Have had THS720A scope since new and it has worked without incident until recently. Believe a portion of the power supply has been compromised but please give other opinions. Some details: Power from 110VAC charger is 14VDC, voltage at NiCad battery terminal is 12VDC. Battery was in poor condition and I'd operated the scope without the battery installed using only the power supply. Channel 1 shows no response but appears on screen as open channel, Channel 2 shows waveform but voltage magnitude is wrong--too low in amplitude. Frequency and scaling (both axes) is correct and can be changed with associated switches. A 9VDC battery attached to DMM terminals returns 4.5VDC when this mode is activated. Battery voltage was confirmed using a separate handheld meter. Tested scope probes on another instrument without issue.
There are 4 electrolytic caps mounted in main board power supply region. Have not replaced these but there is no evidence of deformation or leakage. No components appear damaged and no evidence or arc tracks. Was using scope on 10VDC signal, channel 1 only when failure occurred.
Don't quite understand how this power system should work. Voltage to battery seems very high and shouldn't a negative supply be generated? Thanks in advance for your thoughts on troubleshooting.