Another update ... and a few questions.
With the 2213 on an isolation transformer, I probed the indicated test points. Since my serial number was below B02XX, I needed to probe the circuit without U920. Points 26, 29, and 30 were spot on. Points 27 and 28 were completely opposite. In the schematic, point 27 is at the cathode of the 10V zener VR914. Point 28 is on the anode side. YET, the wave profiles show the peak of 28 to be GREATER than 27. In my case point 27 looks like 28 in the manual. Point 28 looks like 28, but with a 10 volt peak. Based on VR915, point 28 should not exceed 10 volts with respect to TP915. Error in the manual?
I am also seeing issues on points 31 and 32. instead of 80V, I see 60V. I also see something completely unexpected! Q940 and Q942 are triggering at 10kHz. The wave is not square. It is as if the transformer voltage can't keep up to achieve 80 volts. The oscope did start making a higher pitch hum when all this started...
With all of that said...
I 'sense' that Q948, Q954, and Q956 are triggering the signals at points 31 & 32. I do not see how the oscillation is happening or know enough to know if it should be 10kHz.
Considering VR915 (10V Zener) in series with VR914 (10V Zener), shouldn't point 27 have a max of 20V? Wouldn't point 28 have a max of 10V?
UPDATE: As often seems to happen after I post anything..I come across a partial answer. Section 3 discusses the theory of some of my circuit. The 10kHz is expected. The manual references 20kHz triggered by T940 and T942, but it is likely referencing point 33. I will try to upload a screenshot of points 27 & 28 if I can. The voltage across my C937 is 30V instead of the expected 42V. Current Limit Board?