you're right next to the telephone office, it might be as high as 125 ma. or so.
DBM 900ohm TERM = dBm with a 900 ohm termination. Same as DBM BRDG except the meter provides the termination.
Regarding the batteries, I've never used a TTS-28, but 9V feels like it's too low. If you replace the 9V battery with a variable power supply, how high can you set the voltage and still get it to zero?
Ed