After following DaJMasta's Repair log for his SMIQ06B I'm finally getting started on repairing mine.
for his repair log see here ->
https://www.eevblog.com/forum/repair/rohde-schwarz-smiq06b-repair-log-6ghz-board-alc-error/50/Like DaJMasta I got mine of eBay with a stated status of a Level Preset error but output is stable and with-in .5dBm of requested value.
I do not have a SA or Scope that can go to 6.4Ghz but I can go to 3 Ghz on my frequecy counter and if I put a divide by 4 between the SMIQ and my SA I can at least see the signal but can not use it for amplitude checks.
I have a E4418A and can check power and in deed it's is in the 0.5dBm range and the Level Preset is the only error reported.
I produced two items to help troubleshoot this
I cleaned up and stitched the E6GHZ Service Manual and it is available here
https://drive.google.com/open?id=0B0n5ltUlE8YHZ1laejFma245V00 I hope
I also created a excel spreadsheet to log Test data which you can get here
https://drive.google.com/open?id=0B0n5ltUlE8YHbjlQUG4yNlJuNlU i hope
here's a screen shot of the spreadsheet. the TP's are not in numeric order but in order of the signal path from left to right.
this helps make it clearer where the problem is located. (at least I think so)
My initial thought based on the test point data available in the diagnostic menu indicated a problem in the IF900 path somewhere around or before N4.
I connected up some test points as seen here and measured bias voltages
TP Value Actual
N39 4.9 ± 0.1 4.67
V91 4.7 ± 0.1 6.28
N3 4.8 ± 0.1 4.95
N4 4.8 ± 0.1 4.92
V91 is suspect and I found one of the SHF-0186-K (datasheet here
https://drive.google.com/open?id=0B0n5ltUlE8YHZlZrWkRvQXV6NEEI have the SHF-0186 on order and should be here this week
Also N39 is a bit low 4.8 to 5.0 is specified value in the service manual. the service manual though is for a 01 version and I have a 02
So as I am writing this up I decided with all the test points wired up I'd run a full calibration the bloody thing calibrated
Doing some quick tests in the 0-3.3Ghz range its within -0.16dBm on my un-calibrated E4418A
in the 3.3-6.4 range it is now off 2dBm.
So thinking about this I am still going to replace the V91 as it's voltage is very high. and at that point I will also re-run Level Preset Cal + ALC Table Cal.
and see what changes if anything.
I do plan once this part is fixed to also tune the various voltages on the unit.
here's some various images I took while working on it
This is the area i'm focusing on. the amp marked H1 is the V91
This shows N3/N4 which is after the mixer in the IF900 section.
View of top of the unit
View of bottom of the unit
Power reading at 3.301Ghz. Soon as i go above this it goes about 2dBm off
close up of the hybrid output section
close up of the hybrid input amp
Pic to remember how the connections go to the E6GHZ board
Here's N4 - This is the last amp before the Test Point that was the first one out of tolerance
Just because it looks neat
This has all the inputs/output on the back, I want to move them to the front
Question?
The rigid coax for the main output. appears long enough to be moved from the rear to the front panel. but I would have to re-bend the coax to make it work. will doing so mess anything up?
edit: made a few updates and hopefully you can see the images now.
my next planned update is after i get the v91 replaced hopefully this week (waiting on part)