That doesn't sound so good.
Check all semiconductors of series pass element A7A1.
Then also check the pre-regulator board, if all semiconductors are still fine, and if the small signal transistors have sufficient beta.
I assume, that the HV pre-regulation is not working correctly, or is not steered correctly, so that the discharge reed-relays K1A or K2A are energized and shortens the HV output over the 2k power resistor R19.
If the pre-regulator A7A2 is damaged, specially the power zener diode CR14, or Q1, the diode bridge will be shortened and HV will be generated full power, which the series pass element then tries to clamp down.
I had that latter problem..
For a quick check, you may also remove (pull out) the preregulator A7A2 board first, and apply an external voltage from a PSU, current limited, between -output, and at the F1 fuse, at the node K1A/K2A /R10.
The voltage should be at least 10V higher, than the programmed output voltage, for checking the correct output regulation.
That's safe, as no HV is present , but you have to engage the safety switches, if the aluminum cover is removed.
What happens on K1A/K2A?
Check, if the uni-junction circuit Q3 across Q8 produces pulses if the voltage across Q8 is too low..
I think, 36V is required at least, but I don't remember any more correctly.
Frank