Hi!
I'm new here but always been a fan of Dave's videos, so I look forward to being active on this forum as well
I have got a Tektronix P6042 DC-50MHz current probe with some issues and decided to repair it.
This unit was designed in the mid 60's and it's very well constructed, almost entirely discrete-transistor based with documentation available from Tek. It is still a very useful piece of test gear as it can go from DC to 50 meg and supports currents as low as 1mA.
Forum user PA4TIM had repaired one and has very useful info on his website -
http://www.pa4tim.nl/?p=3135Original Tek manual is here -
http://www.atecorp.com/ATECorp/media/pdfs/data-sheets/Tektronix-P6042_Manual.pdfWhen I got my unit I noticed two things:
1) Probe cable was broken near the front panel strain relief rubber, this was an easy repair just shortening the cable and soldering the wires inside.
2) R11 missing, instead there was just a piece of wire there, and it was disconnected at one of the little studs, i.e. open circuit in place of R11.
R11 is a factory-selected resistor that has something to do with the temperature compensation for the hall effect probe. I tried using the same value as R13, a short and open... this has some effect on the drift of the output but doesn't seem to be related to the main problem I have.
The custom diff. amp looks OK, and now the whole system kind of works. I can adjust the cur/div balance circuits and play with the calibration trim pots inside while holding the tongue at the right angle
But there is a little problem that just drives me crazy... the P6042 output is designed to be 50-ohm terminated. This is what I do - I either use my internal TDS540's termination or an external terminator, but the reading I get from the probe is about half of what it should be! and playing with the gain calibration pot has very little effect on this issue (not much dynamic range).
When not terminating the probe's output the reading seems to be correct... or at least I can somewhat calibrate the probe to give a correct reading without being 50-Ohm loaded.
I thought maybe someone hacked a 50 ohm resistor inside for some reason, and nope.
This problem exists on both low and high frequency amps inside, I get the same "half voltage" output regardless of if it's DC or say 20 MHz input (at 50 Mhz I should expect a 3db drop).
Anyway, I'm pretty sure when it says "0.1A/Div" (when scope is set to 50mV/Div), it should be one graticule line in amplitude for 0.1A of current... not for 0.2A!!
I have measured the voltage rails and tested some points along the signal path. Also inspected the attenuator and it was NOT modified in any way.
I really can't track the problem down as I don't know what signal levels to expect along the path. All I know is that the final output is no good.
The manual doesn't mention any test points along the signal path for some given reference input. Only some basic performance checks and calibration procedures, all based on final output stage.
I welcome any ideas. Thanks in advance!
I will be glad to share more info, upload photos etc...
Omer