Interesting fault. I have a Tektronix 2402A Tekmate, which is basically a 286 PC + GPIB board in a case that can be attached to the top of the Tektronix 2430/2440 digital scopes.
Now this one seems to have a faulty motherboard. The board is an Ampro A60220 LittelBoard 286. And it does not really seem to do anything. I have removed all boards, and ruled out the PSU (this can be run off 5V, and did run from a bench PSU).
I see that the CPU starts executing, as there is activity on the bus, but no beeps, no, screen, nothing. After some investigation I found out the the CPU clock is not good. It should be 16MHz, but sometimes it is 32Mhz (that is the main oscillator) or rapidly changing, or has jitter between the two values. So more or less I have concluded that the whatever is responsible for generating the system clock from the 32MHz is faulty. This is likely to be somewhere in the chipset, as it should be software configurable to run on 16 or 8Mhz, but looks like the first divide by 2 of the 32MHz does not always work.
I even tried removing the 32Mhz oscillator and feeding in 16MHz. In this case the clock was sometimes 16, sometimes 8 Mhz seemingly random. So this is not a deterministic fault, and I have no clue how this may happen.
I'm aware that I could just swap the motherboard or use any old PC, I'm not interested in that. I take this as a challenge to see if I can fix the board. Anyone familiar with the internals of this motherboard? I have the Ampro and the Tecktronix manual, but those do not go down to details wrt the motherboard internals.
Any ideas?