Author Topic: Teardown : Keithley 2510 TEC SMU repair and calibration [complete]  (Read 13996 times)
Offline TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Here Keithley gear goes again. I probably already won title of Keithley hoarder multiple times.  :P
This time it's about TEC-oriented SMU. For those who don't know - TEC is a semiconductor module working based on Peltier principle,
which is basically describing forces heat transfer effects by using electrical P-N junctions from cold side to hot side.
This SMU capable of directly controlling 10V@5A TEC element, to a specific temperature setpoint.
Model 2510 is pretty expensive and rather specialized device to be in most of lab. But if unit is broken with diagnosis "Does not power on",
then it's tempting to get one and try to fix. I was thinking to build small temperature chamber for testing voltage references and
measuring tempco of resistors, and small PCBA's, so if I can manage to fix this TEC SMU - it will make controlling temperature for such a project an easier job.
Let's see how this evolve. I expect to receive it somewhen next week.
If anyone have experience with KI 2510, or have any internal photos - those are very much welcome!
So far I only have this photo  :)
Service manual
User manual
For calibration of Keithley 2510 we will need next standards:
Resistors
1?, ±0.1%, 100W
100?, ±1%
1k?, ±1%
10k?, ±1%
100k?, ±1%
For measurement they recommend using Keithley 2002 (which I luckily have, along with calibrated this year Model 2001).
2002's spec:
DC Voltage  +20V: ±6.8ppm
Resistance
20?: ±23ppm
200?: ±19ppm
2k?: ±7.4ppm
20k? ±7.4ppm
200k?: ±29.8ppm
Project tracker
« Last Edit: January 23, 2015, 03:25:31 pm by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
Offline Vgkid
Re: Teardown : Keithley 2510 TEC SMU repair
« Reply #1 on: November 19, 2014, 12:08:31 pm »
I liik forward to the repair.
Offline ManateeMafia
« Reply #2 on: November 19, 2014, 02:23:04 pm »
TiN,
PM sent. It was 15 minutes to get inside. Much easier than the 2001.
« Reply #3 on: November 19, 2014, 03:01:22 pm »
I uploaded my pics to your ftp server.  I need better lighting but I can take more tomorrow. Let me know if you need anything else and I can get them for you. Unfortunately, I don't have an adapter(yet) for my programmer or I would make a backup of the firmware. I will look into this as well for my GQ-4X.
Good luck.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #4 on: November 19, 2014, 03:28:00 pm »
Very nice, thanks a lot. You don't mind if I post them here as well?
Is it possible to get a closeup around U125,U172? These are Voltage references according to service manual.
U125 have just LT005 marking if I see it correctly. I wonder what that is.
U172 is near output connector and three-legged metal can current source (LM334)
Based on photos it's easy to spot Digital and ADC section borrowed from Keithely 2000 :)
Wonder if there could be way to hack current/voltage limits. Also autotuning function from 2510-AT model :)
I also have Model 2000, would be fun to see what are differences on schematics level between two.
So far my plan :
* Get a unit, check for any physical damages. Check VFD.
* Check any leaking caps, burnt/missing components, relays.
* Power unit on, check voltages, +10.4V, +4V, +6.3V, -12.5V, +12.5V, 0V, 0.5V, 5V
* Check front panel (I am 99% sure front panel from 2001 will just work fine :))
* Debug issues as they come...
« Last Edit: November 19, 2014, 03:35:41 pm by TiN »
« Reply #5 on: November 19, 2014, 04:14:12 pm »
I uploaded two more pics. You can share them here if you like. I did not see any unpopulated pads, I am willing to bet the AT is all in FW, maybe it can be unlocked. I did see the jumper on the board but it did not look like it was documented.
It definitely has LT005 on both references.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #6 on: December 04, 2014, 05:17:32 am »
Alright, I got my 2510 today.
Done..
No visual damage anywhere. Dirty outside, but everything nice inside. Front display board seem to be replaced, as datecodes on it are 2007, but mainboard is dated 1999.
Back of it:
Same board with onboard HV voltage boost is used in Keithley 2002 and 2400. 2001,7001 have HV voltage boost for VFD on their boards, not front panel board.
Face:
Display board and VFD are fine and working (supplied +5.000VDC from KI2400, VFD draws 0.45ish Amps, and showing "No Comm Link" which is what MCU on display board outputs if no serial link data received.
Done, everything seems fine visually. Relay coil is fine.
Power on - nothing is happening. No any segments lit on front panel, no reaction on output button toggle, no buzz or fuzz.
