Are you sure: The chip is toasted? I`m just curious.
If the misfunction shows wrong ranges (e.g. set to V, it displays Hz), then it´s not the ES51922.
Hi,
I'm curious now, because that is exactly the case! On several positions the range is wrong and I found a workaround with a manual switch to pull-low or high FC1 of the Cryustek to 3V in order to put it into the desired range when it was not. But then in the other ranges the multimeter switches off. Also this worked but is results in a high battery drain.
Originally I found the two protection diodes D1 and D3 had to be replaced and did so. This could speak for a current overload as origin of the error.
Next thing was the wrong ranges. So I found that FC1 was not was it's supposed to be according to the switch position (Ref Table in the chematics):
Volt => OhmsmV => FaradOhms => Ohms
Hz => Hz
yA => Ohms
mA => OhmsA => A (OK)
So I do not get the FC1 to be High
But in the positions where the range is faulty the FC1 is not closed to another contact.
FC1 is connected through the double diode D4 that tested 0.3V forward and OL in reverse on both diodes. I Guess the diodes are working.
After that I'm losing the overview..
Ps: The contacts in the range switch look ok and are all there.
What is your experience with the wrong ranges problem?