I'm surprised that although you own a T922 scope you don't really know how to operate a scope.
Please read the operator's manual:
http://exodus.poly.edu/~kurt/manuals/manuals/Tektronix/TEK%202213A%20Operator.pdfTo set up the scope read page 3-3 initial adjustments.
What I saw in the video:
1. Like above you did not read user manual to set up or set the switches/controls.
2. Using XY trigger? Without input signal to x or y.
Edit: OK at XY you should see a bright dot instead of a blob .
adjust R 875 for auto focus and R 887 for astigmatism (input coupling at DC). Both adjustments will interact so have to adjust back and forth for best sharpness. Please read service manual page 5-3 onwards for full understanding. Refer to schematic 9 and also in A10 main board adjustment location.
If however it's not possible to get sharp focus then power off scope and measure the 6 resistors of 1Mohm. R878 - R882 and R884. One or more probably exceed/drifted 1Mohm +5% in value. Or suspect transistor Q877.
3. You do not know the spec. of the calibration signal.. 1Khz. 500mV p-p. As you increase volts/div waveform will shrink.
4. The scope is technically in good order except for trace need adjustment back to horizontal using the trace rotation pot at front panel of scope.
5. The cal signal does not have flat tops because the probe have not been adjusted to compensate it's capacitance. Read user or service manual how to do it.
After you have done all that is required report back here.
Edit: You have left ground clip of probe unconnected.