Checking test points using my KI2001 as per service manual:
J1033 front panel +5V: 4.715 VDC (below -5% specified in manual, but I think it's OK, just voltage for front panel)
J1003 p3 : 14.985 VDC (OK)
TP102 : 4.980 VDC (OK)
TP112 : 5.010 VDC (OK)
TP101 : 14.958 VDC (OK)
TP105 : -15.162 VDC (OK)
TP106 : 5.020 VDC (OK)
TP107 : 4.974 VDC (OK)
Popped firmware PLCC32 into MiniPro to read it (ATMEL AT49F040 chip), as it have sticker 2510-800-A11, and
tada...
It's blank (all bytes 0xFF)..
I flashed it with C32 firmware from Keithley 2400, now it shows No Comms Link...
So it's a "progress", meaning it might be just borked firmware...
And I need our fellow community help, a firmware dump of AT49F040 PLCC32 ROM to be exact, as mine seems to be blank (all bytes are 0xFF).
All voltages seem to be just fine (did not check ripple/noise, just measure with K2001).
Keithley website, nor their forum does not have 2510 firmware available for download :(
As a appetizer for a thread, some hi-res photos (clickable).
Overview on mainboard:
GPIB and firmware flash chip (ATMEL AT49F040).
Isolation and digital I/O
Output connector and circuitry around it. According to service manual, U172 is voltage reference. Anyone knows what that part is?
Multislope custom ADC chip leveraged from Keithley 2000.
CPU and SRAM
Infamous TF-245 resistor network, same as Keithley 2000. Some analog stuff around..
Voltage regulator circuits and front panel connector.
Voltage regs switcher
LDO's for +15. -15 and 5V.
Another switcher.
Amp controller
Voltage reference (U125), some PTF56 resistors and analog circuits.
More output circuitry..
DC-DC taking +15V from mains AC/DC and downconverting into +5V
That's it for today, hope it's just missing firmware...
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #7 on: December 04, 2014, 10:01:21 am »
Posted a request on Keithley forums :)
« Reply #8 on: December 04, 2014, 10:18:21 am »
I see you have version A11. I hope you can get an update as mine is apparently A03. I will verify it when I get home. Maybe mine was updated after is what manufactured.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #9 on: December 04, 2014, 10:21:03 am »
Well, I have none, as firmware chip is blank :)
So I need anything to get it flashed and try if unit works.
Offline krivx
  • Country: ie
« Reply #10 on: December 04, 2014, 10:51:50 am »
How could the chip be erased? Failed upgrade? AFAIK any ageing or failure of the part would cause minor corruption.
« Reply #11 on: December 04, 2014, 02:44:58 pm »
My 2510 is reporting version A03. I am looking for an adapter for my programmer. Hopefully I can get one on order for tomorrow. I will need to backup my 2001's too.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #12 on: December 04, 2014, 03:33:21 pm »
Help is appreciated! :-+
Also for 2001 I have dumps from my units available in repository here, as well as calibration EEPROMs.
A08 (regarding 2001) is the one you want, as it have fixed some TRMS operation issues, without which Tek service was unable to calibrate my 2001.
Offline Vgkid
« Reply #13 on: December 04, 2014, 04:34:54 pm »
What is the voltage reference chip?  I'm bad at figuring them out without a more useful part number.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #14 on: December 04, 2014, 08:48:38 pm »
As Linear Top markings document shows, 0425 code corresponds to LT1004CS8-2.5 20ppm/°C voltage reference zener.
So that mystery did not hold for very long :)
Offline Vgkid
« Reply #15 on: December 05, 2014, 09:01:40 am »
Thanks for that. I was wondering why ltc100 was not turning anything useful up. I thought that  the bottom number was a date code.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #16 on: December 05, 2014, 11:44:48 am »
Folks from Keithley forums shared firmware for 2510, so as expected, flashed it to ROM, everything booted up fine.
Boot message
Calibration dates
I think calibration might be corrupted, as serial number is lost?
Hooked it up with my 2002, output seems off too much
I think it's time to find good resistors and calibrate this thing :)
« Last Edit: December 05, 2014, 08:57:02 pm by TiN »
« Reply #17 on: December 05, 2014, 12:24:13 pm »
I just noticed someone posted the fw to the Keithley forum. Good job!
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #18 on: December 06, 2014, 04:53:49 pm »
Let's calibrate it...
Resistors:
1 Ohm : Vishay VPR5 1R0000 0.1% 5W 10ppm/K
100 Ohm : VPG VHP202Z 120R00 0.05% 0.2ppm/K
1 KOhm : VPG VHP202Z 1K0000 0.01% 0.2ppm/K
10 KOhm : VPG VHP202Z 10K000 0.01% 0.2ppm/K
100 kOhm : Fluke wirewound 142349 100K 0.1% 5ppm/K
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #19 on: December 07, 2014, 06:40:07 am »
Resistors characterization, +26 room temp, 10 NPLC, 8.5 digit Model 2002.
