Hmm, unplanned repair log
My
EDC MV106 DC voltage standard stopped working suddenly without visible reason. Was fiddling with freshly received
Keithley Model 182, sourcing various voltages from EDC while suddenly overload lamp lit and now unit stuck to it. Disconnected everything, toggled AC power, reseated sense shorts, no any improvement.
Output is stuck to 13.18V on 10V range, or -13.18 in inverse polarity switch, or 0.125V/-0.125 in 100mV or 13.18mV/-13.18mV in 10mV range. Changing knobs does not affect voltage output, and lamp overload stays on.
Measured zener voltage, it's same voltage as I had it measured in February 2014, 6.1698V, so it's likely output amp went dodgy?
Bummer