When calibrating a meter I noticed that I was having a lot of difficulty nulling the input amplifier. In order to investigate I attached an oscilloscope, and noticed that the output of the amplifier is oscillating. The thing is that it is just barely unstable - reset the meter and everything seems good for a while until after a significant amount of time (5 minutes to two hours), without any repeatable warning, the output starts oscillating. It always ends up at 18 kHz, 500mV peak-to-peak.
Looking at the schematic, I can see that they already added capacitor C1201 after the first version - in the meter currently on the bench it is bodged on the bottom across the pins, on a second (newer) meter, there is an additional footprint and pcb traces provided for this capacitor.
My initial thought was that maybe the capacitor had failed or at least was degraded in performance, but after replacing it with a new capacitor of similar specifications (I think actually exactly the same capacitor from philips, just newer), the problem was still there.
Ofcourse, I can just slap on some more capacitance in the hope that it further stabilizes the amplifier but it is hard to know how much I should add. In addition, I don't want to alter the circuit too much as I don't want to negatively impact the performance.
In addition, how do I actually know I fixed the problem? If it is unstable it could be it take 5 minutes, 5 hours, 5 years before it actually starts oscillating...