Various calibration steps, entering measured resistance values into K2510
Finish, enter new dates...
Checking now output voltage with calibrated (in Feb 2014) Keithley 2001.
REL to null output offset (2510 output ON, Set voltage 0.00000 VDC)
Checking current output source:
Repair and calibration is complete, easy score here.... :)
Now need to build little thermally insulated chamber with TEC element and RTD (I have few Honeywell HEL-705's) and ready to measure
tempco of resistors and devices for my KX project!
In worst case Model 2510 can be used as 10V 5A four-quadrant power supply :D
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #20 on: December 07, 2014, 06:44:13 am »
Video of stepping from 0V to +10V and to -10V
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #21 on: December 07, 2014, 01:18:37 pm »
ManateeMafia, is your 2510 have fan running when power on?
I suspect fan on mine is broken, as it does not operate.
But not sure if it supposed to be working all the time (like on all of my 2001's, 2002) or just turn on after reaching certain loading/temperature.
« Last Edit: December 07, 2014, 01:37:59 pm by TiN »
Offline TiN
  • Country: tw
    • xDevs.com
Re: Teardown : Keithley 2510 TEC SMU repair and calibration [complete]
« Reply #22 on: December 10, 2014, 09:54:08 pm »
Noone really probably care, but I was using 2510 last few days for some testing and gotcha few scope screenshots of
voltage and current behaviour on turn on/turn off source :)
It's pretty slow during ramps, doing it in funky steps, but when it get's there, it's pretty damn accurate :)
Ramp to 2.0000 Amps into 1.7 ohm resistor
Turn off
Step measurements..
« Reply #23 on: December 11, 2014, 09:15:15 am »
TiN,
I checked my 2510 and the fan does not seem to come on by default. Hopefully that is a good sign. Great work on the cal and graphs.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #24 on: March 04, 2015, 11:40:47 pm »
Just for completeness, combined all data into article, including enabling to 2510-AT :)
« Reply #25 on: March 05, 2015, 01:52:05 pm »
Your post reminded me that I bought a plcc adapter for my 2510 firmware.
I tried the secret menu but my rev A03 fw has no AT enable feature. I backed up my firmware and was going to flash the fw from your website until I noticed you were using the s file format.
Do you happen to have a bin file instead? I tried mot2bin.exe and it gave me a much larger file. I have not used this utility before today. Any help would be appreciated.
I will email you the A03 file when I get a chance.
« Reply #26 on: March 05, 2015, 04:24:14 pm »
Never mind, I figured it out. I was able to use my programmer even though it wasn't immediately obvious. I did find another utility that would also work. It is 010 editor.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #27 on: March 07, 2015, 02:52:39 am »
I used WinHEX.
Binary was posted before in repository as well: http://dev.xdevs.com/projects/rnd/repository/show/Keithley/2510/firmware
 :bullshit:
Thanks for A03.
I am pretty positive that you may need recalibration, so beware.
« Last Edit: March 07, 2015, 02:54:25 am by TiN »
« Reply #28 on: March 07, 2015, 10:58:43 am »
You are correct, calibration constants appear to be lost. This is not a big deal, the only caveat is that I don't have a 1ohm 100W resistor. I have the rest in one form or another.
The serial number was also lost. I found your dump of the serial eeprom and it looks like the serial number is stored in the first set of bytes. I have ordered another adapter for my programmer.
I am looking at building a resistor array using some Riedon FPR2A-10RF1 or lower ohm value across a long heat sink. I have found a few enclosures and heat sinks I could mod. I have already done this for another test jig and it turned out fine.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #29 on: March 07, 2015, 07:55:14 pm »
You don't really need 100W rated resistor, SMU does not give that much power and it's pretty quick anyway.
I used 7W ones posted in thread above, worked well, and resistor was not even heat up.
« Reply #30 on: March 07, 2015, 11:33:40 pm »
That is good to know. I have a 1 ohm vhp-4 resistor but it has a max of 3A. I will find something a little beefier without the need for a another major project.
« Reply #31 on: March 11, 2015, 04:30:22 am »
TiN,
Have you powered the 2510 through a GFCI outlet? So far, I have only had my 2510 trip the breaker. It happens almost immediately so it makes me think the SMPS inside the 2510 is causing the issue.
I only have one GFCI in the room and I will probably pick up a different adapter but I would like to know if you had any experience with this.
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #32 on: March 11, 2015, 04:55:32 am »
I had to google what is that first. :)
No, i power all my gear from standard earthed outlets, which run into standard apartment breakers. Its 110VAC 60Hz here in Taiwan.
